Old 06-22-2011, 09:31 PM   #1
dobbr493
Junior Member
 
Location: New Zealand

Join Date: Jun 2011
Posts: 5
Default mRNA on Bioanalyser

Hi there

We are looking at doing a transcriptome experiment with very limited starting material (usually around 10ug total RNA of good quality RIN > 8). We have tried both the Roche magnetic bead kit and Ambion spin column kit for extracting mRNA from total RNA and are unsure as to what the mRNA should look like on the Agilent Bioanalyser. An example of what we have been seeing is attached - is this mRNA or just rubbish?

Cheers
Attached Files
File Type: pdf 2100 expert_mRNA Nano_DE04103720_2011-06-20_12-31-53.pdf (859.3 KB, 179 views)
dobbr493 is offline   Reply With Quote
Old 06-22-2011, 11:17 PM   #2
RCJK
Senior Member
 
Location: Australia

Join Date: May 2009
Posts: 155
Default

I'd say that's just rubbish. What does the ladder lane look like? Did you run any blank lanes to see if the noisy baseline flattens out?
RCJK is offline   Reply With Quote
Old 06-22-2011, 11:18 PM   #3
RCJK
Senior Member
 
Location: Australia

Join Date: May 2009
Posts: 155
Default

Also, after those kits, are your samples in a buffer that is compatible with the Bioanalyser assay you are running?
RCJK is offline   Reply With Quote
Old 06-23-2011, 09:50 AM   #4
mnkyboy
Member
 
Location: Seattle, WA

Join Date: Mar 2009
Posts: 87
Default

mRNA should look like an evenly distributed mound without the ribosomal peaks (I cant find an image easily). What you are showing is noise and either you concentration is too low (or there is nothing there) and/or your buffer is incompatible. You can see that type of result if your salt concentration is too high.
mnkyboy is offline   Reply With Quote
Old 07-07-2011, 05:13 PM   #5
dobbr493
Junior Member
 
Location: New Zealand

Join Date: Jun 2011
Posts: 5
Default

I was wondering what kind of quantities of total RNA have people used for extracting mRNA from? At our lower limits, we have 5ug, max usually 10ug. Is this a common amount to be using with success?

Also, having had trouble with the purification of mRNA we are now looking at amplifying it to get the quantity we need. Has anyone done this with success from the quantities of total RNA above? Any comment on the potential bias introduced with the PCR and selection steps in terms of NGS results would be appreciated.

Thanks!
dobbr493 is offline   Reply With Quote
Old 07-08-2011, 08:12 AM   #6
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Yes that is just background noise. Your nano chip is not sensitive enough to detect your mRNA. You could run your mRNA undiluted on a RNA Pico chip that might pick something up. Or you could check fluorometrically (with a single stranded fluor) to get an idea of what your mRNA concentration is.

Maybe you have less total RNA to start with than you think? For the Nano chip you ran on the total RNA, what was the chip estimate of the concentration of your total RNA concentration? One common issue with estimating total RNA amounts is that even tiny amounts (1 ul per ml) of phenol will give you very strong readings at 260 nm, even in the absence of any nucleic acids.

--
Phillip
pmiguel is offline   Reply With Quote
Old 07-08-2011, 08:15 AM   #7
shurjo
Senior Member
 
Location: Rockville, MD

Join Date: Jan 2009
Posts: 126
Default

I'm attaching a picture from poly(A)-selected mRNA (from 10ug of total RNA) run on a Pico Chip. As you can see, the ribosomal molecules are still noticeably present, but then those are very hard to remove completely.
Attached Images
File Type: jpg IMAG0083.jpg (18.5 KB, 83 views)
shurjo is offline   Reply With Quote
Old 07-08-2011, 08:51 AM   #8
sisch
Member
 
Location: Dusseldorf, Germany

Join Date: Jun 2011
Posts: 29
Default

Quote:
Originally Posted by mnkyboy View Post
mRNA should look like an evenly distributed mound without the ribosomal peaks (I cant find an image easily). What you are showing is noise and either you concentration is too low (or there is nothing there) and/or your buffer is incompatible. You can see that type of result if your salt concentration is too high.
Find attached one of our result files. To the left is totalRNA and to the right is purified mRNA with 1&2 being Qiagen Oligotex Kit purified (Pain in the ass imho) 3&4 Invitrogen Dynabeads.
So that's what we believe mRNA should look like on a Bioanalyzer (up to 8% of residual rRNA rendered no problem in our experiments)

By the way, we usually start off with 30 - 70ug totalRNA and have 0.5ug mRNA afterwards. The limit for sequencing w/o amplification is to my knowledge about 200ng mRNA.

Cheers,
Simon
Attached Images
File Type: jpg bioanalyzer.JPG (25.7 KB, 82 views)

Last edited by sisch; 07-08-2011 at 08:53 AM.
sisch is offline   Reply With Quote
Old 07-10-2011, 05:19 PM   #9
dobbr493
Junior Member
 
Location: New Zealand

Join Date: Jun 2011
Posts: 5
Default

Quote:
Originally Posted by shurjo View Post
I'm attaching a picture from poly(A)-selected mRNA (from 10ug of total RNA) run on a Pico Chip. As you can see, the ribosomal molecules are still noticeably present, but then those are very hard to remove completely.
That's very interesting - what was the kit that you used with that amount of total RNA? Also, what was the end-quantity of mRNA? Thanks for your help!
dobbr493 is offline   Reply With Quote
Old 07-11-2011, 09:03 AM   #10
shurjo
Senior Member
 
Location: Rockville, MD

Join Date: Jan 2009
Posts: 126
Default

Quote:
Originally Posted by dobbr493 View Post
That's very interesting - what was the kit that you used with that amount of total RNA? Also, what was the end-quantity of mRNA? Thanks for your help!
I used the DynaBeads kit from Invitrogen with 2 rounds of purification. mRNA yield from 10ug of total RNA with RIN=10 was about 120ng.

HTH,

Shurjo
shurjo is offline   Reply With Quote
Old 04-25-2012, 08:09 AM   #11
Ecap
Junior Member
 
Location: Maryland

Join Date: Apr 2012
Posts: 2
Default

Hello,

I've observed bioanalyzer signatures very similar to yours when I've analyzed post-poly(A) selected total RNA samples in high salt and/or incompatible buffers.

For me, ethanol precipitation followed by sample reconstitution in a clean buffer (or water) totally solved the problem.
Ecap is offline   Reply With Quote
Old 04-25-2012, 08:09 AM   #12
Ecap
Junior Member
 
Location: Maryland

Join Date: Apr 2012
Posts: 2
Default

Quote:
Originally Posted by Ecap View Post
Hello,

I've observed bioanalyzer signatures very similar to yours when I've analyzed post-poly(A) selected total RNA samples in high salt and/or incompatible buffers.

For me, ethanol precipitation followed by sample reconstitution in a clean buffer (or water) totally solved the problem.
I forgot to mention that in this case I was also using samples whose concentration were equal to or lesser than yours
Ecap is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 11:51 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO