  • 20% adapter get sequenced using MiSeq

    Hi all,

    I am new to MiSeq sequencing and it really frustrates me lately.

    I had two MiSeq run recently and they all turned out to have about 20% adapter get sequenced. I've been using the TruSeq PCR-free DNA prep kit for library preparation. So the first time I followed the protocol for library prep ant it gave me 20% adapter sequenced. The second time I did an 8 cycle PCR enrichment for my library, followed by gel extraction of DNA fragments ranging from 500bp to 1500bp, hoping to remove all adapter dimers. However it gave me about 20% adapter sequence as well, which also leads to an uneven distribution of my libraries (percent reads pass filter for all samples ranging from 0.2% to 24%).

    I wonder if anyone knows why this happens? We don't have bioanalyzer, so I only used qPCR for library quantification using KAPA kit. Also if anybody knows where I could find a commercial bioanalyzer? I would like to pay for them to run a couple of samples for me, in order to get a better MiSeq result.

    Thank you!!!

  • #2
    I wonder which of following you are referring to:

    1- whole read was adapter sequence for 20% of reads
    2- 3' end of reads had adapter sequence which was 20% of output

    The first case indicates incomplete removal of adapters after ligation while the second case would happen if read length were longer than insert size. If you have not use BA then how would you know the average size of libraries which is essential to calculate molar concentration from qPCR results.
    PF% is very low which could be result of over clustering among other causes. You need to run sample in BA or a similar device to obtain correct average library size for optimum loading and clustering.
    Last edited by nucacidhunter; 05-07-2016, 12:54 PM. Reason: typo was fixed

    Comment


    • #3
      It's number 1. The most recent MiSeq was run using 2x300bp, in about 20% of the sequencing output, I get about 60bp read length followed by low quality nucleotides, which I think indicate number 1 happening.

      For read length calculation, I ran the enriched library on DNA gel so I could get an estimate of average fragment length. I should correct myself, the cluster density (~1100K/mm2) and pass filter rate (~90%) for the MiSeq run was great. The problem I had was distribution of % reads. For some of my samples, they were sequenced for only 0.2% to 1% among all sequencing samples; for some other samples, they were sequenced 24%. Theoretically if I have 18 samples and I loaded them equally, I am expecting about 5.5% for each get sequenced.

      Comment


      • #4
        Qubit for sample concentration along with bioanalyzer for size estimation is essential to get balanced pools. You will always find some variation in the actual number of reads you get for each sample (due to pipetting/conc estimation errors).

        As you discovered yourself smaller inserts (adapter dimers qualify) cluster very efficiently compared to longer real inserts. So the only way to avoid that is to get rid of them by purification.

        Comment


        • #5
          To GenoMax,

          Unfortunately we don't have BA and can't afford to have it for now

          To remove the smaller inserts, I did gel extraction (to remove fragment length less than 500bp), but it didn't seem to work

          Comment


          • #6
            You could pay an additional charge and have your sequencing facility QC your samples on BA.

            I am not on experimental side of things but bead purification would be better to remove primer dimers.

            Comment


            • #7
              Yeah, so we don't have instrument for beads purification either... and we don't have a sequencing facility, so I did everything by myself in-house.

              I wish I could pay sequencing company/facility just for BA usage.

              Comment


              • #8
                I don't think you need an instrument for bead purification.

                If you are in the Dallas area I am sure there will be local facilities that offer BA analysis as a "fee for service" option.

                Comment


                • #9
                  Cool! I will look into that. Thank you!

                  Comment

                  Latest Articles

                  Collapse

                  • seqadmin
                    Current Approaches to Protein Sequencing
                    by seqadmin


                    Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
                    04-04-2024, 04:25 PM
                  • seqadmin
                    Strategies for Sequencing Challenging Samples
                    by seqadmin


                    Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
                    03-22-2024, 06:39 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by seqadmin, 04-11-2024, 12:08 PM
                  0 responses
                  18 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-10-2024, 10:19 PM
                  0 responses
                  22 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-10-2024, 09:21 AM
                  0 responses
                  17 views
                  0 likes
                  Last Post seqadmin  
                  Started by seqadmin, 04-04-2024, 09:00 AM
                  0 responses
                  49 views
                  0 likes
                  Last Post seqadmin  
                  Working...
                  X