  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #16
    Originally posted by GenoMax View Post
    Did you try looking at a previous run (since you had said that the most recent run was analyzed off-line by CASAVA)?
    There are no runs at all listed in the Analyses tab. There should be 5, I think -- at least that is the number of runs in that folder if I look at the file system.

    --
    Phillip

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #17
      Firing up IE opened up the "isis viewer" automatically. Once I clicked on the analysis tab lots of previous runs were listed, with the "re-queue" option.

      BTW: Illumina released a new update for the MCS software (v.2.1.x) yesterday.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #18
        Yes, if any runs were displayed in isus view I am confident I could "re-queue" them. Obviously there is a missing file, or something is mis-configured. Guess I could contact tech support...

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #19
          Okay, maybe the information I need would be:

          MiSeq Reporter v2.1 Theory of Operation.

          You will need to have a myIllumina account to view...

          --
          Phillip

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #20
            Okay, I got it. I needed to delete the "queuedforanalysis.txt" file from the run folder. Then it will immediately begin processing. (I had put the new sample sheet in the run folder.)

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #21
              By the way, as far as the dual + single indexed libraries demultiplexing via MCS/RTA, it does work:



              The first 3 libraries are single index genomic DNA libraries. Also included an amplicon set (generated by a customer) with 96 samples using the "TruSeq Custom Amplicon" dual indexes.

              Oh, just in case:

              Oligonucleotide sequences © 2007-2012 Illumina, Inc. All rights reserved.

              The MiSeq happily demultiplexed all 99 index pairs at a >97% efficiency.

              --
              Phillip

              Comment

              • Mosiep
                Junior Member
                • May 2014
                • 3

                #22
                Is it possible to combine/mix samples together for a single run in 454 sequencing? is it practical ?if so what are the benefits? I am new to 454 sequencing

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #23
                  Originally posted by Mosiep View Post
                  Is it possible to combine/mix samples together for a single run in 454 sequencing? is it practical ?if so what are the benefits? I am new to 454 sequencing
                  454? This is an Illumina thread.

                  --
                  Phillip

                  Comment

                  • Jessica_L
                    Senior Member
                    • Feb 2010
                    • 117

                    #24
                    I'm trying to do something similar. I added a single indexed library (6bp P7 index) into a pool of tru seq amplicon libraries (8bp P7 index) and need to demultiplex. I tried putting two "N"s at the end of the 6bp index (i.e. GATCAGNN) on the sample sheet but that doesn't appear to have demultiplexed correctly. Or should it have and maybe the library failed to cluster? It's entirely possible the library failed library construction (colleague generated the library; it was his first ever).

                    I'm thinking I'll try demultiplexing via CASAVA when the dataset is done just to be sure I set the indexes correctly in the sample sheet. I don't expect MSR to have much for me analysis-wise to requeue since this was a generate fastq workflow.

                    Does anyone have any thoughts?
                    Last edited by Jessica_L; 10-22-2015, 06:46 AM. Reason: Fixed a typo.

                    Comment

                    • JBKri
                      Member
                      • Jan 2014
                      • 80

                      #25
                      Originally posted by Jessica_L View Post
                      I'm trying to do something similar. I added a single indexed library (6bp P7 index) into a pool of tru seq amplicon libraries (8bp P7 index) and need to demultiplex. I tried putting two "N"s at the end of the 6bp index (i.e. GATCAGNN) on the sample sheet but that doesn't appear to have demultiplexed correctly. Or should it have and maybe the library failed to cluster? It's entirely possible the library failed library construction (colleague generated the library; it was his first ever).

                      I'm thinking I'll try demultiplexing via CASAVA when the dataset is done just to be sure I set the indexes correctly in the sample sheet. I don't expect MSR to have much for me analysis-wise to requeue since this was a generate fastq workflow.

                      Does anyone have any thoughts?
                      I believe it should work if you add "AT" at the end of the 6 bp index and repeat the demultiplexing in MiSeq Reporter.
                      Or you can leave the index with only the 6 bases (remove the "NN) and demultiplex; this should demultiplex those samples. Then, if needed, you can do another demultiplexing for the amplicon libraries, using 8 bases in the sample sheet.

                      Comment

                      • Jessica_L
                        Senior Member
                        • Feb 2010
                        • 117

                        #26
                        Originally posted by JBKri View Post
                        I believe it should work if you add "AT" at the end of the 6 bp index and repeat the demultiplexing in MiSeq Reporter.
                        Or you can leave the index with only the 6 bases (remove the "NN) and demultiplex; this should demultiplex those samples. Then, if needed, you can do another demultiplexing for the amplicon libraries, using 8 bases in the sample sheet.
                        Thanks so much-- I'll give it a try!

                        Comment

                        • JuNimey
                          Junior Member
                          • May 2017
                          • 2

                          #27
                          I found this thread since I was searching for sequencing Nextera and TruSeq LT libraries on one lane in a MiSeq. I only need the i7 read for the nextera libraries and the TruSeq libraries I have are modified and contain 8-base barcodes (as the Nextera libraries do as well). Can you give me an advice on what to enter in the sample sheet, so that this will work properly? It's my first time sequencing on a MiSeq and I would appreciate any help! Thanks a lot!

                          Comment

                          • thermophile
                            Senior Member
                            • Apr 2015
                            • 243

                            #28
                            In IEM, select truseq single index to see how the sheet should be laid out. Then use that header and column names for the sheet you put together
                            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                            Comment

                            • GenoMax
                              Senior Member
                              • Feb 2008
                              • 7142

                              #29
                              For reference. IEM = Illumina experiment manager (windows only).

                              Comment

                              • thermophile
                                Senior Member
                                • Apr 2015
                                • 243

                                #30
                                oh didn't realize that.

                                if poster isn't windows, I can search for an example sample sheet
                                Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                                Comment

                                Latest Articles

                                Collapse

                                • GATTACAT
                                  Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                                  by GATTACAT
                                  Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
                                  07-01-2026, 11:43 AM
                                • SEQadmin2
                                  Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                                  by SEQadmin2


                                  I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                                  Here are nine questions we think about, in roughly the order they matter, before...
                                  06-18-2026, 07:11 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, Yesterday, 11:08 AM
                                0 responses
                                6 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 06-30-2026, 05:37 AM
                                0 responses
                                11 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 06-26-2026, 11:10 AM
                                0 responses
                                19 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 06-17-2026, 06:09 AM
                                0 responses
                                53 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...