Old 10-21-2011, 08:36 AM   #1
NuGEN
Registered Vendor
 
Location: bay area

Join Date: Jun 2010
Posts: 22
Default New! Encore Complete RNA-Seq Library Systems from NuGEN Technologies

The Encore Complete RNA-Seq Library Systems offer a number of advantages for researchers engaged in RNA-Seq studies:

An end-to-end solution for strand-specific RNA-Seq – All required components for low input preparation of strand-specific RNA-Seq libraries from moderate amounts of total RNA without the need for rRNA reduction or poly(A) selection
Superior strand retention – Using a range of sources for total RNA researchers can expect to achieve >95% read mapping in the sense orientation for accurate and complete understanding of transcript profiling results
Affordable and scalable kit configurations – The barcoding capability provided with Encore Complete RNA-Seq Multiplex Systems I and IB improve sample throughput and reduce sequencing costs as your projects grow

http://www.nugeninc.com/nugen/index....brary-systems/

Best,
Steve

Last edited by NuGEN; 10-21-2011 at 11:57 AM. Reason: Adding link to product web page
NuGEN is offline   Reply With Quote
Old 02-13-2012, 06:00 AM   #2
seq2no
Junior Member
 
Location: Philadelphia

Join Date: Oct 2011
Posts: 1
Default

Hi Steve,

We recently sequenced Nugen ENCORE complete RNA seq library samples on our HiSeq 2000. The q30 scores were not great (between 70-80% - we normally see q30s in the 90s). The phasing was high on some of the lanes and the prephasing was high on other lanes. We also happened to run 3 Illumina prepped samples that had normal q30s and normal phasing/prephasing. Have you heard of issues like this before? Do you know why the NuGen prepped samples would behave this way?

Thanks,
Jeanne

Last edited by seq2no; 02-13-2012 at 06:01 AM. Reason: spelling error
seq2no is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 05:33 AM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO