Unconfigured Ad

Collapse
X
 
  • Filter
  • Time
  • Show
Clear All
new posts
  • soleulloa
    Member
    • Aug 2011
    • 34

    Discordance between Cluster Density and Thumbnail images

    Hi all,

    I ran MiSeq using a 500 cycles kit.
    The value for cluster density was 310K/mm2 but if I'm going to look the thumbnail images, I think I have much more cluster than that.

    Illumina tech looked the results and says "you are overclusterized".
    But for me cluster density is OK when I've reviewed the images.

    I'm attaching here 4 images from cycle 1.

    Could you tell me your opinion?
    Thank you!!
    Soledad
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    To me reported cluster density is what I would expect from the images. The run has been underclustered for a library that looks high diversity although I am assuming that the base diversity of other cycles are similar to the first cycle.

    Comment

    • Brian Bushnell
      Super Moderator
      • Jan 2014
      • 2709

      #3
      What's the scale on the little yellow boxes?

      Comment

      • misterc
        Member
        • Jan 2016
        • 46

        #4
        That looks about right to me for 300-500K/mm2 raw density (and NOT overclustered). Those intensities look reasonable to me too for cycle 1 given the MiSeq ramps up exposure on the cameras at each cycle to counter the typical decreasing intensity plots on their instruments.

        Comment

        • soleulloa
          Member
          • Aug 2011
          • 34

          #5
          Ok! Thank you.
          I'm using 10 pM (from 4 nM) to get this result.
          What do you think I have to use now? 12 or 15 pM to get a good run?

          Comment

          • microgirl123
            Senior Member
            • Jun 2012
            • 199

            #6
            What's your percentage passing filter?

            Comment

            • soleulloa
              Member
              • Aug 2011
              • 34

              #7
              Hi,

              95.27% PF

              Comment

              • microgirl123
                Senior Member
                • Jun 2012
                • 199

                #8
                If 95% of them are passing filter, then you are definitely not overclustered. If you were overclustered your percentage passing filter would be low.

                Unfortunately, the relationship between loading concentration and clusters is not linear so it's always a guess. I find that getting good cluster density is the Achilles' heel of the Illumina Miseq. My guess would be to go with the 15 pM since your density was so low.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by Brian Bushnell View Post
                  What's the scale on the little yellow boxes?
                  I don't know, but when I looked into it in the past, it seemed that the size of a cluster was around 1um in diameter. But that may have been on a HiSeq. I'm not sure whether the scales would be different on a MiSeq.

                  Old-timers like @ECO would be more likely to know. (Alas, I don't think @ callouts actually work on SeqAnswers...)

                  --
                  Phillip

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    Illumina has a note on optimizing cluster density available here.

                    @soledad: Have you checked what the images look like in the middle and at the end of run?
                    Last edited by GenoMax; 02-17-2017, 08:51 AM.

                    Comment

                    • soleulloa
                      Member
                      • Aug 2011
                      • 34

                      #11
                      Someone has quantified libraries before dilution? When you prepare 2 nM dilutions?
                      I've quantified before and after that final dilution now because this is the first run when I have this problem and I'd like to compare these results.
                      Always I was between 1000-1200 K/mm2 using a 300 cycles kit.
                      This time I did the same process... and the only variable is I'm using 500 cycles kit.
                      But the library concentration will be the same.. so I don't understand what happens.

                      Comment

                      • nucacidhunter
                        Jafar Jabbari
                        • Jan 2013
                        • 1250

                        #12
                        It is most likely overestimating library concentration if you have used the same type of library. qPCR is the best method for quantification, other methods that quantify dsDNA are inadequate for libraries that contain non-sequenceable dsDNA fragments.

                        Comment

                        Latest Articles

                        Collapse

                        • mylaser
                          Reply to Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                          by mylaser
                          Kheloyaar: The Complete Guide to Kheloyaar Loginand Kheloyaar ID
                          The online gaming industry has transformed the way people enjoy digital entertainment. As technology continues to improve, players are looking for platforms that offer convenience, security, and a seamless user experience. Kheloyaarhas gained attention among users who value an easy-to-use platform, quick account access, and a simple registration process.
                          Whether you're exploring Kheloyaar for the first time or want to understand...
                          Yesterday, 09:27 PM
                        • SEQadmin2
                          Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                          by SEQadmin2



                          Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                          ...
                          Yesterday, 11:10 AM
                        • SEQadmin2
                          Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                          by SEQadmin2



                          Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                          There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                          07-08-2026, 05:17 AM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by SEQadmin2, Yesterday, 10:04 AM
                        0 responses
                        8 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 07-08-2026, 10:08 AM
                        0 responses
                        7 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 07-07-2026, 11:05 AM
                        0 responses
                        11 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 07-02-2026, 11:08 AM
                        0 responses
                        31 views
                        0 reactions
                        Last Post SEQadmin2  
                        Working...