    Two lane failures and low cluster density - expired reagents?

    Hello everyone!

    I am currently in the middle of running a 72bp paired end, multiplexed run on a GAIIx. So far, the run is stopped after the first index read, so it's halfway through.

    After looking at the run report, I noticed that two of the lanes (3 & 6) didn't have any clusters at all and the remaining lanes, including Phix, had lower than normal data. There was also something weird going on with lane 8. The left side of the pictures had a verticle sharp line and bright clusters while the right side had more dim clusters. I thought it might be a line of oil, but I checked and everything seemed fine.

    When I ran the cluster station, all the lanes were flowing fine and nothing seemed out of the ordinary. I called Illumina and they looked at the data. Even they didn't seem to know what was wrong, saying that the intensities looked fine, and overall quality is not horrible.

    What bothers me though, is that they seemed to not take the situation as seriously when we found out the cluster kit reagents I was using were expired by 3 months. I was using kits bought by someone else, and the expiration date was removed, so I didn't know until Illumina checked the lot numbers.

    So my questions are:
    - Can these types of problems really be caused by expired reagents? And do these reagents really expire exactly when Illumina says they do?
    - Is there any other explanation that could be causing no clusters in some lanes, and lower data in the other lanes?

    I hope I provided enough information. I have a year experience running the GAIIx, but this is my first time troubleshooting, so I don't know all the terminology yet...

    Thanks for all your help!

    ~Angela
  • phoss
    Member
    • Aug 2011
    • 12

    #2
    Hi Angela,
    We had a similar problem to what you outlined (we also ran a multiplexed run on the IIx, several actually). Was the sample concentration in lanes 3 and 6 the same across all other lanes, or did 3 and 6 have differing concentrations?
    It is possible to have low/zero clusters or reads in one lane and plenty of reads in another. What we have learned through such troubleshooting is that over-clustering is a problem, hence a possible motive for low/no reads.
    Have you changed your multiplex kit since your last run or tried a different one?
    Last edited by phoss; 08-24-2011, 08:00 PM.

    Comment

    • Turnerac0987
      Member
      • Aug 2011
      • 15

      #3
      The concentrations in lanes 3 and 6 were the same as all the other lanes.

      At first we (and Illumnia) thought that this was an overclustering problem. However, on closer examination, we realized that the SAV images are autocontrasted so we were actually looking at the oligo lawn surface rather than an overclustered flowcell. So it was confirmed that lanes 3 and 6 had no sample present.

      I'm still not sure what caused lanes 3 and 6 to fail and all the rest of the data to be about a third of what it should be. I'm continuing with the run but I'm hoping to get some insight on what the problem could be so this doesn't happen in the future.

      Does anyone have any experience with using reagents that are expired? Illumina had an expiration date for those reagents at 6 months, but I figured that it wouldn't be an exact date. And now they are saying the reagents expire in a year, not 6 months, so I'm hesitent to believe this was the actual cause...

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
        by SEQadmin2



        CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

        Despite this, “CRISPR helped turn genome editing from a specialized technique into
        ...
        07-31-2026, 11:01 AM
      • SEQadmin2
        Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
        by SEQadmin2


        Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

        The systematic characterization of the human proteome has
        ...
        07-20-2026, 11:48 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 08-06-2026, 07:41 AM
      0 responses
      23 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 08-03-2026, 10:13 AM
      0 responses
      33 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-31-2026, 02:55 AM
      0 responses
      43 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 07-24-2026, 12:17 PM
      0 responses
      26 views
      0 reactions
      Last Post SEQadmin2  
      Working...