Seqanswers Leaderboard Ad

Collapse

Announcement

Collapse
No announcement yet.
X
 
  • Filter
  • Time
  • Show
Clear All
new posts

  • Single cell RNA-seq trouble

    Hi folks,
    I'm trying to set up a single-cell RNAseq platform for human primary cells, so far semi-successfull. Hopefully you can give me some feedback on how to optimize and fix the problems.

    Here's what I do, somewhat following the Trombetta protocol http://onlinelibrary.wiley.com/doi/1...2s107/abstract :
    1) FACS sort 1 cell/well into PCR plate with 10ul TCL buffer (+bMerc)
    2) 2.2x RNA SPRI
    3) RT with Maxima RNaseH-minus RT, 90min at 52C, then 10cycles 2min-50C, 2min-42C
    4) WTA with KAPA HiFi Hotstart, 20 cycles
    5) 2x 0.8x DNA SPRI

    The good: I tested the protocol with Jurkat cells and got acceptable BioA traces for both a bulk control as well as 1cell/well samples (see attachment).

    The bad: Using primary cells (human effector T cells or regulatory T cells, with the Tregs being my goal population) it went south. It works with >200 cells but I either get nothing out of my single cells or a weird pattern of cDNA distribution with 2-3 small bumps (I wouldn't even call them peaks).

    What can I try to get it to work with 1 primary cell? And what causes this weird bumpy pattern?

    Thanks
    Alex
    Attached Files

  • #2
    Single cell RNA-seq trouble

    That would be due to significant lower RNA quantities from T cells than Jurkat cells

    Comment


    • #3
      Primary lymphocytes have a significantly lower level of mRNA than cultured cells.

      I've found that scRNA-seq is very sensitive to sample handling. Are you using fresh or frozen T cells? How long is your sort?

      Also, how many PCR cycles did you perform for each of these samples? For other lymphocyte types, I've needed to exceed 21 cycles to see a peak.

      Comment


      • #4
        I have tried increasing the PCR cycles up to 23 but still got this double-bump pattern.
        The samples were frozen, I sort directly into lysis buffer (max 5min per plate), directly spin the plate down and then freeze on dry-ice and transfer to -80C until processing. For processing, I start out with bringing the RNA beads to room temp, RNAzap the work area, let the plates thaw on ice for 10min and then start the protocol.

        Comment


        • #5
          The double peak looks like degraded RNA. So there must be something that is causing the single primary cells, but not pooled sample or single jurkats to degrade. If you can identify that factor, you can probably fix it!

          Comment


          • #6
            Originally posted by alekzs View Post
            I have tried increasing the PCR cycles up to 23 but still got this double-bump pattern.
            The samples were frozen, I sort directly into lysis buffer (max 5min per plate), directly spin the plate down and then freeze on dry-ice and transfer to -80C until processing. For processing, I start out with bringing the RNA beads to room temp, RNAzap the work area, let the plates thaw on ice for 10min and then start the protocol.
            I don´t have access to the paper you refer to but vaguely remember it.
            I have some comments and suggestions:
            - why do you sort in such a large volume of buffer. We generally sort in 2 ul (384w plates) or 4 ul (96w plates).
            - use RNAse inhibitor at all stages, even during cell isolation. Is your FACS sorter clean...really clean? Always use negative controls in your test: 1 droplet of FACS buffer in a well with lysis buffer, to check that nothing gets amplified.
            - use a dead/live marker. PI, CellTracker, whatever works (no Hoechst. it´s toxic and gives often issues)
            - sorry my silly question but does "2.2X SPRI" mean? are you doing a bead purification before RT? I wouldn´t transfer and/or remove the content of the plate until after PCR, once you have enough DNA.
            - I hope you have a mistake in the short protocol you posted and are not doing RT at 52 deg with Maxima H- but at 42 instead...
            - It´s also not necessary to do 2 rounds of bead purification after PCR. If you are concerned with primer dimers use biotinylated primers and decrease the amount. But this of course, has no influence on the quality of your RNA of course...it just saves you time.

            That said, frozen cells always have a much lower viability than fresh ones (a nice comparison here: PMID:28249587). For T cells we generally do 23-24 PCR cycles and you are doing the same, so no issue here. And yes, the "double bump" is a mix of degraded DNA (RNA) and leftover primers probably not removed after purification. Most likely you have a huge excess of unused primers after PCR.

            Best,
            Simone

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Essential Discoveries and Tools in Epitranscriptomics
              by seqadmin




              The field of epigenetics has traditionally concentrated more on DNA and how changes like methylation and phosphorylation of histones impact gene expression and regulation. However, our increased understanding of RNA modifications and their importance in cellular processes has led to a rise in epitranscriptomics research. “Epitranscriptomics brings together the concepts of epigenetics and gene expression,” explained Adrien Leger, PhD, Principal Research Scientist...
              04-22-2024, 07:01 AM
            • seqadmin
              Current Approaches to Protein Sequencing
              by seqadmin


              Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
              04-04-2024, 04:25 PM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, Yesterday, 11:49 AM
            0 responses
            15 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-24-2024, 08:47 AM
            0 responses
            16 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-11-2024, 12:08 PM
            0 responses
            61 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 10:19 PM
            0 responses
            60 views
            0 likes
            Last Post seqadmin  
            Working...
            X