Default Custom primers on a MiniSeq?

We sent a custom read 1 seq primer that works fine on our NextSeq and MiSeq to two groups with MiniSeqs, and the result is a total failure to find clusters for both. Control runs (Phi X and priming with standard seq primer) confirm the libraries and instruments are working fine, and Illumina TS says annealing temperatures are the same on NextSeq and MiniSeq.

What else could be going wrong here? Has anyone successfully used a custom primer on a MiniSeq?
jfostel is offline   Reply With Quote
Old 07-20-2016, 11:36 AM   #2
jfostel
Junior Member
 
Location: Boston, MA

Join Date: Aug 2010
Posts: 7
Default

Just an update - the Custom Primer Guide for MiniSeq contained an error. These seem to be the corrected positions:

Custom Read 1 primer: Position 15
Custom Read 2 primer: Position 14
Custom Index 1 primer: Position 13
Custom Index 2 primer: Position 13
jfostel is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 06:53 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO