#1 |
Senior Member
Location: Oxford, Ohio Join Date: Mar 2012
Posts: 253
|
![]()
Hi,
I'm having endless issues updating/patching the Windows7 Embedded system on the MiSeq. Constantly getting errors from Windows Update about it not being able to update, and of course, being M$, the error codes are next to useless. Is there anyway around this? I just don't like the idea of an un-patched system on the Internet running M$ Windows. Regards, Andor |
![]() |
![]() |
![]() |
#2 |
Senior Member
Location: USA, Midwest Join Date: May 2008
Posts: 1,178
|
![]()
In my opinion it is unwise to install software patches which have not been tested for compatibility by Illumina. We do not patch any instrument computers with anything other than patches/upgrades provided by the instrument vendor.
Your MiSeq and all networked laboratory computers should be protected by a firewall with very stringent access rules. |
![]() |
![]() |
![]() |
#3 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,080
|
![]()
You should leave your MiSeq as is. Isolating instrument computers from open internet with a strong firewall (preferably via a defined/separate VLAN) is the best option for securing instrument computers.
|
![]() |
![]() |
![]() |
#4 |
Senior Member
Location: Oxford, Ohio Join Date: Mar 2012
Posts: 253
|
![]()
Okay, that works in a "closed" facility. What happens if we have an "open" facility and any user can insert a USB drive? Do I just cross my fingers?
I guess I'd like to know how to prevent usb spawned malware. |
![]() |
![]() |
![]() |
#5 |
Member
Location: Livermore, CA Join Date: Jan 2016
Posts: 44
|
![]()
You could open it up and remove the internal USB cable that passes through to the motherboard for the USB port on the right hand side of the instrument. I forget how the USB port on the back of the MiSeq is plumbed, but that one might be part of the motherboard (and not a cable).
|
![]() |
![]() |
![]() |
#6 |
Senior Member
Location: East Coast USA Join Date: Feb 2008
Posts: 7,080
|
![]()
@misterc: Problem is a lot of times users (or even the facility techs) may create samplesheets using IEM elsewhere and then bring those to the machine at the time of run-start. People may also want to off-load data using USB external drives (if networking is not available).
@cement_head: User education generally works well. You should put up signs (if users can walk in any hour of the day) about using good USB hygiene and absolute need to ensure a clean environment. If you can't trust certain users to follow this basic protocol then you would need to be a "bad cop" and make sure they don't access the sequencer unattended or first scan their USB's on a separate machine before they are allowed to plug those into sequencer. |
![]() |
![]() |
![]() |
#7 |
Junior Member
Location: Atlanta, GA Join Date: Jan 2019
Posts: 3
|
![]()
770-490-3265 we have one that is being retired. A illumina MiSeq DOM 2013.
|
![]() |
![]() |
![]() |
#8 | |
Senior Member
Location: Oxford, Ohio Join Date: Mar 2012
Posts: 253
|
![]() Quote:
|
|
![]() |
![]() |
![]() |