#21 |
|
Member
Location: Spain Join Date: Dec 2010
Posts: 13
|
Hi FGponce,
what was the accuracy at 600 bp (98% match) of your first control test fragment run? The number of reads above 98% perfect match at 600 bp should be at least 56% for a control run (this can change when using libraries). Below that number you could have an instrument problem. You can ask Roche for a validated E coli XL+ library they prepared in Branford recently. |
|
|
|
|
|
#22 |
|
Junior Member
Location: New Zealand Join Date: Dec 2011
Posts: 9
|
Hi madseq,
So the machine looked fine with avg 62% above 98% match for the control run. The data for the sample run attached above actually contains reads from 5 different biological samples/organisms. I may expect one or two to be recalcitrant due to long stretches of AAAs etc in the genome but not all five. Also, if the machine is working well why don't the control beads act as controls and produce long reads? Surely a control that doesn't act as a control during a run with bad libraries isn't actually a control?! I'll try and get a library that has run well on a FLX+ supplied by Roche. We havent done much FLX seq is it normal to see large sample dependent read length variation from libraries of similar fragment size? |
|
|
|
|
|
#23 | |
|
Member
Location: Spain Join Date: Dec 2010
Posts: 13
|
Quote:
Good runs in the FLX+ give readlength averages of 650-700. Donīt expect readlenths of 1000-1500. And yes, the control beads perform differently depending on the sample youīre running, loading density, etc... They help to diagnose problems but they do not behave always exactly the same. You should try to have your run analized by customer support, they have the tools to tell you if your failed run can be due to the samples used or if you are having issues with the instrument. |
|
|
|
|
|
|
#24 | |
|
Senior Member
Location: Brazil Join Date: Nov 2010
Posts: 143
|
Quote:
FAS has recommended us to change the nebulization conditions from 15 psi for 1 min to 20 psi for 1 min. This sort of recommendation made me think the real problem on our preps are the really long fragments rather possibity of having small ones. I repeated the libraries today and we got a larger % of fragments around 650-1200 bp than before. I sent the traces to Roche, and I am waiting on their ok to titrate and do the LV. We are perfoming the protocol for evaluation of fragments size after the first melt during enrichment. Only if we get fragments in the right range, we will sequence. They did tell me that metagenomic samples can be harder to get to work on Plus. |
|
|
|
|
|
|
#25 | |
|
Senior Member
Location: Australia Join Date: May 2009
Posts: 143
|
Quote:
Actually, I'd like to start using our Covaris to shear the DNA, but am not quite sure what size to go for or what settings to use. I'm sure they'll just say that it's not supported and not to do it. I've used it for XLR70 RL preps just fine, but am a bit hesitant to proceed with it for the XL+ RL preps. On a side note, I did hear from Roche last week that they are recommending 6-8% enrichment recovery for best XL+ results. They said to still proceed with beads up to 20% enrichment, but that over ~8%, the signal intensities on the beads starts to get too high. I've asked that they put out some sort of bulletin or something with all their tips for a successful XL+ run since they seem to be highly variable. Good luck and let us know how the run goes! Jason |
|
|
|
|
|
|
#26 |
|
Senior Member
Location: Brazil Join Date: Nov 2010
Posts: 143
|
Thanks, Jason. I will most definitely let you guys know.
I havenīt heard anything about enrichment rates. Our bad run it was with beads with very low enrichment. Problem is we used a box of emPCR reagent that expired on nov/11, this way we cannot get replacement for our run. I honestly donīt think that was the problem but I cannot argue with Roche on that, as using of expired reagents are not supposed to happen. Anyway, now we are doing all this tests under their supervision, with non expired reagents. We aslo got rid of the bead recovery reagents lots that have been showing problems. I agrre theu should release a bulletin with tips, like they did for amplicon sequencing. Could you do me a favor, and post the Agilent traces from your best and worse run, so I can see how they look like? |
|
|
|
|
|
#27 |
|
Member
Location: Branford Join Date: Apr 2011
Posts: 17
|
Hi Jason,
The recommended enrichment percentage for the GS FLX+, GS FLX and GS Junior platforms remains between 5% and 20%. Within this window, there is no correlation between increasing enrichment percentage and signal per base. A higher signal per base is typically indicative of other issues. Please feel free to contact your local Roche representative with any issues or concerns and visit our website at www.my454.com/my454 for the most up to date guidelines. Regards, Nicole Technical Support Scientist 454 Life Sciences, A Roche Company |
|
|
|
![]() |