Hi All,
I'm having some issues running a custom small RNA sequencing protocol on the HiSeq4000, when the same library works perfectly fine on the 2500. Illumina may not have tested the TruSeq Small RNA protocol using the 4000, so I was wondering if anyone in the community has successfully done sequenced small RNAs on the 4000. Thanks! |
![]() |
![]() |
![]() |
#2 |
Senior Member
Location: Stanford Join Date: Jun 2009
Posts: 179
|
![]()
I can't help you, but can you tell us what issues you had and whether you'd already tried the same libraries on a different machine?
|
![]() |
![]() |
![]() |
#3 |
Junior Member
Location: San Francisco Join Date: Nov 2015
Posts: 2
|
![]()
Should've been more specific, yes the libraries are simply not clustering! and they do cluster on the 2500. But I'm just wondering if it is due to the minor differences (2-3nt shorter) in the clustering arms or whether the size of the whole molecule is important in the 4000 due to the differences in clustering.
|
![]() |
![]() |
![]() |
#4 |
Junior Member
Location: Switzerland Join Date: Apr 2012
Posts: 1
|
![]()
Hope you have solved your problem already. If not....small RNA will not cluster on 4000 as the fragment size need to be atleast 300bp to cluster on 4000.
|
![]() |
![]() |
![]() |
#5 |
Senior Member
Location: US Join Date: Dec 2010
Posts: 422
|
![]()
Even adapter dimers (insert size 0) cluster and sequence just fine on the Hiseq 4000 - there is no problem with the insert length. Small RNA libraries can work. The OP problems might relate to the "custom" part.
|
![]() |
![]() |
![]() |
#6 | |
Senior Member
Location: USA, Midwest Join Date: May 2008
Posts: 1,172
|
![]() Quote:
TruSeq Small RNA libraries require different sequencing and indexing primers than other TruSeq libraries (e.g. DNA, mRNA, etc.). The HiSeq2500 clustering and sequencing reagent kits (both Rapid and High Output) include all required primers for all TruSeq library types as well as Nextera libraries. While Illumina has not chosen to share with me personally I would bet my favorite P20 Pipetman* that the TruSeq Small RNA primers are not included in the HiSeq4000 reagent kits. If true then your observation that the libraries worked on the 2500 but not the 4000 is entirely expected. Always check kit & platform compatibility before loading anything on a sequencer. *Since I rarely if ever do actual bench work anymore its not much of bet really. |
|
![]() |
![]() |
![]() |
#7 |
Senior Member
Location: US Join Date: Dec 2010
Posts: 422
|
![]()
If the Illumina kit indeed does not work on the HS 4000 - this one does: http://www.biooscientific.com/Next-G...rary-Prep-Kits
|
![]() |
![]() |
![]() |
#8 | |
Senior Member
Location: USA, Midwest Join Date: May 2008
Posts: 1,172
|
![]() Quote:
The Bioo NEXTflex Small RNA Kits produce libraries with the same adapter sequences as the TruSeq Small RNA Kit. If you have successfully sequenced Bioo NEXTflex small RNA libraries on a HiSeq 4000, using the stock clustering/sequencing reagent kits, then I'll gladly retract my earlier supposition that the 4000 kits lack the Small RNA primers. |
|
![]() |
![]() |
![]() |
#9 |
Senior Member
Location: Bioo Scientific, Austin, TX, USA Join Date: Jun 2012
Posts: 119
|
![]()
I contacted a collaborator today who has been using Bioo Scientific small RNA kits and just had samples sequenced on a HiSeq 4000, and although they haven't analyzed the data in detail yet everything appears normal on first pass. I will update if they run into any problems. As kmcarr mentioned, the adapter sequences in the Bioo kits are identical to TruSeq (except Bioo's contain random bases to reduce ligase bias), so since the Bioo kit worked then TruSeq should as well.
As a disclaimer, I work for Bioo Scientific. |
![]() |
![]() |
![]() |
#10 |
Junior Member
Location: Ann Arbor Join Date: Jun 2015
Posts: 4
|
![]()
Custom primers don't work on the HiSeq 4000. Were you trying to use custom primers?
|
![]() |
![]() |
![]() |
#11 |
Junior Member
Location: San Francisco Join Date: Apr 2012
Posts: 9
|
![]()
Our TruSeq Small RNA library works very well on HiSeq 4000. It is not listed as compatible with HiSeq4000 on Illumina website because the protocol is prone to produce adapter dimer. Check bioanalyzer and make sure there is no even trace of adapter dimer.
|
![]() |
![]() |
![]() |