Old 07-30-2016, 01:15 AM   #1
Seqbob
Junior Member
 
Location: Uk

Join Date: May 2013
Posts: 2
Default low miseq clustering after kapa library quantification

Hi,
Sorry if duplicating a thread.
I have performed 2 runs now with nebnext libraries (approx 350bp in size).
Both times I used kapa library quantity kits to quantify in triplicate over 2 dilutions (1000 & , 10000). Then I diluted multiplexed libs to total concentration of 4nm before following illumina protocol for library dilution to 15pm and loading on miseq.
On both runs the cluttering ws low (550 and 197 pf). I checked things like naoh concentration using ph paper and that was ok.
5% phi x was on both runs but miseq calculates final alignment of phi x as 20 and 25% respectively on each run suggesting my concentration of libraries was much lower than team 15pm I thought I loaded.
I had small amount of adaptor dimers in one library but surely this would not cause low clustering?
Could anyone give me suggestions on what I am doing wrong.
Seqbob is offline   Reply With Quote
Old 07-30-2016, 03:44 AM   #2
nucacidhunter
Jafar Jabbari
 
Location: Melbourne

Join Date: Jan 2013
Posts: 1,174
Default

Dimers cluster better than actual library. Higher ratio of PhiX indicates that it is either library itself or the quantification or both .
nucacidhunter is offline   Reply With Quote
Old 07-30-2016, 10:29 PM   #3
jdk787
josh kinman
 
Location: Austin

Join Date: Apr 2014
Posts: 61
Default

Since your PhiX read percentage is 4 to 5X higher than expected, I would suggest confirming your qPCR result with Qubit to make sure that something isn't wrong there.
jdk787 is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 03:20 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2018, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO