  • Loss of entire sample set after SPRI cleanup

    Hello all,

    I was prepping samples for WGS Using the Kapa HyperPrep workflow. After the final cleanup with 0.8x AMPure SPRI beads, my gel confirmation showed that most of my samples were lost.

    I'm trying to mentally troubleshoot what could have gone wrong, but I don't have extensive experience with library prepping to know where to look for the problem.

    I washed the samples+beads in 80% EtOH twice for 30s and eluted in 10mM TE+0.05% Tween20 buffer warmed to 55°C. I air dried the beads for 2 minutes between the wash and elution. I incubated in buffer for 3 minutes before removing the beads and used a plate shaker to assist with DNA re-suspension. The cleanup was done with a bead extractor to minimize sample loss.

    I redid the cleanup using pre-warmed 10mM Tris-HCl + 0.05% Tween20 instead of TE as the buffer, but the outcome was the same.

    Does anyone have suggestions on what could have happened to effect the affinity of the DNA and beads?

    Any ideas are welcome.
    Last edited by AvantS; 05-23-2019, 10:42 AM. Reason: Clarification
    ƧΛVΛПƬ

  • #2
    How fresh was the 80% Ethanol, if ethanol evaporated off you could have had elution during the washes. It is recommended that 80% Ethanol is not more than a week old.

    Comment


    • #3
      @SeqSuave. It was made on the same day as the bead cleanup.
      ƧΛVΛПƬ

      Comment


      • #4
        I am assuming you are doing the PCR-free workflow? What was the input concentration and the concentration after final clean-up. You might have to increase your input concentration or do a PCR step to amply your library.

        Also during the drying step if the beads are over-dried and you see hairline cracks, I would incubate the beads for 5 - 6 minutes to help increase yield.

        Comment


        • #5
          @itstrieu The workflow includes 3 PCRs: End repair & A-tailing PCR, Adapter Ligation, and an Indexing. Input concentration was 100ng/uL per sample. Quantification after cleanup was not done, since the gel-confirmation showed few successful sample bands.
          ƧΛVΛПƬ

          Comment


          • #6
            Is the workflow

            ER,AT, Ligation --> 0.8X bead cleanup --> Elute in 25 uL follow by Library Amplification --> Post-amplification Cleanup using 1x bead? With the concentration being 100ng/uL, how much total DNA is going into library prep?

            Was their a QC check after fragmentation? One common mistake I see is using elution buffer/water during the wash step instead of ethanol which could wash away your DNA. Not on purpose but accidental since both on them are clear liquids.
            Last edited by itstrieu; 05-24-2019, 10:54 AM.

            Comment


            • #7
              Originally posted by itstrieu View Post
              Is the workflow

              ER,AT, Ligation --> 0.8X bead cleanup --> Elute in 25 uL follow by Library Amplification --> Post-amplification Cleanup using 1x bead? With the concentration being 100ng/uL, how much total DNA is going into library prep?

              Was their a QC check after fragmentation? One common mistake I see is using elution buffer/water during the wash step instead of ethanol which could wash away your DNA. Not on purpose but accidental since both on them are clear liquids.
              Our post-amp cleanup used 0.8x bead. The samples will be diluted to 20nM for sequencing.

              It's not possible that the elution buffer was used instead of EtOH because each U-bottom plate used for cleanup was labeled with the reagent it contains (Sample, EtOH W1, EtOH W2, Etc)and I added the warmed elution buffer to its plate while air drying the beads.
              ƧΛVΛПƬ

              Comment


              • #8
                The only two things I can think of is DNA in buffers containing high concentrations of EDTA, other chelating agents or salts may inhibit the
                end repair and A-tailing reaction. The other is something going wrong with library amplification (maybe your index primers).

                If you can, check the concentration or run a gel/tape of the adapter-ligated library to see if the yield is reasonable going into library amplification.

                Comment


                • #9
                  Originally posted by itstrieu View Post
                  The only two things I can think of is DNA in buffers containing high concentrations of EDTA, other chelating agents or salts may inhibit the
                  end repair and A-tailing reaction. The other is something going wrong with library amplification (maybe your index primers).

                  If you can, check the concentration or run a gel/tape of the adapter-ligated library to see if the yield is reasonable going into library amplification.
                  Thank you for the great advice. I'll run a gel confirmation to see if my adapter ligated libraries still have sample.
                  ƧΛVΛПƬ

                  Comment


                  • #10
                    I've used Kapa Hyperplus 3-4 times a week for well over a year and what causes me to decrease yield or lose samples have almost always been the the frag enzyme activity or sometimes beads.

