  • Ladder-like bioanalyzer trace on supposed 150bp DNA sample library

    I recently started working on very old hyde tissue, extracting DNA, running a HS Bioanalyzer chip. My results showed that the samples were heavily degraded down to 30bp to 50bp fragments. When I made the libraries, theoretically I should see 130bp to 150bp fragments with the adapters attached. I did a PCR amplification to enrich the libraries. I didn't size select because there was no larger pieces than 50bp in the original bioA run.

    However, when I ran the second HS bioA chip with these libraries, I got this strange ladder-like read on my samples with extremely high basepairs.

    Has anyone else experienced this same strange result?

    Click image for larger version

Name:	Lampshade_Page_12.jpg
Views:	1
Size:	55.4 KB
ID:	307782

  • #2
    First, you bioA chip results are hosed in some of the lanes. You can see the fuzzy 10.38kb upper marker has not been correctly identified in several of the lanes. You need to view the .xad file with 2100 Expert software and go to each of those traces, switch to the peak table tab and figure out which peak is your real 10.38 kb band, right-click on it and choose "manually set upper marker". This will give you the correct size of your peaks.

    That said, looks to me like the right four lanes have 2 different MW markers in them. Either you samples are contaminated with them, or the pins of your bioA were contaminated with them.

    That is my take, anyway.
    --
    Phillip

    Comment

    Latest Articles

    Collapse

    • seqadmin
      Current Approaches to Protein Sequencing
      by seqadmin


      Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
      04-04-2024, 04:25 PM
    • seqadmin
      Strategies for Sequencing Challenging Samples
      by seqadmin


      Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
      03-22-2024, 06:39 AM

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by seqadmin, 04-11-2024, 12:08 PM
    0 responses
    18 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 10:19 PM
    0 responses
    22 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-10-2024, 09:21 AM
    0 responses
    17 views
    0 likes
    Last Post seqadmin  
    Started by seqadmin, 04-04-2024, 09:00 AM
    0 responses
    49 views
    0 likes
    Last Post seqadmin  
    Working...
    X