Old 10-13-2011, 03:09 AM   #1
GraemeFox
Member
 
Location: Manchester

Join Date: Oct 2011
Posts: 14
Default Rapid Library Library Quantitation using concentration not RFU

Hi,
I am attempting the Library Quantitation phase of the Rapid Library Prep. set up
ie. Producing a standard curve of fluorescence through 8 serial dilutions of RL
Standard.

However the fluorometer I am using (Invitrogen Qubit 2.0) does not seem to give the option of giving results in RFU.

So....I am able to measure the concentration of my sample in ng/uL.

BUT....the 'Rapid Library Calculator' only accepts measurements in RFU.

AND...I need to ultimately calculate the concentration of my library in molecules/uL.

Is there a way to work around this that anybody knows of?
GraemeFox is offline   Reply With Quote
Old 10-13-2011, 05:24 AM   #2
n2onme
Junior Member
 
Location: US

Join Date: Jan 2011
Posts: 6
Default

Try the link below. It is used to convert ng/uL to molecules/uL with fragment size correction. You could always get a Kapa Biosystems qPCR Kit to quantitate your libraries.

http://www.uri.edu/research/gsc/resources/cndna.html
n2onme is offline   Reply With Quote
Old 10-13-2011, 05:33 AM   #3
TonyBrooks
Senior Member
 
Location: London

Join Date: Jun 2009
Posts: 298
Default

Doesn't the Qubit use Picogreen chemistry for fluorescence?
Therefore, fluorescence should be proportional to mass of DNA within the range of the Qubit standard curve.

I never trust the fluorescence readings for quantification as you assume all library fragments have adapters ligated at both ends. I always use the Kapa qPCR for my rapid libraries
TonyBrooks is offline   Reply With Quote
Old 10-14-2011, 12:13 AM   #4
GraemeFox
Member
 
Location: Manchester

Join Date: Oct 2011
Posts: 14
Default

Thanks for the suggestions.

Having 'explored' the fluorometer a little more thoroughly I found that it does indeed give me RFU values.

This despite a phone call to Invitrogen who said that it wasn't possible on this unit!

Thanks for your help.
GraemeFox is offline   Reply With Quote
Old 10-16-2011, 08:34 AM   #5
madseq
Member
 
Location: Spain

Join Date: Dec 2010
Posts: 13
Default

Hi,

GraemeFox, Ive heard that the Qubit fluorometer is not able to measure RL FAM (I dont know if this is because of wavelength or sensitivity) . Do you obtain reasonable numbers when you measure serial dilutions of the standard?
madseq is offline   Reply With Quote
Old 10-17-2011, 12:16 AM   #6
GraemeFox
Member
 
Location: Manchester

Join Date: Oct 2011
Posts: 14
Default

Madseq - when performing the serial dilutions I used picogreen so I didn't have any problem.

However, my library itself was quantified using the FAM incorporated into the rapid library primers and worked fine on the Qubit.
GraemeFox is offline   Reply With Quote
Old 10-17-2011, 11:31 AM   #7
madseq
Member
 
Location: Spain

Join Date: Dec 2010
Posts: 13
Default

Hi GraemeFox,

When you finally measured your library FAM fluorescence, did you quantify serial dilutions of the RL standard? Roche states that the Qubit is not valid for RL FAM quantification, Ive just asked to a Roche application specialist and he says that the problem is about sensitivity.
madseq is offline   Reply With Quote
Old 10-17-2011, 06:18 PM   #8
RCJK
Senior Member
 
Location: Australia

Join Date: May 2009
Posts: 155
Default

Hi GraemeFox,

I was wondering how are you able to obtain the RFU values from the Qubit? Thanks!
RCJK is offline   Reply With Quote
Old 10-21-2011, 12:23 AM   #9
GraemeFox
Member
 
Location: Manchester

Join Date: Oct 2011
Posts: 14
Default

Quote:
Originally Posted by RCJK View Post
Hi GraemeFox,

I was wondering how are you able to obtain the RFU values from the Qubit? Thanks!
Hi, On the 'Data' page on the unit itself you can scroll across to the right using a fairly inconspicuous scroll bar just below the rows of data.

Alternatively if you export your data via the USB port and open it up in Excel, the RFU values are also given.

Hope that helps!
GraemeFox is offline   Reply With Quote
Old 10-21-2011, 12:27 AM   #10
GraemeFox
Member
 
Location: Manchester

Join Date: Oct 2011
Posts: 14
Default

Quote:
Originally Posted by madseq View Post
Hi GraemeFox,

When you finally measured your library FAM fluorescence, did you quantify serial dilutions of the RL standard? Roche states that the Qubit is not valid for RL FAM quantification, Ive just asked to a Roche application specialist and he says that the problem is about sensitivity.

I didn't have a problem quantifying the FAM fluorescence of my library, or at least I didn't seem to.
Will certainly be looking into the sensitivity problem you mention though.
GraemeFox is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts