  • Weird bioanalyzer results

    Hey all,

    Yesterday, for the first time ever i was about to slam my head into the bioanalyzer. I ran 8 samples prepared with nexteraxt protocol after pcr cleanup and size selection using 29µL of ampure xp beads. The results did not make ANY sense at all. The presence of HUGE fragments that are even bypassing the upper ladder! The concentration of the samples on qubit was the following:
    Sample 1 = 2.05
    Sample 2 = 2.55
    Sample 3 = 3.96
    Sample 4 = 7
    Sample 5 = 3
    Sample 6 = 4.75
    Sample 7 = 6.68
    Sample 8 = 7.3

    if those HUGE fragments are truly present, than the concentration doesnt really correlate with them does it?
    Has anyone seen this phenomena using the hs DNA chip?

    Best,
    Exo
    Attached Files

  • #2
    Hi!
    It might simply be that you have some contaminants that are driving the Bioanalyzer crazy. We see this kind of weird things all the time and in our case is due to traces of ethanol left over from the bead purification or to an excess of salt in the elution buffer (HS chips tolerate a max of 10 mM Tris as buffer, according to the specs). Also some leftover detergent from the previous steps can interfere and screw up the chip.
    Since all the samples are run on the same lane of the Bioanalyzer, when one goes bad the others also tend to go bad, especially the one that is immediately after the bad one.
    Good luck with your troubleshooting!

    Comment


    • #3
      Hi,

      In a Illumina webminar they showed something like this one, https://www.dropbox.com/s/eom2x2u78i...2022.45.44.png

      You can try to clean-up again, but if its an artifact will stay there.
      Last edited by DRYTCYV; 03-13-2014, 02:59 PM. Reason: missing image

      Comment


      • #4
        The smaller the ratio of ampure beads used, the larger the fragment you will obtain.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM
        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        18 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        22 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        17 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        49 views
        0 likes
        Last Post seqadmin  
        Working...
        X