I haven't done it... but did ask the question. It can be done with no problem.
The v3 cluster kit advancement is more reads (bigger flow cell/more clustering real estate), and improved GC resolution (chemistry changes). The v3 sequencing kit advancement is simply better stability, you can load 209 cycles of ICR, rather than split for a 2x101. TruSeq v1 seq kits can be used with v3 cluster kits with no reduction in q scores (presumably), you just have to split ICR for a 2x101