    #16
    I would caution against blindly filtering your metagenomic data against any database of contaminants. Ideally, you would have negative controls run alongside your samples that could be checked for the presence of these contaminants instead....

    Comment

    • Brian Bushnell
      Super Moderator
      • Jan 2014
      • 2709

      #17
      I won't say either one of you is right or wrong, and negative controls are always a good idea. But, the reason I put together the bacterial contaminant file is because JGI is not capable of distinguishing between actual samples, and contaminants in that file. With sufficient amplification (like single cells), some wells may have high levels of a contaminant that is zero in other wells, since it only takes one particle. Some of them, like Pseudomonas, are present in reagents. Others, like E.coli, are present on human skin, and often make their way into the libraries.

      I have seen dozens of posters that incorrectly claim Pseudomonas or various other common contaminants are endemic to some environment. But it's likely artifact of poor quality control. So, I encourage you to be very cautious.

      *Edit. For reference, JGI no longer sequences anything on that list.
      Last edited by Brian Bushnell; 07-17-2017, 08:58 AM.

      Comment

      • ybukhman
        Junior Member
        • Jan 2011
        • 6

        #18
        RE: filtering soil metagenomics data

        Thank you, fanli and Brian! I guess one possibility is not to filter initially, but then check the final assembly against the contaminant files just to find out if some of the species that I am detecting are known contaminants?

        Comment

        • Brian Bushnell
          Super Moderator
          • Jan 2014
          • 2709

          #19
          That's certainly a good possibility. If you have zero or trivial amounts of common contaminants, don't bother filtering. If you have a lot... then, try to figure out whether you have an exact strain match, which greatly boosts the likelihood of it originating in a reagent.

          Comment

          • ybukhman
            Junior Member
            • Jan 2011
            • 6

            #20
            Thanks a lot, Brian! I'll give it a try.

            Comment

            • yang zhang
              Junior Member
              • Aug 2017
              • 2

              #21
              Hi Brain,

              This might be a silly question, but do we need to index the reference every time for every sample?

              Thanks!
              Last edited by yang zhang; 08-27-2017, 11:27 AM.

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #22
                Originally posted by yang zhang View Post
                Hi Brain,

                This might be a silly question, but do we need to index the reference every time for every sample?

                Thanks!
                No. You can create an index of the reference up-front by doing
                Code:
                 bbmap.sh in=ref.fa
                that will create a "ref" directory, which will contain all index files. Do not worry about the contents of the folder since they are arranged in a way BBMap requires them.

                In future when you want to use this index replace "ref=" with "path=/path_to_directory_containing_ref_folder" in your command line.

                Comment

                • yang zhang
                  Junior Member
                  • Aug 2017
                  • 2

                  #23
                  I see. Thank you for the clarification, GenoMax!

                  Comment

                  • Coyk
                    Member
                    • Jun 2011
                    • 15

                    #24
                    Hi Brian,
                    I downloaded the files above and have successfully indexed the cat and dog file. I tried the command:

                    bbmap.sh minid=0.95 maxindel=3 bwr=0.16 bw=12 quickmatch fast minhits qtrim=rl trimq=10 untrim Xmx23g in=cleanAR1.fastq.gz outu=clean2AR1.fastq.gz outm=catAR1.fq

                    and got the following error:
                    Exception in thread “main” java.lang.NumberFormatException: null
                    at java.lang.Integer.parseInt(Integer.java:542)
                    at java.lang.Integer.parseInt(Integer.java:615)
                    at align2.AbstractMapper.parse(AbstractMapper.java:449)
                    at align2.AbstractMapper.<init>(AbstractMapper.java:54)
                    at align2.BBMap.<init>(BBMap.java:42)
                    at align2.BBMap.main(BBMap.java:30)

                    The same happened when I tried the dog masked file. I was able to successfully remove the human contamination using the masked file you provided and the commands above. Is this command not applicable to the cat, dog, and mouse files you provided? Is there an extra step I am missing? I am not versed in java so I don't know how to interpret the error.

                    coyk

                    Comment

                    • Kristian_Andersen
                      Junior Member
                      • Jul 2015
                      • 4

                      #25
                      Hi Brian,

                      Just reviving an old thread here. I have been testing out a lot of different methods to clean human reads and I really love BBMap because it's such a well thought-out program. However, when I try to clean human reads with the settings you have specified, I routinely get a ton of reads remaining - upwards of 70% (so only 30% are cleaned). I have tried to adjust the various parameters, but the only thing that seems to make a difference for depletion is the 'minid' setting. Setting that at 0.50 (which is *very* low) depletes around 95% of reads. As a comparison, a default run with bwa mem depletes 100%.

