  • Anyone had success with Illumina V2 Rapid 2x250?

    Our institute is having trouble completing these long runs. They seem to die towards the end of both reads, camera goes out of focus, etc. They are being run on an upgraded HS2000. FYI, the samples all passed on MiSeq 2x300. We are looking for slightly more yield meta, amp.Has anyone had similar issues, any success, any advice?

  • #2
    I guess it depends on how good of success you're after. We had a 2x250 v2 rapid run finish this morning and the results were relatively good. There was decreased quality toward the end of each read, but this was due primarily to two tiles (I think they were the exact same ones for each of the reads), so it turned out that the vast majority of reads were quite good. I should note that there was a bigger issue with the end of read 2, where a large block of tiles had problems. The latter issue combined with the same tiles showing some quality problems in both read 1 and 2 would be compatible with a camera problem, since there was an obvious per-tile difference...thereby suggesting that the problem isn't due to the chemistry itself.

    I can't give any advice, as I don't personally deal with the sequencer, but that's at least the sort of thing that we're getting at the moment.

    Comment


    • #3
      With an n=1 we had no problems. Got about 250M reads per lane.

      Comment


      • #4
        Are you guys using a native 2500 V4 capable instrument, or an older upgraded 2000?

        Comment


        • #5
          Our run was on an upgraded older 2000.

          Comment


          • #6
            Originally posted by GenoMax View Post
            With an n=1 we had no problems. Got about 250M reads per lane.
            So, 125M pass filter clusters per lane, right?

            --
            Phillip

            Comment


            • #7
              Originally posted by pmiguel View Post
              So, 125M pass filter clusters per lane, right?

              --
              Phillip
              Correct. ~140 M per lane. 79% > Q30.

              Comment


              • #8
                I hope the problem has been solved by now!
                Anyway: With 2 native HS2500 we regularly get >300M cpf per flowcell and between 80 and 85% Q30 for amplicon data.

                Comment


                • #9
                  We've given up for the time being in favor of V1 rapids. Our instrument is an older one that we bought in 2011 and then upgraded to 2500. We're thinking that the camera/optics aren't capable of running 500 cycles?

                  Comment

                  Latest Articles

                  Collapse

                  • seqadmin
                    Current Approaches to Protein Sequencing
                    by seqadmin


                    Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                    04-04-2024, 04:25 PM
                  • seqadmin
                    Strategies for Sequencing Challenging Samples
                    by seqadmin


                    Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                    03-22-2024, 06:39 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by seqadmin, 04-11-2024, 12:08 PM
                  0 responses
                  22 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-10-2024, 10:19 PM
                  0 responses
                  24 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-10-2024, 09:21 AM
                  0 responses
                  19 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-04-2024, 09:00 AM
                  0 responses
                  50 views
                  0 likes
                  Last Post seqadmin  
                  Working...
                  X