  • MiSeq: Sudden drop in %> Q30 halfway read 1

    We have a run going on the MiSeq as we speak (2x250 bp). I just had a look at the run in the SAV.

    A sudden drop in Q30 after about 75 cycles caught my attention. Has anyone seen anything like this?

    *Pic attached*
    Attached Files

  • #2
    How are you preparing your library? We do a lot of amplicon sequencing, and sharp drops in quality in the 75-150bp region typically indicate some level of PCR artifact (primer-dimer) in the library. Since the artifacts have adapters, they load and sequence, but once the short fragment is done being sequenced (about 75-150 cycles into the run) the quality score for those clusters drops sharply.

    Short fragments tend to load preferentially, so even a relatively small amount of artifact can have a pretty large effect on the run.

    Comment


    • #3
      Thanks Bryanbriney.

      These were whole-genome frgmented DNA libraries prepared using Kapa + bioo scientific indexes. You are probabbly spot on. There were some adapter-dimers and adapters visible on the bioanalyzer trace. Priot to this we only used gels, so we were unsure whether these adapters were something we had had also in previous runs but failed to spot on the gels. In that case, as previous runs have gone well, we decided to sequence and see what happens. Probably less than perfect AMPure clean-up after adapter ligation and PCR then....

      I guess doing another clean-up of the samples pooled after PCR could be a good option, and then start another run...

      Comment


      • #4
        Check out this thread:
        Bridged amplification & clustering followed by sequencing by synthesis. (Genome Analyzer / HiSeq / MiSeq)


        Illumina has acknowledged an issue with their MiSeq 600 cycle v3 kits, though it has not been resolved yet.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM
        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        17 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        22 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        16 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        46 views
        0 likes
        Last Post seqadmin  
        Working...
        X