                      Any idea how I might get BBMap to more accurately deplete human reads?

                      Comment

                      • GenoMax
                        Senior Member
                        • Feb 2008
                        • 7142

                        #26
                        Have you tried using "bbsplit.sh" with human genome to see if that works better. If you are interested in non-human data then I would use the non-masked genome and risk losing a few additional reads.

                        Comment

                        • Kristian_Andersen
                          Junior Member
                          • Jul 2015
                          • 4

                          #27
                          I have not - however, bbsplit doesn't really seem to be the right tool for removing human reads?

                          Comment

                          • GenoMax
                            Senior Member
                            • Feb 2008
                            • 7142

                            #28
                            "bbsplit.sh" is a general purpose tool that will bin reads into any number of bins (depending on the reference sequences provided, you can provide as many as you want). In this case you would provide human_genome.fa (and any other reference you want to use). If you only use human then reads not mapping to human genome will be collected in other bin.

                            Comment

                            • jylee
                              Junior Member
                              • Jun 2019
                              • 3

                              #29
                              Hello Brian,

                              I am trying to use removehuman.sh on MSU HPCC.

                              Inputs (*_filtered.fastq.gz files) are phix filtered R1 and R2 files using BBduk as following: [leejooy5@dev-intel18 filtered_reads]$ bbduk.sh -Xmx10g in1=NFW_R1_trimmed.fastq.gz in2=NFW_R2_trimmed.fastq.gz out1=NFW_R1_filtered.fastq.gz out2=NFW_R2_filtered.fastq.gz ref=/opt/software/BBMap/37.93-foss-2018a/resources/phix174_ill.ref.fa.gz k=31 hdist=1 stats=GR25_stats.txt threadS=8)

                              As shown in below, error message "Exception in thread "main" java.lang.RuntimeException: Can't find file /global/projectb/sandbox/gaag/bbtools/hg19/ref/genome/1/summary.txt
                              " popped up when I tried to run "removehuman.sh". I tried without additional parameters such as -Xmx and threads, but same error happened. Also, I tried to find the find the
                              file "/global/projectb/sandbox/gaag/bbtools/hg19/ref/genome/1/summary.txt", but I couldn't. Could you tell me what mistake I did or let me know where I can find a solution? Thank you for your time and consideration.

                              Cheers,
                              Joo-Young

                              =======================
                              [leejooy5@dev-intel18 filtered_reads]$ removehuman.sh -Xmx10g in1=NFW_R1_filtered.fastq.gz in2=NFW_R2_filtered.fastq.gz out1=NFW_R1_clean.fastq.gz out2=NFW_R2_clean.fastq.gz threads=8

