Lightbulb Looks like there is a problem with imaging one of the flowcell surfaces...

It looks like there is a problem with imaging one of the flowcell surfaces...

Either focusing issue (Z axis lens mover in a need of some cleaning/re greasing/calibration) or lots of very small bubbles (if it is an upper surface) (but then they would affect the other surface with such width.

Also check tile thumbnail images, and contact illumina if this instrument is on service contract.

You can also try making movie from thumbnail jpg images using image magic and some scripting...