Did a run last week on a v1 hardware MiSeq using a v2 chemistry 500 cycle kit on qPCR titrated libraries. Clustered at 10 pM. Instead of the expected 800-1000 clusters/mm^2 we got 425.
Again these were with KAPA qPCR titrated libraries using the same protocol that usually allows us to nail cluster density. These TruSeq RNAseq libraries were a little odd in that the insert sizes were a mean of 700-800 bp to get the full value of the 250 base reads. But we have gotten dead-on accurate clustering results for libraries with much larger mean inserts sizes than these in the past. (Obviously this does require compensation for the larger amplicon sizes compared to the KAPA standards.) This is the first time we have used v2 chemistry, however. But Illumina Tech support says there should be no major difference in cluster densities with v2. I have another run to do with another set of similar libraries, but I don't want to waste the reagents getting 1/2 the amount of data I should. On the other hand clustering at 20 pM seems too aggressive. Anyone seeing much lower than expected cluster densities running v2 MiSeq chemistry? Especially with fairly long amplicons? -- Phillip |
![]() |
![]() |
![]() |
#2 |
Member
Location: Madison Join Date: Feb 2010
Posts: 18
|
![]()
We had the same problem here. We used to cluster at 8pM and that resulted in about 700k/mm2. Switching to v2, we ended up with 200k/mm2. Loaded the latest run at 15pm and that seemed to do the trick (1000k/mm2). Will probably set up the next run at 13 or 14pM.
|
![]() |
![]() |
![]() |
#3 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
Thanks Bucky,
I have been stripping my gears trying to figure out what was going on. That makes it seem like v2 performs more like a HiSeq with respect to raw cluster density[1]. Are you running v1 hardware, or has your instrument been upgraded to v2? Anyone else see this or not see this? [1]Of course as a matter of efficiency the cBot blows the MiSeq away -- 120 ul/lane for the HiSeq gets you the same density that 600 ul gets you for the MiSeq. -- Phillip |
![]() |
![]() |
![]() |
#4 |
Member
Location: Madison Join Date: Feb 2010
Posts: 18
|
![]()
Hey Philip,
We are using v1 hardware. I should also mention that I pooled my sample at 4pm and then diluted it out more than usual in order to get a lower NaOH final concentration. Maybe that helped as well.. |
![]() |
![]() |
![]() |
#5 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,087
|
![]()
Philip,
Disclaimer: I do not work in the lab. I am merely passing the following info along. We are not seeing this problem with v.2. kits but the inserts in question are not long. |
![]() |
![]() |
![]() |
#6 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
Could you elaborate, please? -- Phillip |
|
![]() |
![]() |
![]() |
#7 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() |
![]() |
![]() |
![]() |
#8 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,087
|
![]() |
![]() |
![]() |
![]() |
#9 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
Okay, this may be a v1 hardware/v2 reagent thing then. I heard from another core that they saw lower cluster densities ("like 1/2") using v2 chemistry on a v1 instrument. But that apparently went away when they were upgraded to v2 hardware.
-- Phillip |
![]() |
![]() |
![]() |
#10 | |
Member
Location: Madison Join Date: Feb 2010
Posts: 18
|
![]() Quote:
|
|
![]() |
![]() |
![]() |
#11 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() |
![]() |
![]() |
![]() |
#12 |
Senior Member
Location: Oklahoma Join Date: Sep 2009
Posts: 411
|
![]()
Illumina put out a new methodology for denaturation with the v2 chemistry. It's in the new Rev E version of the Miseq users guide.
|
![]() |
![]() |
![]() |
#13 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
It is different from the HiSeq -- there the concentration of NaOH used is 1/2 that than for the MiSeq denaturation. Is that new with Rev E of the MiSeq UG? Anyway, we were using the 0.2 N NaOH, just as called for in Rev E. -- Phillip |
|
![]() |
![]() |
![]() |
#14 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
After complaining the Illumina about this issue they told me they did not equipment in-house to test it out. (Meaning, I presume, all their in-house MiSeq have v2 hardware.) So they offered to send a couple of MiSeq kits out to test this hypothesis. Just started the v1 kit this morning.
Clustered at 10 pM and got 758 Kclusters/mm^2. That seems a little lower than normal, I think. Actually, I ran the same library on v2 chemistry 2 days ago at 15 pM and got a cluster density less than 500 Kclusters/mm^2. But Illumina wants the experiment to be done by the same person, same day, etc. I'll let it run until cycle 26 and then abort the run, so I can get the second run started today. -- Phillip |
![]() |
![]() |
![]() |
#15 |
Member
Location: Madison Join Date: Feb 2010
Posts: 18
|
![]()
Cool, let us know what you find out!
|
![]() |
![]() |
![]() |
#16 |
Member
Location: Ft. Detrick, MD Join Date: Aug 2008
Posts: 50
|
![]()
We've been getting low cluster densities since our MiSeq was upgraded. I used to load at 8 pM (by Kapa qPCR) and get ~500-600K / mm2. Our last 2 runs were loaded at 15 pM and only gave ~150-175K / mm2. I tried taking the number of reads obtained from those low density runs to generate a "corrected" concentration. Started that today and am getting 1400K / mm2 with 5% PF. So much for that idea....
|
![]() |
![]() |
![]() |
#17 |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
My v2 run did cluster at a lower density, but nowhere near the difference I saw previously (~50%). V1 and v2 runs were clustered at 758 vs. 652 Kclusters/mm^2, respectively for 10 pM.
I changed a few things. Not sure if they were salient: (1) My library had been over-diluted down to 1 nM. So, for my previous load (a couple of days ago) at 15 pM I had added 1 ul of 2M NaOH to 19 ul of my library to reach the desired denaturation conditions. (0.1 M NaOH.) This clustered at 430 on a v2 500 cycle run. So, I used a speed vap (vacuum only, no heat) to bring 80 ul of the sample split among 4 tubes down to 40 ul. Hence 2nM. This I could use the standard MiSeq denaturation technique on. (2) Our lab is a little chilly -- 19 oC. So I actually set a heat block at 25 oC for the denturation incubation. (3) I pre-chilled the HT1 (on ice) to be used to neutralize the denaturation reaction. So, what is the difference? -- Phillip |
![]() |
![]() |
![]() |
#18 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]()
Just to underline the implications of what Bucky wrote upthread:
Note from pp 74-75 of the rev. E MiSeq manual: Quote:
So to follow both the recommended v2 load concentration (~12.5 pM) and load at a final NaOH concentration of no more than 1 mM would require an initial library concentration of > 5nM!!! -- Phillip |
|
![]() |
![]() |
![]() |
#19 | |
Member
Location: Germany Join Date: Dec 2010
Posts: 80
|
![]() Quote:
Starting with 2.5nM Denaturation: 1.25 nM Dilute 100 fold to 12.5 pM |
|
![]() |
![]() |
![]() |
#20 | |
Senior Member
Location: Purdue University, West Lafayette, Indiana Join Date: Aug 2008
Posts: 2,317
|
![]() Quote:
That is not as bad. But does mean using the standard protocol, that calls for 2nM starting concentration, the maximum you can cluster at is 10 pM. To cluster at above that concentration, you should start with more concentrated library. -- Phillip |
|
![]() |
![]() |
![]() |