    Bioanalyser traces + QC postPCR. ChIPseq Illumina protocol

    Hi everyone.
    I have been trying to get ChIPseq libraries. I started with 10ng as suggested by Illumina and got a Bioanalyser trace. Problems: the trace is actually two quite-fused peaks. The first at around 165bp and the second at around 224bp (I cut the gel from 125-150bp to 300bp).
    Is this normal?
    Another question would be if anyone does a QC after the protocol's PCR. imagine you start with the INPUT sample (pre-immunoprecipitated DNA. Your ChIP background control, basically) which is normal fragmented genomic DNA. Before library preparation, you have the same amount for all genomic locations (as seen on RT-PCR). But after library PCR, there is a quite strong bias for some regions compared to others. Is this normal? do you do this QC step or go direcly into sequencing if the library looks fine on the Bioanalyser?
    Cheers everyone in advance!
  • Susanne
    Member
    • Aug 2009
    • 33

    #2
    Did you try to run the IP-ed material (on the Bioanalyzer) before sample prep.? This may give you a clue if the size distribution looked like this before adding the adapters.
    Sorry, not the complete solution, but maybe it give a hint...

    Comment

    Latest Articles

    Collapse

    ad_right_rmr

    Collapse

    News

    Collapse

    Topics Statistics Last Post
    Started by SEQadmin2, Yesterday, 11:58 AM
    0 responses
    10 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-05-2026, 10:09 AM
    0 responses
    25 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-04-2026, 08:59 AM
    0 responses
    35 views
    0 reactions
    Last Post SEQadmin2  
    Started by SEQadmin2, 06-02-2026, 12:03 PM
    0 responses
    58 views
    0 reactions
    Last Post SEQadmin2  
    Working...