I would trust the size indicated by gel. Three issues with Tape:

1- Tape has expired and sometimes that will affect the sizing accuracy
2- The used Tape (HS D5000) accuracy for sizing short fragments is very low. Those should be run on HS D1000 tape.
3- Markers on some is misidentified (not the main issue in this case)

Last edited by nucacidhunter; 10-27-2017 at 05:17 PM.