Originally posted by vanessamata
View Post
Unconfigured Ad
Collapse
X
-
I don't really see a pattern in the tag/plate failures, either. But if I understand correctly, the 6, 8, 9, 10 tags that failed the first time also failed the second? And Tag 7, which worked the first time, worked this time, too?
The only thing that makes sense to me here is that you're having an issue with the structure of your custom primers in the first PCR (the ones that are adaptor + 5bp tag + primer). I assume you've double checked all the sequences and didn't reverse complement something you weren't supposed to, otherwise you probably wouldn't have a 50% success rate. But have you checked the primers for things like dimers and hairpin loops? Given the fact that you have a low number of reads mapping to the failed tags (~50 reads per sample, you said in your first post) that could be random errors causing misidentification or it may be the result of a strong primer dimerization with only a few oligos available for amplification.
Comment
-
-
Yes. 6,8,9, and 10 failed on both runs. And 7 worked on both runs.Originally posted by Jessica_L View PostI don't really see a pattern in the tag/plate failures, either. But if I understand correctly, the 6, 8, 9, 10 tags that failed the first time also failed the second? And Tag 7, which worked the first time, worked this time, too?
The only thing that makes sense to me here is that you're having an issue with the structure of your custom primers in the first PCR (the ones that are adaptor + 5bp tag + primer). I assume you've double checked all the sequences and didn't reverse complement something you weren't supposed to, otherwise you probably wouldn't have a 50% success rate. But have you checked the primers for things like dimers and hairpin loops? Given the fact that you have a low number of reads mapping to the failed tags (~50 reads per sample, you said in your first post) that could be random errors causing misidentification or it may be the result of a strong primer dimerization with only a few oligos available for amplification.
Yes, the only thing that makes sense to me is that for some (molecular) reason the structure of these primers/tags are not allowing the sequencing to occur. I tested everything on agarose gel. I have no primer dimer and the amplifications are all of the expected size (and the same) with all primers/tags. They all incorporate the nextera index as well.
Comment
-
-
I'll admit, this one has me stumped. If the library construction appears to have been successful, these should be sequencing. A few last questions:
Have you qPCR'd the individual libraries with the 5-bp tags? i.e. a separate set of reactions for each of tags 6, 8, 9, 10 with no pooling? It sounded like when you did the qPCR initially that it was on multiple libraries pooled together and I'm curious if you have done, or if you can do, some without pooling them. If the qPCRs work(ed), there's absolutely no reason for the sequencing to fail as any kind of molecular inhibition would/should affect the qPCR as well.
Lastly, can you isolate the ~50 reads from the tags that fail? Can you get them through an analysis pipeline for variant calling? What do they look like in IGV? Are the tags there and complete? The only two culprits I can think of at this point are things like binding to the flow cell and whether the sequencing primer binds correctly.
Comment
-
-
@vanessamata: I am not an experimental expert so take this with a grain of salt until someone confirms it but have you thought about heating your sample before loading? That should help break up any strange structures that may be forming. I remember this working for some difficult to sequence samples in past.
Comment
-
-
That's a good point, GenoMax-- for the TruSeq Amplicon v1 that I've used, the library that's eluted off beads post-normalization is single stranded in dilute NaOH and that requires a short heating followed by snap cooling on ice prior to loading. It's not a bad idea to try. The Illumina protocol for the TruSeq Amplicon calls for 96 degrees C for 2 minutes, followed by 5 minutes in an ice water bath, if you wanted to try something similar.
Comment
-
-
Considering discussion so far the most likely cause would be errors in overhang primer design or synthesis such as a 3' single base mismatch which prevents priming sequencing. I would check sequences on the oligo tubes to see if it matches with the intended sequences. Also 5 out of 6 unsequenced tags start with T but it would less likely be the cause.
Comment
-
-
@GenoMax and Jessica-L
I think you might be right!
The 16S metagenomics illumina protocol also mentions this step! For some reason the technician responsible for preparing the libraries to load into MiSeq was not doing it!
Next time I will try to heat up the library and see if this still happens or not!
Thanks a lot!
Comment
-
Latest Articles
Collapse
-
by SEQadmin2
I’m not a sequencing expert. I’m a purification scientist who uses NGS to evaluate workflows my group develops. With this perspective, we think about the sample first and the NGS workflow second. The sequencer is an exceptionally honest reporter, but it can only report on what you give it, so whether you get clean, interpretable data from an NGS workflow is largely determined before you begin.
Here are nine questions we think about, in roughly the order they matter, before...-
Channel: Articles
06-18-2026, 07:11 AM -
-
by SEQadmin2
Data variability is still an issue in sequencing technologies despite the advances in reproducibility and accuracy of these platforms. But the problem does not originate in the sequencing itself, but in the previous steps, before the sample reaches the sequencer.
The first step is collection, followed by preservation and sample preparation for analysis. Most scientists overlook those steps, but not being careful might just be skewing the experiment’s results.
...-
Channel: Articles
06-02-2026, 10:05 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Started by SEQadmin2, Yesterday, 05:37 AM
|
0 responses
6 views
0 reactions
|
Last Post
by SEQadmin2
Yesterday, 05:37 AM
|
||
|
Started by SEQadmin2, 06-26-2026, 11:10 AM
|
0 responses
16 views
0 reactions
|
Last Post
by SEQadmin2
06-26-2026, 11:10 AM
|
||
|
Whole-Genome Sequencing Traces Faroe Islands Ancestry to a North Atlantic Founder Population
by SEQadmin2
Started by SEQadmin2, 06-17-2026, 06:09 AM
|
0 responses
51 views
0 reactions
|
Last Post
by SEQadmin2
06-17-2026, 06:09 AM
|
||
|
Sequencing the Two-Toed Sloth Genome Reveals Jumping Genes Tied to Its Extreme Metabolism
by SEQadmin2
Started by SEQadmin2, 06-09-2026, 11:58 AM
|
0 responses
110 views
0 reactions
|
Last Post
by SEQadmin2
06-09-2026, 11:58 AM
|

Comment