Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • MiSeq Error stopped run

    I started a run on the MiSeq on Friday and by Monday morning I noticed that an error caused the MiSeq to stop just less than 2 hrs into the run. The error saids "No useable signal found in the images; it is possible that clustering has failed." Has anyone recieved this error before? I was running a few Nextera XT libraries. I have used this procedure before with success. This error is a first.

  • #2
    Have you checked images in the "Thumbnail_Images/L001/C1.1" directory to see if there are clusters?

    Comment


    • #3
      If you spike in PhiX you will be able tell if it's a chemistry or instrument error. Without spiking it is difficult to troubleshoot.

      Comment


      • #4
        No thumbnail images were created and I didn't use PhiX. What I can tell you is that the machine went under rountine maintenance and a replacement of the motor that controls the focus mirror. The machine worked very well after the service job, but we used a non-expired flowcell and reagents. For this run we used an expired flowcell (june 6th expiry) and expired reagent cartridge. Do you think that the expired flowcell may have caused this level of failure?

        Comment


        • #5
          It may be best to open a ticket with Illumina tech support so they can remote-in to look at this run.

          If there was a hardware failure it should show up in the "Logs" directory. You could look through those files (*.log) while you wait to hear back from Illumina.

          Expired kit (unless it was mishandled in terms of storage) should not cause a total failure.

          Comment


          • #6
            The reagents should still have been fine even though they were "expired".

            I had this problem once with a run, and the issue was ultimately chalked up to a "bad flowcell". That excuse gets thrown around a bit by Illumina when they don't have any other explanation, but they'll usually replace the kit so you can re-run the libraries.

            As was noted above, you should always have phiX spiked into your samples so you can troubleshoot more effectively. Right now you don't know if your libraries or the kit or the MiSeq were at fault, although my suspicion would be the libraries.

            Did you do any library QC before trying to run them or did you just follow through the XT protocol? We use XT extensively (no pun intended), and we've had issues when following the denaturing and normalization parts of the protocol. We've since switched to stopping after the post-PCR cleanup and doing Qubit and Bioanalyzer to QC the library, followed by standard pooling and denaturing like these were standard libraries. Since we've switched to doing things this way, we've been getting more even pooling and have more control over clustering.

            Comment


            • #7
              I contacted Illumina and they had me run a number of diagnostic tests to make sure there was nothing wrong with the MiSeq. All tests passed. Therefore I was told that these errors are caused by the presence of bubbles at the wrong place and time. For example, a bubble present near the inlet port, where the optical system attempts to find focus, of the FC can cause focusing problems and lead to a premature run stop. Apparently, these errors do occur from time to time.

              Thanks everyone for your help.

              Comment

              Latest Articles

              Collapse

              • seqadmin
                Strategies for Sequencing Challenging Samples
                by seqadmin


                Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                03-22-2024, 06:39 AM
              • seqadmin
                Techniques and Challenges in Conservation Genomics
                by seqadmin



                The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

                Avian Conservation
                Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
                03-08-2024, 10:41 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by seqadmin, Yesterday, 06:37 PM
              0 responses
              8 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, Yesterday, 06:07 PM
              0 responses
              8 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 03-22-2024, 10:03 AM
              0 responses
              49 views
              0 likes
              Last Post seqadmin  
              Started by seqadmin, 03-21-2024, 07:32 AM
              0 responses
              67 views
              0 likes
              Last Post seqadmin  
              Working...
              X