    Hi all,
    first the TLDR:
    Do you know of any factors that might cause PhiX denaturation to fail or reverse, outside of typical mistakes (NaOH concentration/pipetting mistakes/age)?

    We have performed nearly one hundred MiSeq runs to date in our lab, and mostly they perform as expected. However, in some recent amplicon runs we have seen far less PhiX align than expected based on the % added, e.g. 1-2% aligned rather than 20% expected. This has happened both with an old and a fresh stock of PhiX. It cannot be explained by wrong library quantification, as the cluster densities were approximately as expected from the loading concentration. I quantified the new PhiX stock by qPCR, and it is very close to 10 nM. Also I verfied the insert size from what little did align. I checked the pH of the NaOH stock and dilution (made fresh every time) and it is OK. We have also done a system check of the instrument (optics alignment, volume test, thermal test) and everything passed.

    This leads me to believe something sometimes goes wrong with the denaturation of the PhiX. Once we had some old 12.5 pM aliquots of denatured PhiX (several months at -20), and we saw that the % aligned gradually declined, as expected for old aliquots. But when I made fresh aliquots from the stock it got much worse; almost nothing aligned. At the time I though the stock had gone bad. But then the same thing happened to a completely new stock. I am at a loss to understand why this is happening. I did contact Illumina about this problem and other amplicon run problems we had and got some help, but nothing specifically about the PhiX problem.
    Jon

    Hellenic Centre for Marine Research

Latest Articles

Collapse

  • SEQadmin2
    Nine Things a Sample Prep Scientist Thinks About Before Sequencing
    by SEQadmin2


    I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

    Here are nine questions we think about, in roughly the order they matter, before...
    06-18-2026, 07:11 AM
  • SEQadmin2
    From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
    by SEQadmin2


    Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


    The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
    ...
    06-02-2026, 10:05 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 06-17-2026, 06:09 AM
0 responses
36 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-09-2026, 11:58 AM
0 responses
100 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-05-2026, 10:09 AM
0 responses
120 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 06-04-2026, 08:59 AM
0 responses
113 views
0 reactions
Last Post SEQadmin2  
Working...