                    Fragmentation: Under or over fragmentation will definitely mess up your yield. Make sure your kit is not expired, enzymes are always stored on ice and never vortexed, and buffers have no precipitation remaining. Kapa is also super temperature sensitive so maybe there was an issue there with frag enzyme activity? If i had to bet, I would say this is probably what happened to you.

                    Beads: Make sure the beads are well suspended and at the correct temperature. I personally elute in 10 mM Tris, ph 8.0 and think tween just gets in the way. I think its a good idea to never use TE for library preparation.

                    Comment


                    • #11
                      Originally posted by beancat View Post
                      I've used Kapa Hyperplus 3-4 times a week for well over a year and what causes me to decrease yield or lose samples have almost always been the the frag enzyme activity or sometimes beads.

                      Fragmentation: Under or over fragmentation will definitely mess up your yield. Make sure your kit is not expired, enzymes are always stored on ice and never vortexed, and buffers have no precipitation remaining. Kapa is also super temperature sensitive so maybe there was an issue there with frag enzyme activity? If i had to bet, I would say this is probably what happened to you.

                      Beads: Make sure the beads are well suspended and at the correct temperature. I personally elute in 10 mM Tris, ph 8.0 and think tween just gets in the way. I think its a good idea to never use TE for library preparation.
                      That sounds very likely to me as well. My team and I will redo the pipeline keeping those suggestions in mind. Thank you.
                      ƧΛVΛПƬ

                      Comment


                      • #12
                        Originally posted by beancat View Post
                        I've used Kapa Hyperplus 3-4 times a week for well over a year and what causes me to decrease yield or lose samples have almost always been the the frag enzyme activity or sometimes beads.

                        Fragmentation: Under or over fragmentation will definitely mess up your yield. Make sure your kit is not expired, enzymes are always stored on ice and never vortexed, and buffers have no precipitation remaining. Kapa is also super temperature sensitive so maybe there was an issue there with frag enzyme activity? If i had to bet, I would say this is probably what happened to you.

                        Beads: Make sure the beads are well suspended and at the correct temperature. I personally elute in 10 mM Tris, ph 8.0 and think tween just gets in the way. I think its a good idea to never use TE for library preparation.
                        Hi,

                        Interestingly I had a similar problem with the HyperPlus, my post can be found here:
                        Techniques and protocol discussions on sample preparation, library generation, methods and ideas


                        beancat, since you are an experienced user of KAPA Hyperplus I would like to ask you what was your average pre-capture yield? I got an average yield of 500 ng, however the protocol says we should get more than 1000 ng. I was also thinking I was loosing sample during the SPRI clean ups, I mean using the same 100ng input with other kits I get higher pre-capture yields.

                        In anycase I will be following this thread for any updates.
                        Thanks!

                        Comment


                        • #13
                          Originally posted by jmc887 View Post
                          Hi,

                          Interestingly I had a similar problem with the HyperPlus, my post can be found here:
                          Techniques and protocol discussions on sample preparation, library generation, methods and ideas


                          beancat, since you are an experienced user of KAPA Hyperplus I would like to ask you what was your average pre-capture yield? I got an average yield of 500 ng, however the protocol says we should get more than 1000 ng. I was also thinking I was loosing sample during the SPRI clean ups, I mean using the same 100ng input with other kits I get higher pre-capture yields.

                          In anycase I will be following this thread for any updates.
                          Thanks!
                          So I cut my reaction volume considerably with a lot of other deviations from the normal protocol (including PCR-free) so I don't think a direct comparison is useful. The only thing I can say is that it performs a lot better (nearly 25% better) than Qiagen's QIAseq FX kit with the same template input.
                          Last edited by beancat; 06-25-2019, 05:52 AM.

                          Comment

                          Latest Articles

                          Collapse

                          • seqadmin
                            Strategies for Sequencing Challenging Samples
                            by seqadmin


                            Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                            03-22-2024, 06:39 AM
                          • seqadmin
                            Techniques and Challenges in Conservation Genomics
                            by seqadmin



                            The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

                            Avian Conservation
                            Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
                            03-08-2024, 10:41 AM

                          ad_right_rmr

                          Collapse

                          News

                          Collapse

                          Topics Statistics Last Post
                          Started by seqadmin, Yesterday, 06:37 PM
                          0 responses
                          8 views
                          0 likes
                          Last Post seqadmin  
                          Started by seqadmin, Yesterday, 06:07 PM
                          0 responses
                          8 views
                          0 likes
                          Last Post seqadmin  
                          Started by seqadmin, 03-22-2024, 10:03 AM
                          0 responses
                          49 views
                          0 likes
                          Last Post seqadmin  
                          Started by seqadmin, 03-21-2024, 07:32 AM
                          0 responses
                          67 views
                          0 likes
                          Last Post seqadmin  
                          Working...
                          X