Default shearing question using Covaris

Would like to understand something when I use Covaris for shearing.

I have DNA that has been diluted to 5ug in 120ul for shearing or 41ng/ul. I shear and measure concentration by Picogreen. But it is about 30ng/ul after shearing. Why does the concentration drop? It is only sheared DNA, not cleaned. Is it due to the DNA being less compact after shearing? Does it have anything to do that DNA is fragmented now and this causes the concentration to drop? I have tried other DNAs and their concentrations drop just like this.
seqgirl123 is offline   Reply With Quote
Old 04-27-2011, 08:11 PM   #2
ECO
--Site Admin--
 
Location: SF Bay Area, CA, USA

Join Date: Oct 2007
Posts: 1,358
Default

Single stranded regions. (?)
ECO is offline   Reply With Quote
Old 04-28-2011, 05:51 AM   #3
LMcSeq
Member
 
Location: Maryland--we have a HiSeq too!

Join Date: Feb 2009
Posts: 33
Default

Agreed-PicoGreen is intended to measure double stranded DNA. If the shearing resulted in any single stranded material, that would not be detected.
LMcSeq is offline   Reply With Quote
Old 04-28-2011, 11:17 AM   #4
Hamid
Senior Member
 
Location: San Diego, CA

Join Date: Sep 2009
Posts: 105
Default

Hi seqgirl123,

What you are noticing is how pico green method is used for concentration determination. Two points of interest:

1. Pico green binding to DNA and its intensity is dependent on the size of DNA. It is recommended that you generate a standard curve using DNA in in the same size range as the distribution of your fragments after shearing. If you use a standard curve that has been generated using large DNA such as lambda, the concentration from that standard curve will not apply accurately to the smaller sheared material.
2. Quite a few service labs use a combination of pico green and A260/A280 to check the quality and concentration of their samples. If you test the samples you mentioned on a nano drop or equivalent system, you will notice that there is very very minimal loss of your samples.

Thank you

Hamid
Hamid is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 07:24 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO