    fastqc shows strange GC content

    Hi,

    we have a 36-samples ChIP-Seq experiment with a very strange GC-content behavior (s. images)
    The samples are from yeast. This strange behavior comes in the IP samples, some of them shows (the first image is from such a file) in the fastqc report an over-representative accumulation of certain sequences, such as
    Code:
    GGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
    TGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG	
    TTGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
    AGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
    But some of them don't ( the second image is from a file with no(!) over-rep. sequences at all.
    Other than that the report looks quite good with no abnormalities (except the GC content).

    I was thinking that this curve might be due to the over-rep. polyG sequences, but what about the samples where there is no apparent accumulation of such sequences. Is there a possible biological reason for this thing?

    The problem occurs only in the IP samples. Is it possible that lack of biological material for the sequencer for this samples causes the machine to put a G (or identify a G ) intread of the real nt?
    we are using a nextseq500 with a high kit from Illumina

    thanks
    Assa
    Attached Files
  • frymor
    Senior Member
    • May 2010
    • 151

    #2
    I just wanted to add that this case is even stranger, as i have this sample, where the GC content is classified as good, even though it is completely shifted to the left.

    Q: How does fastqc calculate the theoretical GC content?
    Attached Files

    Comment

    • nucacidhunter
      Jafar Jabbari
      • Jan 2013
      • 1250

      #3
      The plots that you are referring should show normal distribution around the GC content of genome in a random library. But IP is not expected to be random so there would be a bias. Extreme GC at the start of reads could also be due to library construction method where some non-template bases are added.

      Stretches of G in NextSeq data indicates that there was not any signal in those cycles which could be due to short inserts and adapter dimers. F

      From FastQC manual:

      “This module measures the GC content across the whole length of each sequence in a file and compares it to a modelled normal distribution of GC content.”

      “An unusually shaped distribution could indicate a contaminated library or some other kinds of biased subset. A normal distribution which is shifted indicates some systematic bias which is independent of base position. If there is a systematic bias which creates a shifted normal distribution then this won't be flagged as an error by the module since it doesn't know what your genome's GC content should be.”

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        Originally posted by frymor View Post
        Hi,

        we have a 36-samples ChIP-Seq experiment with a very strange GC-content behavior (s. images)
        The samples are from yeast. This strange behavior comes in the IP samples, some of them shows (the first image is from such a file) in the fastqc report an over-representative accumulation of certain sequences, such as
        Code:
        GGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
        TGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG	
        TTGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
        AGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGGG
        But some of them don't ( the second image is from a file with no(!) over-rep. sequences at all.
        Other than that the report looks quite good with no abnormalities (except the GC content).

        I was thinking that this curve might be due to the over-rep. polyG sequences, but what about the samples where there is no apparent accumulation of such sequences. Is there a possible biological reason for this thing?

        The problem occurs only in the IP samples. Is it possible that lack of biological material for the sequencer for this samples causes the machine to put a G (or identify a G ) intread of the real nt?
        we are using a nextseq500 with a high kit from Illumina


        thanks
        Assa
        frymor,

        You correctly identified the cause in your last statement above. The Illumina NextSeq500 (and NovaSeq) use 2 Color Chemistry for tagging and identifying bases. Using this system G's are defined by the absence of a fluorescent signal. This means that anything which can cause a cluster to stop producing a signal with result in a polyG output. A much better and more thorough explanation can be found in this QC Fail post.

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          Stretches of G in NextSeq indicates lack of signal in those cycles and it could be result of short inserts and adapter-dimer.

          In a shot gun library GC content peak will be representative of genome GC but IP pull down is not random and some bias is expected depending on what sequences are pulled down and enriched.

          From FastQC manual:

          "This module measures the GC content across the whole length of each sequence in a file and compares it to a modelled normal distribution of GC content."

          "An unusually shaped distribution could indicate a contaminated library or some other kinds of biased subset. A normal distribution which is shifted indicates some systematic bias which is independent of base position. If there is a systematic bias which creates a shifted normal distribution then this won't be flagged as an error by the module since it doesn't know what your genome's GC content should be."

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
            by SEQadmin2


            Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

            The systematic characterization of the human proteome has
            ...
            07-20-2026, 11:48 AM
          • SEQadmin2
            Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
            by SEQadmin2



            Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
            ...
            07-09-2026, 11:10 AM
          • SEQadmin2
            Cancer Drug Resistance: The Lingering Barrier to Rising Survival
            by SEQadmin2



            Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

            There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
            07-08-2026, 05:17 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 07-24-2026, 12:17 PM
          0 responses
          30 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 07-23-2026, 11:41 AM
          0 responses
          23 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 07-20-2026, 11:10 AM
          0 responses
          213 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 07-13-2026, 10:26 AM
          0 responses
          79 views
          0 reactions
          Last Post SEQadmin2  
          Working...