    Shoulder in Truseq ChIP sample preparation libraries

    Hey,

    I am a newbie who has been trying to build ChIP libraries using the TruSeq ChIP Sample preparation kit.
    After doing protocol optimization we decided to use MinElute columns for pre-PCR purification instead of using Ampure beads XP. After the PCR, we did a Ampure XP clean-up to remove adapter dimers with the recommended ratio of 1:1.

    For the Gel size selection, we are using the E-Gel Size-Select Gels to select fragments between 250-300bp of size.

    However, now we are having another type of problems, which is the reason of this post. The libraries that we prepared using this method show a shoulder before the peak and the peak is of a much higher weight than the expected one.

    Does anyone know the reason for these shoulders?
    Do they affect sequencing quality?
    Is it possible to eliminate them?

    I used the search function and couldn't find a satisfactory answer.
    Attached Files
    Last edited by Rosmano; 01-14-2013, 05:43 AM.
  • kwaraska
    Senior Member
    • Nov 2008
    • 131

    #2
    The first issue is that the y-adapters of TruSeq do not run true to size on an E-gel. We were off by anywhere from 100 to 300 bases.

    We now use a robot for the size selection but I would suggest either find a Pippin Prep instrument or look for double sided SPRI bead for size selection. (It uses the AMPure beads at differing concentrations to remove High MW then low MW)

    Comment

    • Rosmano
      Member
      • May 2012
      • 28

      #3
      I wouldn't mind to try that, but if possible in the short term I would like to understand what issues may this shoulder on samples 2 and 4 cause.
      Why it may be happening and ways to avoid while trying again an E-gel based purification.

      I don't have access to a Pippin prep. I could try the double sided SPRI method but is it as tight as gel size selection?
      Last edited by Rosmano; 01-11-2013, 05:07 PM.

      Comment

      • Rosmano
        Member
        • May 2012
        • 28

        #4
        So, I suppose that these Bioanalyzer curves don't have any issue?
        Could this shoulder be derived from Y-shaped adapters?

        I would like ot know if that shoulder may interfere with sequencing, since last time I sequenced libraries that had that shoulder (and small adapter dimer peaks) I had a really poor quality sequencing list.
        Last edited by Rosmano; 01-14-2013, 06:34 AM.

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          The high MW peak are just bubble products, no? That is, the products that anneal to each other, instead of being replicated during the last cycle of PCR. Since they are a normal duplex only in the adapter region they migrate more slowly than the full double stranded library molecules.

          If you google

          "bubble product" site:seqanswers.com

          you will see probably a hundred posts on this topic. The good news is that bubble products sequence fine. You just have to get the titre right, which can be tricky unless you use qPCR.

          --
          Phillip

          Comment

          • SLeigh-Brown
            Junior Member
            • Dec 2011
            • 3

            #6
            Looking at your bioanalyzer trace, Rosmano, I believe that these are unlikey to be bubble products. We had a lot of trouble with bubble products a few months ago using the Truseq RNA library prep kit, and we were consistently told that bubble products will run at twice the size of your library peak - so you get two separate peaks on your gel, one at ~300bp as you expect, the other at ~600bp.

            What you have shown here is a 'shoulder' which is slightly smaller than your expected library size. I often see this on libraries cut by hand from a slab-gel, but it is less common in library size selected with an e-gel.

            As long as that shoulder is DNA from your library, your library should sequence without any problem. A mix of small and large DNA products should not influence sequencing quality, as long as quantification is accurate.

            One or two words of caution before you proceed:
            - Are you pooling libraries with different indexes? If you are, it will be important to check that all of your libraries have the same profile - if not, your relative ratios of each library in the final pool are unlikely to be accurate.
            - If you have access to a Nanodrop, I recommend you run a ul of each library and check your 260/230 ratio is low (this will test if you have phenol or other aromatics contamination - http://www.nanodrop.com/Library/T042...ity-Ratios.pdf)
            - If you have a library with a complex profile as opposed to a single peak, I would not recommend using the agilent bioanalyzer for quantification. You will have best results using qPCR.
            - I suggest you mention the 'shoulder' to your sequencing provider, and send them the trace so that they can take any steps they feel necessary.

            Good luck, and well done on your first try - I have seen many worse libraries than this from someone's first ChIPseq prep!

            Comment

            • Rosmano
              Member
              • May 2012
              • 28

              #7
              Hello!

              Thanks everyone to the replies. Well, we continued to troubleshoot the protocol and managed to get rid of that shoulder by doing two things:

              1 - Do an Ampure XP beads purification with a 1:1 ratio instead of a MinElute purification before running the e-gel. The Bioanalyzer profile gets slightly smaller shoulders.

