  • Larger size TruSeq nano prep

    Hello,
    I followed the protocol for the 550bp size TruSeq nano library preparation but my final libraries were much larger than the expected 670bp average size (see attached pic).

    Has anyone else come across larger than expected libraries?
    Do you have any advice as to what may have happened?

    Thanks in advance,
    Anna.
    Attached Files

  • #2
    I also saw a larger than expected product. I ran the Covaris to shear at 500 bp, and the final libraries had Bioanalyzer peaks at 780 bp. I don't have an explanation.

    Comment


    • #3
      Truseq pcr free libraries do run larger on a BA due to the tailed adapters but I have not seen this before with nano libraries. One explanation my be incomplete mixing of the size selection beads meaning that the incorrect amount of beads were added to the cleanup and a poor size selection. Might also be an idea to run the sheared samples on the BA before cleanup to check the size distribution from the shearing.

      Comment


      • #4
        Hello, I asked Illumina tech support and they said to try diluting the libraries as the chip looked to be overloaded. I made a 1 in 20 dilution and repeated the chip. An example of a trace is attached. The average size has gone down by around 100bp but it's still not the 670bp average stated in the manual. I have sent tech support these diluted library results and asked if they think they will sequence.

        Anna.
        Attached Files

        Comment


        • #5
          I have the same problem. I make a change on the fragmentation protocol whit covaris (from a 550 pb insert to a 400 pb insert) but the mean size was almost the same (900 bp aprox.). I think that could be a problem on the size selection. Any help?. Thanks.

          Comment


          • #6
            Hello,
            It's a size selection issue. The protocol doesn't need to be altered, just make sure your SP beads are vortexed a lot (I mean a lot!). I usually mix them in between each sample prep as a pre-caution. Also we've found that the libraries run larger on the tapestation in comparison to the bioanalyser. Libraries usually look around 50bp larger on the tapestation. When the insert sizes are checked in the analysis stage, they are usually around the right size.

            All the best,
            Anna.

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Current Approaches to Protein Sequencing
              by seqadmin


              Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
              04-04-2024, 04:25 PM
            • seqadmin
              Strategies for Sequencing Challenging Samples
              by seqadmin


              Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
              03-22-2024, 06:39 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, 04-11-2024, 12:08 PM
            0 responses
            18 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 10:19 PM
            0 responses
            22 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 09:21 AM
            0 responses
            16 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-04-2024, 09:00 AM
            0 responses
            47 views
            0 likes
            Last Post seqadmin  
            Working...
            X