    Fragment length limits?

    I've been told that with the Ion Fragment Kit, the maximum input fragment size should be 150 bp so that with adapters the fragment is no longer than 210 bp.

    Can anyone with access to Ion's deepest, darkest secret documents (known elsewhere as "package inserts") confirm this? The application note on Amplicon Sequencing mentions nothing on the subject.

    Anyone know if the constraint is imposed in the library prep or the emPCR template prep? What imposes it? Extension times? Nucleotide concentrations? emPCR micelle sizes?

    One might guess this is something Ion will need to address in order to hit 454-like read lengths
  • Marzanna
    Junior Member
    • Mar 2008
    • 7

    #2
    We are still waiting for installation so I can only share theoretical knowledge from the secret documents ;-)

    The fragmentation is done with BioRuptor or Covaris (both described).
    "If the size distribution of the fragmented DNA is in the 150–200 bp range, you can use alternative fragmentation methods such as the Covaris System."

    After end repair and adapter ligation there is size selection step (eg. Pippin Prep).
    "For optimal sequencing results, size-select a DNA library with a mean size of 185–210 bp, with a size distribution of ±20 bp around the mean. Libraries with a mean size >~220 bp yield results of reduced sequencing quality."

    I was also told that too wide insert size range (like size selection on beads for SOLiD) leads to too high percentage of short reads (50-100bp).

    And now I will be taken directly to the IonHell
    see you around
    Marzanna

    Comment

    • The_Dark_Base
      Junior Member
      • Aug 2010
      • 6

      #3
      The small insert size really voids the use of the PGM for their often touted library validation application. Looking at the tech note, you have a library with adapters, then add an oligo on to attach to their particle. With all that only adapter dimers would work on the PGM?

      There must be a size limitation due to how much DNA can be put onto their ionspheres, i.e. longer templates have less on the ionsphere, produces less H+

      Just look at binding on other Life Technology spheres:

      "The amount of biotinylated DNA immobilized will depend on fragment size. Due to steric hindrance, the binding efficiency is significantly reduced when the fragment size exceeds..."

      Comment

      • SeqAA

        #4
        Originally posted by The_Dark_Base View Post
        There must be a size limitation due to how much DNA can be put onto their ionspheres, i.e. longer templates have less on the ionsphere, produces less H+
        this would be my guess as well. High copies per sphere until their algorithms get better?

        Comment

        • ECO
          --Site Admin--
          • Oct 2007
          • 1360

          #5
          Originally posted by SeqAA View Post
          this would be my guess as well. High copies per sphere until their algorithms get better?
          Also sensor/chemistry improvement perhaps? Algo can't detect protons that aren't there from a low level of template.

          Comment

          • zhengz
            Member
            • Aug 2010
            • 24

            #6
            Originally posted by SeqAA View Post
            this would be my guess as well. High copies per sphere until their algorithms get better?
            As Ion prioritizes quick turnaround time, amplicon length has been sacrificed. The advertisement says read length is going to be double in 201x. It has been quite long time since 454 advertised 1000 bp, but still not there yet. Read length is the hard part for emPCR/bridge-PCR based technologies.

            Comment

            • Claudia Stewart
              Junior Member
              • Sep 2008
              • 4

              #7
              Originally posted by krobison View Post
              I've been told that with the Ion Fragment Kit, the maximum input fragment size should be 150 bp so that with adapters the fragment is no longer than 210 bp.

              Can anyone with access to Ion's deepest, darkest secret documents (known elsewhere as "package inserts") confirm this? The application note on Amplicon Sequencing mentions nothing on the subject.

              Anyone know if the constraint is imposed in the library prep or the emPCR template prep? What imposes it? Extension times? Nucleotide concentrations? emPCR micelle sizes?

              One might guess this is something Ion will need to address in order to hit 454-like read lengths
              Apparently, the issue is the micelle size – material larger than 250bp doesn’t drive back to bead well. I suspect that this will no longer be the case with the new One Touch, since there is no oil in the system. At this point, that seems to be the only way to get to longer fragment lengths.

              Comment

              • BBoy
                Member
                • Oct 2010
                • 50

                #8
                Originally posted by Claudia Stewart View Post
                Apparently, the issue is the micelle size – material larger than 250bp doesn’t drive back to bead well. I suspect that this will no longer be the case with the new One Touch, since there is no oil in the system. At this point, that seems to be the only way to get to longer fragment lengths.
                Does de-phasing impose any limits on their read lengths?

