    Hi,

    I would appreciate any help in regard to the following MySeq error:

    Timeout of 70.000 seconds exceeded while waiting for FPGA command "CAMDBLSNAP 390 390 188 188 34390 150" to complete.

    see image:



    The error emerges any time after the run is started and ruins the run. It first appeared a year ago and now the fifth time since.

    The only advice we are continually getting from Illumina support is that we should reboot our myseq prior to the run (we are doing that since the first time the error occurred) and that the error is "very unlikely to repeat".
    Unfortunately, the error is repeating and keeps ruining runs.

    The MySeq's warranty has expired.

    If anybody had experienced the problem himself or has any information regarding the error (what does the code mean?), we would be very happy to hear from you.

    Thank you,
    Lovro

  • #2
    If your MiSeq is not under a maintenance contract then you are living dangerously (or you just don't use it enough to justify the contract).

    This error may be due to some sort of hardware issue that can only be fixed by a visit from an Illumina engineer. You may have to bite the bullet and pay for that visit.

    Comment


    • #3
      What Genomax said.

      I'd add that if you don't have your MiSeq under a maintenance agreement, you may want to consider getting one first, as the service visit and repair can easily exceed the cost of a contract. There are usually some hoops to jump through, but that's probably the cheapest way to get your instrument fixed.

      An initial visit with an FSE might be a good idea, at least, and if you have a decent relationship with them, they could give you a ballpark idea of how much repairs might cost and how that would compare to the cost of a service agreement.

      Comment


      • #4
        Hi, thanks for the answers.
        We have considered the latter.
        However, I was more interested in finding out the meaning of the error code.

        We had a hardware problem with a moveable mirror part (bouncing while snapping into position) while the machine was still under warranty. It was replaced by illumina and might still be under warranty.
        "CAMDBLSNAP" sounds a lot like some part of the camera snapping into position, so I want to investigate the issue.

        The error codes descriptions might be impossible to get a hold of though...?

        tnx

        Comment


        • #5
          Your best bet on that score is probably your field support from Illumina, for a MiSeq probably your FAS. The error codes aren't closely guarded secrets, but I'm not aware of a collection of codes with descriptions that's accessible in any kind of searchable format.

          Comment


          • #6
            CAMDBLSNAP is an FPGA or cabling issue on one of the 2 USB cables to the cameras internally. You can try re-seating the USB cables to the cameras to prevent/reduce the frequency of these errors. The ones you want to re-seat are on the inside of the MiSeq, facing the front of the instrument, up on the top right (behind the monitor), readily accessible when you take the top skin/panel off.

            Comment


            • #7
              Thank you very much. I will mention this to their representative.

              Comment

              Latest Articles

              Collapse

              • seqadmin
                Essential Discoveries and Tools in Epitranscriptomics
                by seqadmin


                The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist on Modified Bases...
                Yesterday, 07:01 AM
              • seqadmin
                Current Approaches to Protein Sequencing
                by seqadmin


                Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                04-04-2024, 04:25 PM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by seqadmin, 04-11-2024, 12:08 PM
              0 responses
              39 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 04-10-2024, 10:19 PM
              0 responses
              41 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 04-10-2024, 09:21 AM
              0 responses
              35 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 04-04-2024, 09:00 AM
              0 responses
              55 views
              0 likes
              Last Post seqadmin  
              Working...
              X