              2 - Pick DNA of a smaller size. The e-gel size select isn't exactly accurate. When I pick the fragments between 250bp-300bp I usually get libraries with the size distribution that my profiles show.
              Afterwards we decided to pick smaller fragments (take DNA when the 200bp marker is in the bottom well) and managed to consistently obtain DNA with a size of 270-285 bps.

              The shoulder disappeared in these samples so I am inclined to believe that it was smaller sized fragments that remain in the bottom well. All the libraries that I prepared have a single peak (the last one has a very small adapter dimer peak).

              The concentration range of the libraries that I have prepared untill now, as measured on Qubit, are 13,6ng/uL to 25ng/uL with samples 2 4 and 7 amplified with 17 cycles and the last one 18 cycles (that sample had other problems).
              To me they seem to have similar profiles.
              Attached Files
              Last edited by Rosmano; 02-05-2013, 09:15 AM.

              Comment

              • Thimble_12
                Junior Member
                • Jun 2012
                • 4

                #8
                Hi SLeigh-Brown,

                I just noticed you mention having trouble with double peaks in your RNA-Seq libraries - did you ever get to the bottom of what was causing them? We've had similar issues and suspected it was a kit issue, but it's frustrating not really knowing how to cure it!

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by Thimble_12 View Post
                  Hi SLeigh-Brown,

                  I just noticed you mention having trouble with double peaks in your RNA-Seq libraries - did you ever get to the bottom of what was causing them? We've had similar issues and suspected it was a kit issue, but it's frustrating not really knowing how to cure it!
                  Which type of double peaks? Bubble products? Just do fewer cycles of PCR.

                  The other issue -- not completely removing the smaller fragments is more complex and depends on your fractionation method.

                  --
                  Phillip

                  Comment

                  • Thimble_12
                    Junior Member
                    • Jun 2012
                    • 4

                    #10
                    We have been getting double peaks, so we suspect it's bubble products.

                    Hmm, we've tried reducing the cycle number to 10 and are still having trouble. We were concerned that it was a kit issue, but Illumina claim not to have had other users report problems, so I was interested to know how SLeigh-Brown dealt with the issue, because they seem to have had the problem over a similar time-frame.

                    Comment

                    • pmiguel
                      Senior Member
                      • Aug 2008
                      • 2328

                      #11
                      Originally posted by Thimble_12 View Post
                      We have been getting double peaks, so we suspect it's bubble products.

                      Hmm, we've tried reducing the cycle number to 10 and are still having trouble. We were concerned that it was a kit issue, but Illumina claim not to have had other users report problems, so I was interested to know how SLeigh-Brown dealt with the issue, because they seem to have had the problem over a similar time-frame.
                      We routinely use 8 cycles with the TruSeq RNA kit.

                      Note that the bubble products normally are not an issue if you titrate your libraries using qPCR prior to clustering and use the left (lower molecular weight) peak to calculate your average amplicon size.

                      Remember, the presence of bubble products is just an indication that the ratio of product to primers in the PCR reaction has gone above a threshold at which substantial numbers of products anneal to one another before a primer can anneal. In a sense it is an indication that the prior TruSeq reactions worked well.

                      I think Illumina over-engineered the kit to work over a wide range of input RNA/DNA amounts.

                      If you really want to avoid bubble products altogether you can do qPCR prior to amplification and then pick a number of cycles that will keep your final concentration in the 2-4 nM range.

                      --
                      Phillip

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                        by SEQadmin2


                        Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                        The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                        ...
                        06-02-2026, 10:05 AM
                      • SEQadmin2
                        Single-Cell Sequencing at an Inflection Point: Early Impacts of New Platforms and Emerging Trends
                        by SEQadmin2


                        With the launch of new single-cell sequencing platforms in 2026, the field stands at an exciting inflection point. This article surveys the most impactful advances in the field and discusses how they’re reshaping research in cancer, immunology, and beyond.


                        Introduction

                        Single-cell sequencing technologies have undergone remarkable advances over the past decade, transitioning from low-throughput experimental approaches to highly scalable platforms capable of...
                        05-22-2026, 06:42 AM
                      • SEQadmin2
                        Environmental Genomics in the Age of NGS: From Microbes to Conservation Strategies
                        by SEQadmin2

                        Studying ecosystems means dealing with complex, multi-species communities that are hard to observe at scale. This complexity, however, hides many important questions to be answered, from how biogeochemical cycles work and how climate change can affect species distribution to how conservation strategies can work best.


                        Genomics, particularly since the expansion of NGS, has transformed ecosystem ecology. By sequencing environmental DNA, we can now assess biodiversity without direct...
                        05-06-2026, 09:04 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 06-02-2026, 12:03 PM
                      0 responses
                      19 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 06-02-2026, 11:40 AM