  • DNA loss on Bioanalyzer

    About two weeks ago I sheared some gDNA samples on the Covaris E210 and purified them with AMPure beads, as instructed by the SureSelect XT protocol. I measured the samples on the Bioanalyzer 2100 immediately following clean-up and received good peaks at the right area. However, I re-measured the samples as a positive control and I see that even though the peaks are still present for my sample, they are much shorter and it appears I have lost some gDNA.

    My first instinct is to consider this nuclease contamination, but if this were the case, wouldn't I not see any evidence of DNA on the Bioanalyzer? I'm also wondering if I lost DNA following the freeze-thaw cycle the samples underwent, for they were stored at -20 degrees Celsius. Does anyone else have any ideas about what could be causing this DNA loss, as well as any suggestions about how I should prevent it? Thanks!

  • #2
    What's the original sample source (organism and if applicable, tissue) and purification method? How much shorter is shorter? Did the whole peak move

    Nuclease contamination is a possibility. Freeze-thaw problems are very unlikely in my experience, especially if you're observing a dramatic change in sample quality from a single cycle.

    I always suspect the Bioanalyzer. =] Any quantitation method that relies on pipetting of 1ul is by definition a ballpark method (unless you run several dilutions/reps etc).

    Comment


    • #3
      The original source was tissue, and the DNA was purified with Agencourt AMPure XP magnetic beads. The peak is shorter by about 15 fluorescence units, but the peak itself did not move.

      I talked to tech support about this and they told me if there is only one visible peak and the peak size hasn't moved to a smaller size range, then I shouldn't worry about nuclease contamination. Tech support then validated your opinion, ECO, and said it was likely discrepancy between the two reads. I guess storing nucleic material in water for a long time rather than a buffer makes me a bit queasy.

      Thanks for the feedback!

      Comment


      • #4
        Do you measure concentration of your samples with another technics? Bioanalyzer not reliable for measuring of concentration.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Essential Discoveries and Tools in Epitranscriptomics
          by seqadmin




          The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
          Yesterday, 07:01 AM
        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        58 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        54 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        45 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        55 views
        0 likes
        Last Post seqadmin  
        Working...
        X