Hi,
I have a sequencing project involving tiling of multiple amplicons across a genome. Each is sized approx 1-1.5 kb. Is it possible to sequence these amplicons directly using 454 titanium? I've been told that 1 kb is too small for fragmentation via nebulization, and too long for amplification in emPCR. Is my best bet to redesign the amplicons to ~500 bp and add the fusion primers to my target-specific primers? Or find restriction enzymes that will make my fragments smaller? Or is there a way to salvage the 1 kb amplicons directly? The region I'm targeting is highly variable so it's difficult to get perfectly spaced primer pairs.

Last edited by greigite; 07-30-2009 at 05:08 PM.
greigite is offline   Reply With Quote
Old 07-30-2009, 06:10 PM   #2
kmcarr
Senior Member
 
Location: USA, Midwest

Join Date: May 2008
Posts: 1,178
Default

The amplicon sequencing kit for Titanium is not released yet. It is supposed to be released this Fall. I don't have any concrete information so take what I say with a grain of salt. Assuming amplicon read lengths on Titanium will be similar to the shotgun reads (400-450nt) then I would think the maximum amplicon size you would want to generate is 700nt (including the template specific portions of the fusion primers). This way you could sequence each amplicon product using the II and III emulsion kits (sequencing from each end of the amplicon) with enough overlap to make certain you adequately cover the entire amplicon.
kmcarr is offline   Reply With Quote
Old 08-04-2009, 08:03 PM   #3
greigite
Senior Member
 
Location: Cambridge, MA

Join Date: Mar 2009
Posts: 141
Default

Thanks, kmcarr. I'm aware that there is no amplicon sequencing kit for Titanium yet, so I was planning on fragmenting the amplicons and using multiplex adapters. Do you know if nebulization works well on these smaller fragment sizes? I'm planning to try it myself, but am curious if anyone else has had success. Unfortunately we do not have a Covaris or Bioruptor.
greigite is offline   Reply With Quote
Old 08-05-2009, 07:51 PM   #4
AKroxy
Member
 
Location: Alaska

Join Date: May 2009
Posts: 20
Default

Are the A and B adaptor sequences the same for titanium and FLX? If so, why would a specific titanium kit be needed?
AKroxy is offline   Reply With Quote
Old 08-07-2009, 03:21 AM   #5
flxlex
Moderator
 
Location: Oslo, Norway

Join Date: Nov 2008
Posts: 415
Default

Latest we heard at the User Meeting in Lisbon is that the software seems to have problems handling amplicons sequenced using Titanium chemistry. We tried once and got suboptimal results. My advice is to wait for the official release.

BTW, A and B adapter sequences for Titanium are NOT the same as for FLX...
flxlex is offline   Reply With Quote
Old 08-19-2009, 12:52 PM   #6
454Sequencing
Member
 
Location: Branford, CT

Join Date: Apr 2009
Posts: 22
Default

Hi everyone,

Allow me to clear a few things up. 454 Life Sciences plans to release methods, reagents, and software for GS FLX Titanium series amplicon sequencing in the fall. Please stay tuned for a Technical Bulletin on the topic that will be available shortly. Included within will be amplicon design guidelines including the new fusion primer sequences, length considerations, and other related details.

Addressing flxlex's question: Rather than a problem with the existing software, it's simply that much as with Standard series a different analysis pipeline is required for Titanium series amplicon sequencing. Using the "normal" i.e. shotgun/paired end signal processing is not recommended for amplicons and may lead to poor results.

In the interim I would recommend following up with your local Roche representative if you have any questions.

Best regards,

Jason
Product Manager, 454 Life Sciences
454Sequencing is offline   Reply With Quote
Old 08-26-2009, 07:32 AM   #7
joa_ds
Member
 
Location: belgium

Join Date: Dec 2008
Posts: 52
Default

hehe, we also tried amplicon sequencing on the Titanium. Poor results. Only 200k reads with poor quality. So should wait indeed for the titanium package.
joa_ds is offline   Reply With Quote
Old 09-02-2009, 03:29 PM   #8
Hamid
Senior Member
 
Location: San Diego, CA

Join Date: Sep 2009
Posts: 105
Default

Hi Greigite,

You can indeed use the Covaris instrument to shear PCR products to smaller fragments. We have good experience with shearing PCR products down to 300-400bp for use library preparation. Feel free to contact me for the settings.

hamid
Hamid is offline   Reply With Quote
Old 09-24-2009, 10:43 AM   #9
kmcarr
Senior Member
 
Location: USA, Midwest

Join Date: May 2008
Posts: 1,178
Default

Quote:
Originally Posted by flxlex View Post
Latest we heard at the User Meeting in Lisbon is that the software seems to have problems handling amplicons sequenced using Titanium chemistry. We tried once and got suboptimal results. My advice is to wait for the official release.

BTW, A and B adapter sequences for Titanium are NOT the same as for FLX...
There's new information on this. Our FAS just sent us a Technical Bulletin which Roche/454 is distributing ahead of the (soon to be released) Titanium Amplicon kits. For amplicons the A and B adapters are different but only in the respect that they are extended at their 5' ends:
Code:
Titanium Fusion Primer-A: 5'-CGTATCGCCTCCCTCGCGCCATCAG-{MID}-{template-specific-seq}-3'
FLX Fusion Primer-A:            5'-GCCTCCCTCGCGCCATCAG-{MID}-{template-specific-seq}-3'

Titanium Fusion Primer-B: 5'-CTATGCGCCTTGCCAGCCCGCTCAG-{MID}-{template-specific-seq}-3'
FLX Fusion Primer-B:            5'-GCCTTGCCAGCCCGCTCAG-{MID}-{template-specific-seq}-3'
The Bulletin states that the new Amplicon kits and protocols will be backward compatible with amplicons created using the standard FLX adapters. However it strongly recommends that you redesign your amplicon primers as the older FLX ones, while they will work, perform sub-optimally producing fewer reads.

Recommended total amplicon size is 200-600bp (including the fusion primers). Products up to 800 bp are possible but you may encounter problems.
kmcarr is offline   Reply With Quote
Old 12-28-2009, 08:28 AM   #10
jffkm
Junior Member
 
Location: new york

Join Date: Oct 2009
Posts: 7
Default 454 Titanium Amplicon sequencing

I was wondering why there is a limitation to the size of the amplicons that can be sequenced. I know that larger (>600bp) amplicons will not get complete sequencing coverage from a one-sided read, but apart from that, I don't understand why there would be a physical reduction in the quality of the reads as emulsion PCR can be routinely done on amplicons exceeding 3kb (nature protocols, etc.). The extension times used in the emPCR procedure also provide more than sufficient rates for covering larger amplicons as well. Any input would be appreciated. thanks!
jffkm is offline   Reply With Quote
Old 01-05-2010, 06:43 AM   #11
sulfobus
Member
 
Location: Denmark

Join Date: Oct 2009
Posts: 12
Default

Is this kit for amplicon sequencing with titanium released yet?
sulfobus is offline   Reply With Quote
Old 01-05-2010, 06:52 AM   #12
454Sequencing
Member
 
Location: Branford, CT

Join Date: Apr 2009
Posts: 22
Default

Dear sulfobus,

Yes, the emPCR kits that support GS FLX Titanium series sequencing of amplicons are available. You can find information on current products at our website, specifically at http://www.454.com/products-solutions/product-list.asp. In this case it is the Lib-A series of emPCR kits that you would want. Additional details are provided at http://www.454.com/products-solution...sequencing.asp and your local Roche representative will be able to supply further guidance.

If you are a current customer I would recommend the my454 customer portal (http://www.454.com/my454). Therein you will find documentation including current user manuals as well as other useful information.

Best regards,
Jason

Quote:
Originally Posted by sulfobus View Post
Is this kit for amplicon sequencing with titanium released yet?
__________________
Technical Product Manager
454 Life Sciences, A Roche Company
454Sequencing is offline   Reply With Quote
Old 01-15-2010, 09:26 AM   #13
sudsf16
Junior Member
 
Location: Columbia, SC

Join Date: Jun 2008
Posts: 4
Default

All but one of our clients want unidirectional amplicon sequencing which we have been providing with the standard FLX chemistry. The new Titanium amplicon kits are bidirectional only. If you need unidirectional sequencing, Kit II or Kit III using FLX chemistry is what's available from Roche.
sudsf16 is offline   Reply With Quote
Old 01-15-2010, 12:00 PM   #14
kmcarr
Senior Member
 
Location: USA, Midwest

Join Date: May 2008
Posts: 1,178
Default

Quote:
Originally Posted by sudsf16 View Post
All but one of our clients want unidirectional amplicon sequencing which we have been providing with the standard FLX chemistry. The new Titanium amplicon kits are bidirectional only. If you need unidirectional sequencing, Kit II or Kit III using FLX chemistry is what's available from Roche.
sudsf,

Jason (454Sequencing) addressed that issue in another thread. There will be enough of each reagent to do a unidirectional run.

Additional info:

O.K. I just went looking for Application Brief which Jason mentioned in that thread. It is posted on the 454 web site and is called APP001-2009_Lib-LunideirectionalAmplicons.pdf. Basically the strategy is to design your amplicon primers using the normal shotgun A & B adapter sequences and then do the emPCR with the Lib-L kit instead of the Lib-A. I believe that this method would not be backward compatible with amplicons originally designed for the standard FLX protocol. The Lib-A emPCR kit is backward compatible.

Last edited by kmcarr; 01-15-2010 at 12:12 PM. Reason: Add information about alternate method for unidirectional sequencing.
kmcarr is offline   Reply With Quote
Old 03-25-2011, 11:54 AM   #15
laurahug
Junior Member
 
Location: Toronto, ON

Join Date: Mar 2011
Posts: 1
Default

Hi all,

I have the same question as greigite, two years later: has anything changed? Can I sequence a 1-1.5kb PCR amplicon with any of the currently available 454 technologies? I'm working with a gene family as well, and require the amplicon length to be that long to avoid hypervariable regions to which primer design is near-impossible.

I don't, however, need a full over-lapping sequence from the analysis, it is more like pyrotagging from an analysis standpoint.

So has emPCR overcome the ~800 bp limit? Do LV or MV kits make a difference in that respect?

Thanks for any information, I've looked everywhere.

Laura
laurahug is offline   Reply With Quote
Old 03-25-2011, 12:28 PM   #16
kmcarr
Senior Member
 
Location: USA, Midwest

Join Date: May 2008
Posts: 1,178
Default

Laura,

No nothing has changed. In fact Roche just release a Tech Note showing how poor longer amplicons work in the current system.
kmcarr is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 01:45 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2021, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO