  • bioanalyzer and qubit disagreement

    We prepare clinical samples with a formal SOP and do not regularly see irregularities. Today we have a capture pool of samples that quanted at 24.0ng/uL on qubit. To determine fragment size we also run sample pool in duplicate on bioanalyzer DNA1000. Duplicate 1, there is noise around 40bp (<1ng/uL) and no peak in the expected range (300-400). Duplicate 2, there is noise around 40bp and a wide peak between 700 and 1500 (upper marker). We reran the sample pool (4 replicates) on the bioanalyzer with only the noise around 40bp.

    Qubit concentration = normal
    bioanalyzer fragment size = missing?

    Any ideas what is going on? Where is the missing peak?
    Last edited by smsestes; 02-05-2020, 02:47 PM. Reason: clarity

  • #2
    Perhaps you didn't load enough on the Bioanalyzer to get a signal from a smear using a D1000 chip? Try to run on HS DNA chip.

    Comment


    • #3
      DNA 1000 Kit is for 5-50 ng/ul of DNA
      Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

      I think that in HS dna chip will be overloaded...

      Check if everything is ok with gel dye mix in DNA 1000 Kit.

      Comment


      • #4
        Originally posted by Noveline View Post
        DNA 1000 Kit is for 5-50 ng/ul of DNA
        Hs DNA Kit: "NGS sheared DNA or libraries: For accurate determination of DNA concentration, the total DNA in the samples must be between 100pg/μL to 10 ng/μL."

        I think that in HS dna chip will be overloaded...

        Check if everything is ok with gel dye mix in DNA 1000 Kit.
        Hi Noveline,

        When we reran the sample pool, we included a previous sample pool which ran normally and as expected. Thus it is unlikely to be the BioA or its reagents.

        Comment

        Latest Articles

        Collapse

        • seqadmin
          Current Approaches to Protein Sequencing
          by seqadmin


          Proteins are often described as the workhorses of the cell, and identifying their sequences is key to understanding their role in biological processes and disease. Currently, the most common technique used to determine protein sequences is mass spectrometry. While still a valuable tool, mass spectrometry faces several limitations and requires a highly experienced scientist familiar with the equipment to operate it. Additionally, other proteomic methods, like affinity assays, are constrained...
          04-04-2024, 04:25 PM
        • seqadmin
          Strategies for Sequencing Challenging Samples
          by seqadmin


          Despite advancements in sequencing platforms and related sample preparation technologies, certain sample types continue to present significant challenges that can compromise sequencing results. Pedro Echave, Senior Manager of the Global Business Segment at Revvity, explained that the success of a sequencing experiment ultimately depends on the amount and integrity of the nucleic acid template (RNA or DNA) obtained from a sample. “The better the quality of the nucleic acid isolated...
          03-22-2024, 06:39 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by seqadmin, 04-11-2024, 12:08 PM
        0 responses
        18 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 10:19 PM
        0 responses
        22 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-10-2024, 09:21 AM
        0 responses
        16 views
        0 likes
        Last Post seqadmin  
        Started by seqadmin, 04-04-2024, 09:00 AM
        0 responses
        46 views
        0 likes
        Last Post seqadmin  
        Working...
        X