I have run a lot of amplicon samples on the Bioanalyzer and never seen that! Since it's a high sensitivity chip, is it possible that the previous run contaminated the electrodes with something 370 bp in length? I'd pull out the electrode cartridge, clean it thoroughly, and try again.