                Comment

                • krobison
                  Senior Member
                  • Nov 2007
                  • 734

                  #9
                  Originally posted by BBoy View Post
                  Does de-phasing impose any limits on their read lengths?
                  Yes. De-phasing is the ultimate bugbear for any clonal sequencing approach.

                  Note that since my original query Ion has released a much longer dataset, though that protocol won't be public until sometime this fall.

                  Comment

                  • BBoy
                    Member
                    • Oct 2010
                    • 50

                    #10
                    But that does not seem to be the limiter at this time, since they are driving read length up without changes to the chemistry or chip. Or am I missing something? It seems that longer-term read length can be helped by more sensitive readout what would enable smaller beads with fewer strands on them. Assuming that the probability for dephasing is sqrt(N) where N is the number of strands on a bead, things should get better linearly with bead size, no?

                    Comment

                    • krobison
                      Senior Member
                      • Nov 2007
                      • 734

                      #11
                      I'd have to really think through the dephasing, and probably model it.

                      Probability that any given molecule will dephase in a given cycle is the 1-repetitive yield (I'll call this Pdephase). If you are going to model this, one option is to assume that dephased reads are lost forever. Alternatively, one can assume they can be rephased under the right circumstances (indeed, Ion claims their flow order encourages recovering dephased molecules).

                      I'm not convinced that fewer molecules leads to less dephasing until you get to very small numbers. The current beads carry 800K molecules per bead; even with very high repetitive yield you are going to lose some in each round and presumably that is close to a fixed proportion. Clearly, if you get to molecule numbers approach or go below 1/Pdephase then you need to model it as a stochastic process. I'm not sure what Pdephase is, but presumably you could crudely estimate it from the distribution of read lengths.

                      If you get to really small numbers (but not 1), then a possible issue is that the relative contribution of a dephased molecule in the population becomes large -- so it is rare that a bead has any dephased molecules but a small number of dephased molecules kills your signal-to-noise ratio.

                      Comment

                      • lek2k
                        Member
                        • Aug 2011
                        • 32

                        #12
                        I'm going to go off memory on this one so please correct me if I'm wrong.

                        1. dephased reads may be lost for ever but the contributions they make exist. This is evident in a ionogram as the zeromer calls look a little noisy and still exist deep into the read after rephasing.

                        2. According to the guys in Ion Community, the flow order can lead to dramatic improvements. However, this was only a simulation. I have found it very hard to find in the code where they have taken advantage of the 32 flow redundant cycle. Anyways, I suspect the long read data set is a combination of chemistry, flow order and other flow parameters being tweaked. There was also a change in version of ion-Analysis that hasn't been released. If you look at the summary report PDF the signal incorporation plot looks very, very odd.

                        3. Fewer molecules means weaker signal, which means good luck trying to differentiate between a zero-mer and one-mer when things start dephasing. Also I agree, if you have a fewer number then if a few dephase that would be a large proportion of the total making it difficult to rephase and base call. Then again, there could be a biochemical law that says there is a relationship between strands and polymerase and reagents?

                        4. For each chip sequencing run, the carry forward, incomplete extension and droop fitted probabilities are listed in one of the files for each CAFIE region (13x13 wells).

                        Comment

                        • krobison
                          Senior Member
                          • Nov 2007
                          • 734

                          #13
                          Sorry, by lost I meant they no longer contribute signal but do contribute noise.

                          The flow order improvement wouldn't show up in the code. The idea is that if you follow base X with another X after a few other bases, you can catch up some reads. For example, imagine the flow order

                          1 2 3
                          A T A

                          Incompletely extended molecules after cycle 1 will be finished off in cycle 3 and resynchronized in round 3, unless of course the next correct base is T in which case the majority of molecules on the bead will have marched ahead.

                          Comment

                          • simon.jarman@aad.gov.au
                            Junior Member
                            • May 2009
                            • 5

                            #14
                            Has anyone tried the 200 bp kit yet? Using the 100 bp kit on variable size amplicons we see a strong bias against the longer (~200) bp amplicons so that amplicons around 150 bp are over-represented by proportion. This will cause us many problems in analysing environmentally-derived amplicons. Does anyone know how life tech plan to get around this for longer read lengths?

                            Comment

                            • arolfe
                              Member
                              • Jul 2011
                              • 29

                              #15
                              We've used the 200bp kit. It seems to work as advertised and increases the maximum construct size (adapters + insert) from about 230bp to about 330bp. I don't have any data about whether the dropoff in sequencing efficiency is as steep at the top of the size range as it was for the 100bp protocol.

                              The only problem is that there's no OneTouch 200bp kit yet, so you have to do the ePCR by hand for the moment.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                Yesterday, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              31 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              21 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              71 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...