                              removehuman.sh -Xmx10g in1=NFW_R1_filtered.fastq.gz in2=NFW_R2_filtered.fastq.gz out1=NFW_R1_clean.fastq.gz out2=NFW_R2_clean.fastq.gz threads=8
                              java -Djava.library.path=/opt/software/BBMap/37.93-foss-2018a/jni/ -ea -Xmx10g -cp /opt/software/BBMap/37.93-foss-2018a/current/ align2.BBMap minratio=0.9 maxindel=3 bwr=0.16 bw=12 quickmatch fast minhits=2 path=/global/projectb/sandbox/gaag/bbtools/hg19 pigz unpigz zl=6 qtrim=r trimq=10 untrim idtag usemodulo printunmappedcount usejni ztd=2 kfilter=25 maxsites=1 k=14 -Xmx10g in1=NFW_R1_filtered.fastq.gz in2=NFW_R2_filtered.fastq.gz out1=NFW_R1_clean.fastq.gz out2=NFW_R2_clean.fastq.gz threads=8
                              Executing align2.BBMap [tipsearch=20, maxindel=80, minhits=2, bwr=0.18, bw=40, minratio=0.65, midpad=150, minscaf=50, quickmatch=t, rescuemismatches=15, rescuedist=800, maxsites=3, maxsites2=100, minratio=0.9, maxindel=3, bwr=0.16, bw=12, quickmatch, minhits=2, path=/global/projectb/sandbox/gaag/bbtools/hg19, pigz, unpigz, zl=6, qtrim=r, trimq=10, untrim, idtag, usemodulo, printunmappedcount, usejni, ztd=2, kfilter=25, maxsites=1, k=14, -Xmx10g, in1=NFW_R1_filtered.fastq.gz, in2=NFW_R2_filtered.fastq.gz, out1=NFW_R1_clean.fastq.gz, out2=NFW_R2_clean.fastq.gz, threads=8]
                              Version 37.93 [tipsearch=20, maxindel=80, minhits=2, bwr=0.18, bw=40, minratio=0.65, midpad=150, minscaf=50, quickmatch=t, rescuemismatches=15, rescuedist=800, maxsites=3, maxsites2=100, minratio=0.9, maxindel=3, bwr=0.16, bw=12, quickmatch, minhits=2, path=/global/projectb/sandbox/gaag/bbtools/hg19, pigz, unpigz, zl=6, qtrim=r, trimq=10, untrim, idtag, usemodulo, printunmappedcount, usejni, ztd=2, kfilter=25, maxsites=1, k=14, -Xmx10g, in1=NFW_R1_filtered.fastq.gz, in2=NFW_R2_filtered.fastq.gz, out1=NFW_R1_clean.fastq.gz, out2=NFW_R2_clean.fastq.gz, threads=8]

                              Set MINIMUM_ALIGNMENT_SCORE_RATIO to 0.650
                              Set MINIMUM_ALIGNMENT_SCORE_RATIO to 0.900
                              Set threads to 8
                              Retaining first best site only for ambiguous mappings.
                              Exception in thread "main" java.lang.RuntimeException: Can't find file /global/projectb/sandbox/gaag/bbtools/hg19/ref/genome/1/summary.txt
                              at fileIO.ReadWrite.getRawInputStream(ReadWrite.java:906)
                              at fileIO.ReadWrite.getInputStream(ReadWrite.java:871)
                              at fileIO.TextFile.open(TextFile.java:227)
                              at fileIO.TextFile.<init>(TextFile.java:71)
                              at dna.Data.setGenome2(Data.java:822)
                              at dna.Data.setGenome(Data.java:768)
                              at align2.BBMap.loadIndex(BBMap.java:313)
                              at align2.BBMap.main(BBMap.java:32)

                              Comment

                              • GenoMax
                                Senior Member
                                • Feb 2008
                                • 7142

                                #30
                                @jylee: "/global/projectb/sandbox/gaag/bbtools/hg19/ref/genome/1/summary.txt" appears to refer to a location on JGI servers (if that is not your own). You will need to download and provide hg19 reference sequence. You can pre-index the genome with BBMap to use with path= or use ref= option to point to the genome sequence multi-fasta file location.

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                                  by SEQadmin2


                                  Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                                  The systematic characterization of the human proteome has
                                  ...
                                  07-20-2026, 11:48 AM
                                • SEQadmin2
                                  Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                                  by SEQadmin2



                                  Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                                  ...
                                  07-09-2026, 11:10 AM
                                • SEQadmin2
                                  Cancer Drug Resistance: The Lingering Barrier to Rising Survival
                                  by SEQadmin2



                                  Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

                                  There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
                                  07-08-2026, 05:17 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 07-24-2026, 12:17 PM
                                0 responses
                                21 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-23-2026, 11:41 AM
                                0 responses
                                19 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-20-2026, 11:10 AM
                                0 responses
                                26 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 07-13-2026, 10:26 AM
                                0 responses
                                38 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...