  • TruSeq RNA PCR Cycle Number

    Hello,

    I'm seeing two size populations in my final RNAseq libraries, one at ~280 bp which is the expected library size and a larger one at ~700 bp which isn't expected.

    I start with 500ng of intact total RNA and follow the TruSeq protocol with 15 cycles of PCR. When I back the number of cycles down to ten the larger sized population disappears (see picture).

    Has anyone else experienced this?

    Thanks!
    Attached Files
    Last edited by little_henry; 04-26-2016, 11:37 AM.

  • #2
    Hi,

    the extra peak is probably due to overamplification which can give you ssDNA that migrates slower than dsDNA. See e.g. slide 30 here: http://www.mbl.edu/jbpc/files/2014/0..._slideshow.pdf

    Comment


    • #3
      Well, probably not completely ssDNA.

      If the PCR reaction runs low on primers and high on products, then amplicons annealing back to each other becomes kinetically favorable. But since this is a library then the likelihood is that the inserts of the two annealling amplicons will be unrelated. So you end up with "bubble" products. Double-stranded at the ends (where the adapters are) but with little or no annealing in the middle.

      These floppy bubble products apparently migrate slowly and so appear to be larger than they actually are. Happily you can pretty much ignore them as they cluster and sequence fine. (Although they can play havoc with your cluster titrations if you don't understand what they are.)

      This probably the single most common artifact posted to this forum! But in recent years one only sees new posts about it a couple of times per year.

      --
      Phillip

      Comment


      • #4
        The Illumina Truseq protocol must be written horribly - almost all of our customers using the Truseq RNA-seq kit for the first time totally over-amplify their libraries (15 cycles).

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM
        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        18 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        22 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        16 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        47 views
        0 likes
        Last Post seqadmin  
        Working...
        X