    I have the same problem here

    Comment

    • yaximik
      Senior Member
      • Apr 2011
      • 199

      #17
      This may happen if in previous (erroneous) run Z motor hit the bottom of the moving range and cannot recover. Quitting an restarting the software will reinitialize everything including homing Z motor.

      Comment

      • matth431
        Member
        • Jan 2012
        • 36

        #18
        I just had a similar "Object reference" failure about 185bp into read 1. In our case, cluster density was very low (Nextera XT normalised libraries pooled and diluted as per instructions but according to user, library concentration prior to going into normalisation step sounds low - 1-5ng/ul compared to 20-30ng/ul with libraries I've prepared myself).

        Density was only 133K/mm2 which is way lower than our worst run to date (min 700k/mm2). We typically run 900-1300k/mm2.

        Thanks for the tip on getting the MiSeq to analyse the data from the failed run, though - user wants to have a look at it although without the index reads I suspect it will be fairly meaningless.

        Comment

        • yaximik
          Senior Member
          • Apr 2011
          • 199

          #19
          Did you try to restart the run using the same cell and reagent set? It worked several times, I just used shorter run times and even got paired read completed. The reads have a lot of Ns and low quality regions, but with aggressive cleaning substantial set of useful data can be recovered. My feeling is sometimes BioAnalyzer traces are misleading in regard to average library size as contamination with shorter fragments can be significant yet not clearly visible. That results in a large portion of clusters being shorter then anticipated and vertical focus unexpectedly fails hitting the floor. That is usually signified by steep declines in the signal intensities on the graphs.

          Comment

          • matth431
            Member
            • Jan 2012
            • 36

            #20
            Originally posted by yaximik View Post
            Did you try to restart the run using the same cell and reagent set? It worked several times, I just used shorter run times and even got paired read completed. The reads have a lot of Ns and low quality regions, but with aggressive cleaning substantial set of useful data can be recovered. My feeling is sometimes BioAnalyzer traces are misleading in regard to average library size as contamination with shorter fragments can be significant yet not clearly visible. That results in a large portion of clusters being shorter then anticipated and vertical focus unexpectedly fails hitting the floor. That is usually signified by steep declines in the signal intensities on the graphs.
            I wasn't aware it was possible to restart a failed MiSeq run? Surely it rejects the cartridge and flowcell based on the RFID tags?

            Comment

            • yaximik
              Senior Member
              • Apr 2011
              • 199

              #21
              Here is my earlier post how to do this.

              Bridged amplification & clustering followed by sequencing by synthesis. (Genome Analyzer / HiSeq / MiSeq)


              If the run has aborted during read 1, then reagents are OK. However, if run terminated normally, either SE or PE, then reagents cannot be reused as MiSeq spits back something (backflush?) and ruins remaining reagents. What a waste!
              Last edited by yaximik; 03-18-2013, 07:15 AM.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM
              • SEQadmin2
                New Genomics Technologies Take Aim at Long-Standing Limits
                by SEQadmin2


                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                ...
                09-28-2026, 10:25 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Yesterday, 11:13 AM
              0 responses
              19 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, Yesterday, 10:48 AM
              0 responses
              18 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-15-2026, 10:52 AM
              0 responses
              36 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-14-2026, 05:39 AM
              0 responses
              42 views
              0 reactions
              Last Post SEQadmin2  
              Working...