Old 07-20-2010, 11:16 AM   #1
454User
Junior Member
 
Location: Mid-Atlantic, USA

Join Date: Oct 2009
Posts: 5
Default LMW band in paired end libraries?

Hello all,

To those of you that prepare Ti 3 and 8 kb PE libraries, have you ever experienced a prominent, low molecular weight band around 100-110 nt in your ssDNA product? Please see the attached images for examples.

I have seen this band without fail in many PE preps that span reagent lots and operator experience. While its abundance relative to the major library band varies, its presence does not. I am of the mind that this band represents:

1.) Library adaptor dimers (though these adaptors are not phosphorylated and I would not expect PNK to carry through the end polish cleanup)

or

2.) Amplification primer dimers that display this size due to their %GC

Any comments or insight would be greatly appreciated!

-J
Attached Images
File Type: jpg 3kbPE1.JPG (32.6 KB, 27 views)
File Type: jpg 3kbPE2.JPG (36.7 KB, 17 views)
454User is offline   Reply With Quote
Old 07-23-2010, 04:00 AM   #2
gilly
Junior Member
 
Location: East of England

Join Date: Nov 2008
Posts: 8
Default

Hi.

Yes we do see them occasionally I believe the cause is amplification primer dimmers. If you take these libraries into emPCR then you run the risk of getting high enrichment with an increase in very short or even null beads.

This peak can be removed by performing a 4-cycle PCR, using identical conditions to those listed in the paired end manual (except with the cycle number reduced to 4 obviously). This step is to double-strand all of your products. This is then followed by a standard SPRI clean up (1.8x SPRI to sample ratio, as per Agencourt’s protocol).
gilly is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 10:21 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO