  • Strange higher size amplicon in ChIP seq Library!!

    Hi, I am seeing a strange amplicon ~2Kb long in all my libraries beside expected ~250-300bp amplicon (picture attached).

    I perform ChIP using DNA sheared with bioruptor which peaks ~500bp. I work with low cell number so I get ~5-10 ng after ChIP. Post IP this DNA is further sheared using Covaris and processed for library preparation as per the Solid ChIP-seq library preparation protocol. I used to get ~250-300 bp DNA as expected but in all my recent samples I could see a higher size amplicon ~2 KB beside the expected amplicon. I am not able to find any logical explanation for this!! Has anyone seen this before?

    Can I still use these libraries for sequencing after size selection? I am afraid I might introduce bias in my samples. Please help.
    Attached Files
    Last edited by Chiper; 06-22-2010, 01:12 AM.

  • #2
    Your loading dye is tricking you.

    The higher molecular weight DNA you see is just the top of a smear of concatamers of your monomer amplicons. The loading dye is blocking signal from the middle section of the smear. Thus the very top of the smear appears to be a band.

    Over amplification of your amplicons is said to be a source of such a smear.

    --
    Phillip

    Comment


    • #3
      Dear Phillip,
      Thanks for your reply. The band I see is not a smear since I also loaded samples on Bio Analyzer High sensitivity chip and in almost all the samples it is a band around 1.5-2KB region with no DNA in between. In one sample I also saw another faint band around 500 bp.
      I amplified the libraries from 12-14 cycles and given the starting amount I had I don't see much on Lonza gel if I amplify less than this. Do you think that concatamers of monomer amplicons can be a source of such band?

      Comment


      • #4
        Originally posted by Chiper View Post
        Dear Phillip,
        Thanks for your reply. The band I see is not a smear since I also loaded samples on Bio Analyzer High sensitivity chip and in almost all the samples it is a band around 1.5-2KB region with no DNA in between. In one sample I also saw another faint band around 500 bp.
        Ah, well then your loading dye tricked me. I saw the loading dye and just assumed there would be DNA it was hiding.

        Originally posted by Chiper View Post
        I amplified the libraries from 12-14 cycles and given the starting amount I had I don't see much on Lonza gel if I amplify less than this. Do you think that concatamers of monomer amplicons can be a source of such band?
        Well it does not take much DNA -- for each ng there are 5 billion 200 bp amplicons. So it is better not to over-amplify.

        Can you post the high sensitivity chip picture? That would help me decide whether the 2 kb band is a concatamer or something else.

        --
        Phillip

        Comment


        • #5
          Dear Philip,
          Sorry for the late reply. Please find the pictures attached.
          Attached Files

          Comment


          • #6
            Yes that looks like "overamplification", even though you only did 10-14 cycles. I don't know what causes it, but I remember pictures of the phenomenon from the v1 manual we used during SOLiD fragment library construction training in Foster City.

            --
            Phillip

            Comment

            Latest Articles

            Collapse

            • seqadmin
              Current Approaches to Protein Sequencing
              by seqadmin


              Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
              04-04-2024, 04:25 PM
            • seqadmin
              Strategies for Sequencing Challenging Samples
              by seqadmin


              Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
              03-22-2024, 06:39 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by seqadmin, 04-11-2024, 12:08 PM
            0 responses
            25 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 10:19 PM
            0 responses
            28 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-10-2024, 09:21 AM
            0 responses
            24 views
            0 likes
            Last Post seqadmin  
            Started by seqadmin, 04-04-2024, 09:00 AM
            0 responses
            52 views
            0 likes
            Last Post seqadmin  
            Working...
            X