    Poly-G Failed Runs on NextSeq

    Approximately a month ago, our lab performed an RNASeq using a TruSeq Stranded mRNA library prep and a v2 75cyc High Output flowcell. The data that came back was almost entirely G's and was severely underclustered. When we went to re-run the same library, the machine died and would soon thereafter require all of the boards to be replaced.

    We assumed that the poly-G run was an early symptom of the board failure, until we tried to run a genomic library prepared with Nextera DNA reagents on v2 300cyc High Output flowcell and again got nearly all G's (also very underclustered). There were no shared reagents between the two library preps and our lab has performed many of each type of prep without failure.

    Thinking that it was a hardware issues still, we did a phiX run on a 300cyc Mid Output flowcell... which of course came out perfect. Illumina is of the opinion that it is a transient issue and that we should just continue trying, but there's got to be more to it than that. Have any of you run into this problem before or have thoughts on what may be going on?
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    Have you tried to sequence the libraries on a different sequencer?

    Comment

    • djs150
      Junior Member
      • Oct 2015
      • 6

      #3
      We have not tried them on a different machine yet. Illumina was first insistent that it was a library prep issue, so at the time it seemed wasteful to run on a separate machine.

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        You may want to spike one of the libraries with someone else (as long as the tags allow) who may be running on a MiSeq/HiSeq. That would confirm that quality of your libraries before you try a re-run on NextSeq.

        Comment

        • wdecoster
          Member
          • Oct 2015
          • 97

          #5
          Given that NextSeq uses two color chemistry, poly G essentially means an absence of signal...

          Comment

          • Tim NGS
            Junior Member
            • Mar 2015
            • 4

            #6
            Make sure no custom primers were selected during the run setup. Information can be found at bottom of the "runparameters.xml"-file. Just look for "<UsesCustomReadOnePrimer>true"

            Sounds stupid, but we also had it once when the custom primer button was pressed accidentally when initiating an experiment. The system will use the wrong primer for first read sequencing, so no signal is generated and thus no clusters will be detected. Due to the chemistry, there is no signal from G-bases. Run will look severely underclustered and possibly only G's will be called.

            Comment

            • djs150
              Junior Member
              • Oct 2015
              • 6

              #7
              At this point, I welcome things that sound stupid if it means we can get this wrapped up. Unfortunately the run parameters show false for the custom primers and everything else in there looks as expected.

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #8
                Only options I see is to cross your fingers and try a re-run with NextSeq or check the libraries on a MiSeq/HiSeq to confirm that they are sequenceable before coming back to NextSeq.

                Comment

                • djs150
                  Junior Member
                  • Oct 2015
                  • 6

                  #9
                  As an update, we've now run the same library on a different NextSeq and obtained the same results. The library was run with a 1% PhiX control, which sequenced perfectly.

                  We thought that, due to the high GC nature of the libraries, the strands weren't denaturing. The NaOH stock used was over pH 13 and a Qubit before and after denaturing showed that they are certainly single stranded when they are being loaded.

                  So the problem is in the library prep, somewhere. The only common reagent is the Omega Bio-tek beads from the PCR clean up. Is there any rationale that the beads could be causing this problem somehow?

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    This is likely an indication of a bad library prep and you may be better off starting over.

                    As a last try do a MiSeq run, if you can. MiSeq is a resilient sequencer and may be able to sequence this sample. Having 4 color chemistry helps as well. If the library has low nucleotide diversity then increase the phiX concentration.
                    Last edited by GenoMax; 11-29-2016, 05:54 PM.

                    Comment

                    • Brian Bushnell
                      Super Moderator
                      • Jan 2014
                      • 2709

                      #11
                      I think if the same library failed with the same error mode on a different machine, it would be a waste of resources to run it again. Considering the PhiX reads sequenced perfectly, I cannot imagine that a different platform would give different results. It sounds like you just don't have any sequence (since as wdecoster noted, 'G' means no signal). But to clarify, are the sequences 100% G, or are they ~60bp of adapter and then poly-G?

                      Comment

                      • GenoMax
                        Senior Member
                        • Feb 2008
                        • 7142

                        #12
                        I think two separate libraries were originally tried but at least one of the libraries was run again and failed.

                        If this is an irreplaceable sample then try the last ditch effort I mentioned above otherwise I agree that it would be best to start over and make new libs.

                        Comment

                        • Brian Bushnell
                          Super Moderator
                          • Jan 2014
                          • 2709

                          #13
                          Originally posted by GenoMax View Post
                          If this is an irreplaceable sample then try the last ditch effort I mentioned above
                          Good point, I did not consider that aspect. If that's the situation I concur with Genomax.

                          Comment

                          • djs150
                            Junior Member
                            • Oct 2015
                            • 6

                            #14
                            They are 100% G, adapter and all. And yes, there were originally two unrelated libraries that failed and the second was re-run on a different machine.

                            The samples are by no means irreplaceable as we have the original DNA to remake the libraries. However, if I'm missing something obvious that is going to lead to another failed run (i.e. bad reagent), then remaking the libraries isn't going to help.

                            I'm simply at a loss to explain why Qubit and Tapestation indicate there is tons of DNA there and yet I can't seem to get a viable read on in it.

                            Comment

                            • Brian Bushnell
                              Super Moderator
                              • Jan 2014
                              • 2709

                              #15
                              I wonder if...

                              A) The adapters never attached to the reads
                              or
                              B) The reads were to long to bridge-amplify (in which case a different platform might make a difference)

                              I don't know how to diagnose those issues, but they seem like the most likely culprits to me.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                                by SEQadmin2



                                CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                                Despite this, “CRISPR helped turn genome editing from a specialized technique into
                                ...
                                07-31-2026, 11:01 AM
                              • SEQadmin2
                                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                                by SEQadmin2


                                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                                The systematic characterization of the human proteome has
                                ...
                                07-20-2026, 11:48 AM
                              • SEQadmin2
                                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                by SEQadmin2



                                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                ...
                                07-09-2026, 11:10 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Yesterday, 10:13 AM
                              0 responses
                              13 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-31-2026, 02:55 AM
                              0 responses
                              26 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-24-2026, 12:17 PM
                              0 responses
                              20 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-23-2026, 11:41 AM
                              0 responses
                              19 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...