#1 |
Junior Member
Location: Estonia Join Date: Oct 2013
Posts: 8
|
![]()
Hi,
Maybe someone has faced the same problem, as I am: I am using Agencourt AMPure XP to clean my PCR-amplicons from PCR-reaction mixture. I am using AMPure at 1x sample volume, to get rid of all the primers. In the last few months I started noticing, that the AMPure magnetic beads do not pellet properly on the tube wall after the addition of Elution Buffer (10 mM Tris-HCl, pH 8,5). The DNA concentrations of amplicons also decreased. I tried to elute my amplicons in MQ water and it worked well, I also obtained a bit higher DNA concentrations. But I am still wondering, what is wrong with the elution using EB. I checked my Ethanol and tried to use 80% instead of 70%, it did not change anything. I also tried to use different lab plastic, magnet plate and ordered fresh EB and AMPure. Nothing worked and I still got this „haze“ instead of properly pelleted beads. Maybe someone could help me, what am I doing wrong? |
![]() |
![]() |
![]() |
#2 |
Member
Location: East Cost Join Date: May 2011
Posts: 79
|
![]()
Are you using the same PCR polymerase/buffer. I know that having proteins like BSA in the sample causes haze during Ampure cleanup. I cannot explain why your problem disappears with MQ water though.
|
![]() |
![]() |
![]() |
#3 |
Junior Member
Location: Estonia Join Date: Oct 2013
Posts: 8
|
![]()
I am using Maxima HotStart PCR Master Mix. I tried to leave out ethanol purification steps (just to see, what happenes) and it worked - the haze was gone. But I can´t do that with my real samples..
|
![]() |
![]() |
![]() |
#4 | |
Member
Location: USA Join Date: Aug 2013
Posts: 52
|
![]() Quote:
|
|
![]() |
![]() |
![]() |
#5 |
Junior Member
Location: Estonia Join Date: Oct 2013
Posts: 8
|
![]()
Thanks bilyl!
I have tried different incubation times. I have also tried to elute amplicons right after I removed ethanol and also tried to let the beads dry completely, before I added the elution buffer. I did not notice any changes. |
![]() |
![]() |
![]() |
#6 |
Junior Member
Location: Milan - Italy Join Date: Aug 2014
Posts: 3
|
![]()
I' m using AMPure on a Beckman Platform for many years and I used water MQ as elution buffer without any problem. Are you using the original magnet plate?
|
![]() |
![]() |
![]() |
#7 |
Junior Member
Location: Estonia Join Date: Oct 2013
Posts: 8
|
![]()
I am using DynaMag magnet plate (Life Technologies). I am not using MQ water, because sequencing company where we send our samples prefers PCR amplicons in Tris-HCl buffer.
|
![]() |
![]() |
![]() |
#8 |
Junior Member
Location: San Francisco, Ca Join Date: Mar 2013
Posts: 9
|
![]()
I'm not certain what might be going on here, but if MQ seems to work for the elution, could you just elute in water and then dilute in elution buffer?
|
![]() |
![]() |
![]() |
#9 |
Junior Member
Location: Estonia Join Date: Oct 2013
Posts: 8
|
![]()
Thank you Dan (dtm2451)!
I will try this out. But I still would like to know, what could have happned with the purification. My last idea is, that maybe it´s the hot weather, that causes such abnormalities (air conditioner in my lab does not work properly and it is up to 30C/86F) But i refrigirate Ampure (and then bring it to room temperature immediately before using) I also tried to refrigerate ethanol and elution buffer, unfortunately it did not help. |
![]() |
![]() |
![]() |
#10 |
Junior Member
Location: Estonia Join Date: Oct 2013
Posts: 8
|
![]()
My problem seems to be gone now! During last few Ampure purifications I did not notice haze any more. I did not change anything and so I think the haze was a result of way too hot working conditions. Fortunately is summer over now!
|
![]() |
![]() |
![]() |
#11 |
Junior Member
Location: San Francisco, Ca Join Date: Mar 2013
Posts: 9
|
![]()
My lab also gets fairly hot in the Summer due to inefficient AC. Perhaps that was the issue as I'm no longer having SPRI trouble either. I don't like the heat explanation... I really don't get why that would inhibit bead binding, but it might be true. Either way, I'm glad you are no longer having issues either!
|
![]() |
![]() |
![]() |
#12 |
Junior Member
Location: Seattle WA Join Date: Aug 2014
Posts: 8
|
![]()
My Lab uses AMPure beads at 0.7x to remove primer-dimer and any junk below our target after barcoding and final pooling for MiSEQ. We elute into TE and then run the eluted DNA through a QiaQuick spin column to remove any residual beads. You can also adjust the amount of EB you spin through the column if you need a higher DNA concentration. Might be a bit redundant but hopefully it helps!
|
![]() |
![]() |
![]() |