mdegraaf,

You didn't say how many libraries you pooled. If it is more than 12 then it is inevitable that you have more than one library sharing a common i7 index sequence. i7 is the first of the two index sequenced or the ONLY index sequenced if you choose PE150+8+150. You can not use this format if any i7 index is repeated in your pool, you must then use PE150+8+8+150. Honestly I'd always choose the dual index sequencing format; what would be the downside to it?