    RNA-seq libraries - double peak after size selection

    Hi everyone,
    I get some weird results when measuring my RNA-seq libraries on Bioanalyzer and now desperately need an advice...
    The libraries look like a peak of expected length (300-400 or 400-500) and a higher peak (>1000 bp according to BA) (the image is attached).
    I'm using TruSeq RNA v.2 kit, following the protocol (except for using qiagen columns instead of ampure beads and a size-selection step after PCR enrichment).
    This is the first time that I'm making RNA-seq libraries but I have some experience with standard DNA libraries - and none of them looked like this. It was always a peak of expected length, sometimes with traces of adapter dimers at 120 bp - but never any peaks higher than expected.
    What puzzles me the most is that I got this AFTER size selection. I'm pretty sure that I cut the band about 300-500 bp and certainly not 1500-2000 bp! So this should be something generated from my 300-500 bp fragments. What could it be? A single-stranded DNA? If so, will it affect the sequencing resuts? Does it make sense to try to run these libraries or it would be safer to redo them?
    Thank you very much in advance!
    Attached Files
    Last edited by MLog; 02-08-2012, 02:46 PM.
  • ssing
    Member
    • Jan 2009
    • 21

    #2
    Hi there,

    I got the exact same pattern you describe (couldn't see the image). See this post for what I found, what others suggested, and to see the end results. Let me know if you'd like more information.

    Techniques and protocol discussions on sample preparation, library generation, methods and ideas


    Best of luck!

    Comment

    • MLog
      Member
      • Jan 2010
      • 36

      #3
      Originally posted by ssing View Post
      Hi there,

      I got the exact same pattern you describe (couldn't see the image). See this post for what I found, what others suggested, and to see the end results. Let me know if you'd like more information.

      Techniques and protocol discussions on sample preparation, library generation, methods and ideas


      Best of luck!
      Thanks, ssing! This is a very useful link. In my case the reason is not the beads - just because I don't use them. So I will certainly try heating and cooling.
      PS: there was indeed a problem with the attached image but now it should work.

      Comment

      • pbluescript
        Senior Member
        • Nov 2009
        • 224

        #4
        It is due to too many cycles of PCR. When you have a high concentration of DNA fragments with complimentary ends, they can hybridize with each other, leading to the longer fragment sizes. Did you amplify each library with the same number of cycles or did you titrate the cycle number?
        If you titrate the number of PCR cycles, you can be sure you don't get these effects and introduce minimal PCR bias into your samples.
        You can see the gel I have attached to this post for an example of the appearance of these double sized bands as the number of PCR cycles goes from 6-18.

        Edit:
        The gel is 1.5% agarose post-stained with EtBr. The ladder is an Invitrogen 1kb+ ladder, so the sizes visible are 100,200,300,400,500,650.
        Attached Files
        Last edited by pbluescript; 03-05-2012, 04:59 AM. Reason: Added more information about the gel.

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by ssing View Post
          Hi there,

          I got the exact same pattern you describe (couldn't see the image). See this post for what I found, what others suggested, and to see the end results. Let me know if you'd like more information.

          Techniques and protocol discussions on sample preparation, library generation, methods and ideas


          Best of luck!
          Hi ssing,
          Glad that worked for you. However, about the gel you ran after heating/cooling -- was that agarose? If so, did you pre-stain (eg, EtBr in loading dye and/or gel and/or running buffer) or post-stain the gel?

          --
          Phillip

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by pbluescript View Post
            You can see the gel I have attached to this post for an example of the appearance of these double sized bands as the number of PCR cycles goes from 6-18.
            Hi pbluescript,
            That is a great picture. Could we get info about the sieving matrix (agarose or acrylamide?) and whether you pre or post-stained it? And what the stain was (EtBr, SyBR-safe, SyBr-gold, etc)?

            --
            Phillip

            Comment

            • pbluescript
              Senior Member
              • Nov 2009
              • 224

              #7
              Originally posted by pmiguel View Post
              Hi pbluescript,
              That is a great picture. Could we get info about the sieving matrix (agarose or acrylamide?) and whether you pre or post-stained it? And what the stain was (EtBr, SyBR-safe, SyBr-gold, etc)?

              --
              Phillip
              Sure. This one was 1.5% agarose post-stained with EtBr. I've actually tried it with 1-2% agarose post-stained with EtBr, Sybr-safe, and Sybr-gold and get identical results. I have also used the Invitrogen E-gels and see the same pattern.

              I generally take an aliquot of each library pre-PCR and titrate the number of cycles to decide how many each library needs.

              Comment

              • pmiguel
                Senior Member
                • Aug 2008
                • 2328

                #8
                Originally posted by pbluescript View Post
                Sure. This one was 1.5% agarose post-stained with EtBr. I've actually tried it with 1-2% agarose post-stained with EtBr, Sybr-safe, and Sybr-gold and get identical results. I have also used the Invitrogen E-gels and see the same pattern.

                I generally take an aliquot of each library pre-PCR and titrate the number of cycles to decide how many each library needs.
                Then I am not sure what to think. I thought most of the double peak phenomenon was the result of the extra sieving resistance BioAnalyzer chip polymer had for ssDNA vs. dsDNA. But you are showing the same phenomenon in agarose -- agarose without run-time staining. Strange.

                --
                Phillip

                Comment

                • ssing
                  Member
                  • Jan 2009
                  • 21

                  #9
                  Hi Phillip,

                  It was a pre-stained agarose gel. Thanks!

