The nanodrop number is subject to inflation based on other molecules in your sample (salts, free nucleotides, short oligos) (like any spectrophotometer).

How are you cleaning up your sample, and what is the concentration range you're in?

Overall, I have found the bioanalyzer much more accurate for quantitation prior to ePCR.