  • Rapid Run vs. High Output mode in HiSeq 2500

    Hi, we have been considering about using HiSeq 2500 to sequence ~100 cancer genomes. We were wondering why the Rapid Run mode saves so much time compared to the High Output mode (500 cycles in < 3 days versus 250 cycles in 6 days, if I understand correctly). Since the chemical procedure for each cycle seems to be quite similar in both mode (nucleotide incorporation -> wash -> imaging -> remove blocking and fluor group -> wash) and all reactions happen in a parallel manner, where comes the substantial difference in run time?

    Many thanks in advance!

  • #2
    Originally posted by xiang.jiao View Post
    Hi, we have been considering about using HiSeq 2500 to sequence ~100 cancer genomes. We were wondering why the Rapid Run mode saves so much time compared to the High Output mode (500 cycles in < 3 days versus 250 cycles in 6 days, if I understand correctly). Since the chemical procedure for each cycle seems to be quite similar in both mode (nucleotide incorporation -> wash -> imaging -> remove blocking and fluor group -> wash) and all reactions happen in a parallel manner, where comes the substantial difference in run time?

    Many thanks in advance!
    Primarily from having a small surface area that gets imaged. The 'high output mode' flow cell has eight channels while the 'Rapid Run' flow cell has only two. There may be some other differences in the chemistry, but surface area (which translates to imaging time) is the biggest.
    AllSeq - The Sequencing Marketplace
    [email protected]
    www.AllSeq.com

    Comment


    • #3
      Thanks a lot AllSeq! That is exactly what we were looking for! By the way, do you how long the imaging time is for one channel?

      Comment


      • #4
        No, I'm not sure of the specifics. Hopefully someone with a 2500 will chime in.
        AllSeq - The Sequencing Marketplace
        [email protected]
        www.AllSeq.com

        Comment


        • #5
          The imaging time is ~10 minutes per cycle. Also, the chemistry and cycling parameters are different from the high-throughput mode.

          Comment

          Latest Articles

          Collapse

          • seqadmin
            Strategies for Sequencing Challenging Samples
            by seqadmin


            Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
            03-22-2024, 06:39 AM
          • seqadmin
            Techniques and Challenges in Conservation Genomics
            by seqadmin



            The field of conservation genomics centers on applying genomics technologies in support of conservation efforts and the preservation of biodiversity. This article features interviews with two researchers who showcase their innovative work and highlight the current state and future of conservation genomics.

            Avian Conservation
            Matthew DeSaix, a recent doctoral graduate from Kristen Ruegg’s lab at The University of Colorado, shared that most of his research...
            03-08-2024, 10:41 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by seqadmin, Yesterday, 06:37 PM
          0 responses
          10 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, Yesterday, 06:07 PM
          0 responses
          9 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 03-22-2024, 10:03 AM
          0 responses
          49 views
          0 likes
          Last Post seqadmin  
          Started by seqadmin, 03-21-2024, 07:32 AM
          0 responses
          67 views
          0 likes
          Last Post seqadmin  
          Working...
          X