                  Sonal

                  Comment

                  • pbluescript
                    Senior Member
                    • Nov 2009
                    • 224

                    #10
                    Originally posted by pmiguel View Post
                    Then I am not sure what to think. I thought most of the double peak phenomenon was the result of the extra sieving resistance BioAnalyzer chip polymer had for ssDNA vs. dsDNA. But you are showing the same phenomenon in agarose -- agarose without run-time staining. Strange.

                    --
                    Phillip
                    I always thought it was the complimentary adapter ends from separate molecules hybridizing, increasing the overall length of the fragments. As the concentration goes up, this becomes more likely. I haven't directly tested that though.

                    Comment

                    • TonyBrooks
                      Senior Member
                      • Jun 2009
                      • 303

                      #11
                      Originally posted by pbluescript View Post
                      It is due to too many cycles of PCR. When you have a high concentration of DNA fragments with complimentary ends, they can hybridize with each other, leading to the longer fragment sizes. Did you amplify each library with the same number of cycles or did you titrate the cycle number?
                      If you titrate the number of PCR cycles, you can be sure you don't get these effects and introduce minimal PCR bias into your samples.
                      You can see the gel I have attached to this post for an example of the appearance of these double sized bands as the number of PCR cycles goes from 6-18.
                      They should pose no problem during sequencing as you denature before you load onto the cBot.
                      You could perform and additional PCR cycle with extra primers and mastermix if you want to check. The additional peak should disappear.

                      Comment

                      • pbluescript
                        Senior Member
                        • Nov 2009
                        • 224

                        #12
                        Originally posted by TonyBrooks View Post
                        They should pose no problem during sequencing as you denature before you load onto the cBot.
                        You could perform and additional PCR cycle with extra primers and mastermix if you want to check. The additional peak should disappear.
                        I agree. They've never caused me an issue. I use the appearance of the bands more as a guide to the maximum number of PCR cycles I need for a library.

                        Comment

                        • monkey_SEQ
                          Junior Member
                          • Oct 2011
                          • 6

                          #13
                          Hi All

                          This will be my first post…also first time attempting the TruSeq protocol (followed exactly - 4ug total RNA, AmPureXP beads, 15cycle PCR amp). I seemed to have gotten exactly the same result as ssing did (http://seqanswers.com/forums/showthread.php?t=15224), strong band at ±300-500bp with a HMW smear going all the way to the wells (1% EtBr pre-stained agarose gel attached). I had a heart attack when I saw this earlier today, but according to the post I have seen, it seems to be a regular result with the Truseq kit. I will try pmiguel’s advice on denaturing the libraries and rerunning the gel just to see if the smear clears up.

                          It seems then that these libraries are over amplified (if I understand correctly), but what effect will this have on a transcriptome assembling project? I am interested in read counts/ expression levels. Will an over amplified library result in too much of a bias in the results or will it be fine to continue with what I have?

                          I will still size-select for what its worth. What is a good insert size to aim for when attempting a de novo assembly?

                          Thanks for any advice
                          Attached Files

                          Comment

                          • pbluescript
                            Senior Member
                            • Nov 2009
                            • 224

                            #14
                            Originally posted by monkey_SEQ View Post
                            Hi All

                            This will be my first post…also first time attempting the TruSeq protocol (followed exactly - 4ug total RNA, AmPureXP beads, 15cycle PCR amp). I seemed to have gotten exactly the same result as ssing did (http://seqanswers.com/forums/showthread.php?t=15224), strong band at ±300-500bp with a HMW smear going all the way to the wells (1% EtBr pre-stained agarose gel attached). I had a heart attack when I saw this earlier today, but according to the post I have seen, it seems to be a regular result with the Truseq kit. I will try pmiguel’s advice on denaturing the libraries and rerunning the gel just to see if the smear clears up.

                            It seems then that these libraries are over amplified (if I understand correctly), but what effect will this have on a transcriptome assembling project? I am interested in read counts/ expression levels. Will an over amplified library result in too much of a bias in the results or will it be fine to continue with what I have?

                            I will still size-select for what its worth. What is a good insert size to aim for when attempting a de novo assembly?

                            Thanks for any advice
                            Actually, those look fine to me. The smear is probably due to the high concentration of DNA.

                            Comment

                            • lilin001
                              Junior Member
                              • Dec 2010
                              • 3

                              #15
                              Originally posted by pbluescript View Post
                              It is due to too many cycles of PCR. When you have a high concentration of DNA fragments with complimentary ends, they can hybridize with each other, leading to the longer fragment sizes. Did you amplify each library with the same number of cycles or did you titrate the cycle number?
                              If you titrate the number of PCR cycles, you can be sure you don't get these effects and introduce minimal PCR bias into your samples.
                              You can see the gel I have attached to this post for an example of the appearance of these double sized bands as the number of PCR cycles goes from 6-18.
                              Very nice. Could you please provide ladder size? Thanks!

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Nine Things a Sample Prep Scientist Thinks About Before Sequencing
                                by SEQadmin2


                                I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.

                                Here are nine questions we think about, in roughly the order they matter, before...
                                06-18-2026, 07:11 AM
                              • SEQadmin2
                                From Collection to Sequencing: Why Sample Preparation and Preservation Define Sequencing Data
                                by SEQadmin2


                                Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.


                                The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
                                ...
                                06-02-2026, 10:05 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 06-17-2026, 06:09 AM
                              0 responses
                              38 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-09-2026, 11:58 AM
                              0 responses
                              100 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-05-2026, 10:09 AM
                              0 responses
                              121 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 06-04-2026, 08:59 AM
                              0 responses
                              114 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...