    MiSeq v3 2x300 run, Low PF and Q30

    I recently ran a ~200-400bp library prepped with NexteraXT on the MiSeq using reagents kit v3 for a 2x300 run, paired end, dual index reads: 10pM pooled library with 15% 10pM phiX.

    I got a high cluster density of 1185 k/mm2 with only 27% clusters PF and >=Q30 14%.

    Does anyone have any ideas for troubleshooting? Could this come from old kits or bubbles on the flowcell, would this even be visible? I did a qPCR (KAPA library quantification) before dilution all samples to 4nM. The qPCR worked fine, so there must be an error afterwards or with the kit.
    Could it be that there was a DNase destroying the amplicons?

    I'm curious about your ideas and answers.

    Thanks in advance
    Sarah
    Attached Files
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    Have you checked the actual insert size of your library? It is possible that you have inserts significantly shorter than you expect them to be. That can explain the precipitous drop in Q score on R2.

    You can check post #2 in this thread to see how you can calculate that information.

    This run can also be borderline overloaded. That cluster density number is not very accurate once you get next the upper spec limit. Have you had Illumina tech support look at this run?

    Comment

    • SarahAurora
      Member
      • Jun 2017
      • 17

      #3
      Thank you for your reply. We have done the sequencing of this amplicon multiplex for many times and know the actual sizes. But are the moment the demultiplexing is problematic because it's too less reads.
      The run could be overloaded but we have also done the sequencing with equal samples and even more input. It was very overloaded but we did get a lot of clusters passing filter (but a bad Q30). Now we have a very low clusters passing filter. So I think that overloading could not be the main issue.
      No, we have not (yet) been in touch with tech support. I'm still hoping that the clues here could lead to an explanation.

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        If you have done similar sequencing successfully in the past then I suggest that you contact Illumina tech support and have them look at the run. This may turn out to be a hardware related problem on the sequencer.

        Comment

        • thermophile
          Senior Member
          • Apr 2015
          • 243

          #5
          have you looked at the thumbnails? I'd guess your cluster density is higher than reported.
          Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

          Comment

          • SarahAurora
            Member
            • Jun 2017
            • 17

            #6
            Thumbnails of the tiles look ok, or let's say, they don't look as bad as the very overclustered run before. But they don' look balanced.
            Attached Files

            Comment

            • GW_OK
              Senior Member
              • Sep 2009
              • 411

              #7
              Can you post the %base graph?

              Comment

              • SarahAurora
                Member
                • Jun 2017
                • 17

                #8
                This one? I am not sure which one you want
                Attached Files

                Comment

                • SarahAurora
                  Member
                  • Jun 2017
                  • 17

                  #9
                  It's strange that we can read the first bases of R1, but no R4 and just a few Indices (R2+R3). Looks like the sequencing could not get read out like something intereferred with the optics.

                  I am very concerned about the run and my mind starts to melt, so please excuse this question: If I would have had trouble with the Index-PCR, then the adapters never would have gotten fixed on the flowcell, right? I want to exclude errors while preparing the run.

                  Comment

                  • GW_OK
                    Senior Member
                    • Sep 2009
                    • 411

                    #10
                    That's intensity. I was talking about the graph that is from % Base, which is selectable in the menu from the gray button under Intensity.

                    Still, though, in looking at this graph it appears you don't have a short insert but you do have sort've an odd 3-steps down to 0 intensity.

                    It looks to me like a hardware issue.

                    Comment

                    • SarahAurora
                      Member
                      • Jun 2017
                      • 17

                      #11
                      Ok, thank you for your opinion. The longer I check all circumstances, the more it gets clear that the problem was not the DNA. Because at first the bases looked fine.
                      Attached Files

                      Comment

                      • GW_OK
                        Senior Member
                        • Sep 2009
                        • 411

                        #12
                        What's troubling to me is that you don't have any intensity in read 2, which should always show if you have at least a read 1, regardless of index.

                        Have you run any of the onboard diagnostic tests?

                        Comment

                        • GenoMax
                          Senior Member
                          • Feb 2008
                          • 7142

                          #13
                          @Sarah: Do you not have a maintenance contract on this MiSeq with Illumina? If you do then you should go ahead and let tech support have a look at this run.

                          Comment

                          • SarahAurora
                            Member
                            • Jun 2017
                            • 17

                            #14
                            Originally posted by GW_OK View Post
                            What's troubling to me is that you don't have any intensity in read 2, which should always show if you have at least a read 1, regardless of index.

                            Have you run any of the onboard diagnostic tests?
                            Our bioinformatician told me we have 300 bp read 1 and read 2 but he main part is not Q30 or even Q20. So probably the DNA-bases part worked finde but the optics couldn't recognize anything...

                            No I did not know that those tests even exist...

                            Comment

                            • SarahAurora
                              Member
                              • Jun 2017
                              • 17

                              #15
                              Originally posted by GenoMax View Post
                              @Sarah: Do you not have a maintenance contract on this MiSeq with Illumina? If you do then you should go ahead and let tech support have a look at this run.
                              It's not our MiSeq, we just use it from our cooperation partner. I'll check if they have a maintenance contract with Illumina. Nevertheless, I wrote tech support an email with all details.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                                by SEQadmin2



                                CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                                Despite this, “CRISPR helped turn genome editing from a specialized technique into
                                ...
                                07-31-2026, 11:01 AM
                              • SEQadmin2
                                Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                                by SEQadmin2


                                Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                                The systematic characterization of the human proteome has
                                ...
                                07-20-2026, 11:48 AM
                              • SEQadmin2
                                Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                by SEQadmin2



                                Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                ...
                                07-09-2026, 11:10 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 07-31-2026, 02:55 AM
                              0 responses
                              17 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-24-2026, 12:17 PM
                              0 responses
                              15 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-23-2026, 11:41 AM
                              0 responses
                              13 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 07-20-2026, 11:10 AM
                              0 responses
                              25 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...