Old 11-24-2013, 08:22 PM   #1
AndyG
Junior Member
 
Location: Australia

Join Date: Jun 2013
Posts: 4
Default Twin peaks, double peaks, double humps of RNA Seq, it’s all very frustrating...HELP!

Method:

Wang et al. 2011, A low-cost library construction protocol for Illumina-based strand-specific multiplex rna-seq. PLoS One 6(10) e26426. doi:10.1371/journal.pone.0026426

Basics steps:

Extract Total RNA
Extract mRNA
Fragment
1st strand synthesis
2nd strand synthesis
End repair
dA-Tail
Ligate adaptors
Size select
Excise dUTP
Amplify
Bioanalyze with DNA HS Chip

Problem:

Bioanalysis with a DNA HS Chip, reveals a double hump in every library product.



Notes:

Libraries are individual i.e. not yet pooled.
Yes Library 11 has a ~126bp adaptor dimer (already fixed).
100ng mRNA input
Qubit after cDNA synthesis (2nd strand) yields on average 8-10ng total
Qubit after final library completion, anything from 8-30ng total from 14 cycles.
Amplification step is definitely working (trace not shown).
I made many more libraries with this method and the double humps just keep on coming.

Questions:
1. What is happening here, any theories?
2. How do I fix this?


EDIT - 26 Nov 13

Guys I don't believe the double hump problems are related to the following examples:



PCR artifacts are normally twice the expected library size as shown in the image above. Since my libraries are between 200-500bp which is exactly the expected size I don't believe my results are PCR artifact related, but I'm happy to be proven wrong. Notice in my samples the double humps are within the main library peak and not two separate peaks.



There is definitely no indication of bead carry over in my traces.




Things I have tried but made no difference:
• Denaturing the libraries multiple different ways to prevent bird nesting or this so called “bubble effect” as suggested in other posts
• I have tried re-amplifying my libraries for another 2 more cycles as suggest in other posts
• Tried 12x, 13x, 14x, 15x, 16x, 17x cycles.

Last edited by AndyG; 11-25-2013 at 05:45 PM. Reason: More information added
AndyG is offline   Reply With Quote
Old 11-25-2013, 07:55 AM   #2
ZWB
Member
 
Location: Livermore, CA

Join Date: Apr 2011
Posts: 27
Default

In my experience, the double hump is the result of over amplification during your final PCR step. I would try 12 cycles instead of 14 and see if that gets rid of the second hump.
ZWB is offline   Reply With Quote
Old 11-25-2013, 09:49 AM   #3
microgirl123
Senior Member
 
Location: New England

Join Date: Jun 2012
Posts: 196
Default

The double peak is often a "bubble peak," which is often seen when you do lots of rounds of PCR. Basically, two strands of DNA have annealed that don't match in the middle so they bubble out and run strangely on the Bioanalyzer chip. It's not a problem for Illumina instruments (I don't know about others) because your DNA is denatured before loading.
microgirl123 is offline   Reply With Quote
Old 11-25-2013, 03:05 PM   #4
AndyG
Junior Member
 
Location: Australia

Join Date: Jun 2013
Posts: 4
Default

Quote:
Originally Posted by ZWB View Post
In my experience, the double hump is the result of over amplification during your final PCR step. I would try 12 cycles instead of 14 and see if that gets rid of the second hump.
Quote:
Originally Posted by microgirl123 View Post
The double peak is often a "bubble peak," which is often seen when you do lots of rounds of PCR. Basically, two strands of DNA have annealed that don't match in the middle so they bubble out and run strangely on the Bioanalyzer chip. It's not a problem for Illumina instruments (I don't know about others) because your DNA is denatured before loading.

Thanks guys, I just updated my first post with a little bit more information to clarify things.
AndyG is offline   Reply With Quote
Old 11-28-2013, 02:21 PM   #5
AndyG
Junior Member
 
Location: Australia

Join Date: Jun 2013
Posts: 4
Default

Anybody have any ideas?

I have seen some people with the same problem on this forum but never a solution.

Has anyone sequenced libraries with these double humps within the main library peak?
AndyG is offline   Reply With Quote
Old 11-29-2013, 04:03 AM   #6
WhiteSeal
Member
 
Location: Netherlands

Join Date: Jul 2013
Posts: 13
Default

I work with the illumina TruSeq DNA kit (with gDNA to start with) and I have seen the double peaks (in all variations) a lot. Have done lots of samples on the HiSeq and some on the MiSeq as well, but never had any problems sequencing.
WhiteSeal is offline   Reply With Quote
Old 11-30-2013, 10:29 PM   #7
chiaraf
Member
 
Location: Italy

Join Date: Oct 2012
Posts: 10
Default

With truseqRNA I often see these libreries traces but I always sequence them with no problem.
That's what Illumina suggests:

How to fix bubble products
To confirm this phenotype re-anneal the library slowly:
1. Denature the library by heating to 95 degrees
2. Slowly re-nature by cooling a degree a minute in a thermocycler by removing a heat block from an incubator and allowing it to cool to room temperature
The bubble products will often resolve into normal double-stranded library
In any case also without denaturation they are ok to be used and sequenced
chiaraf is offline   Reply With Quote
Old 12-11-2014, 01:17 PM   #8
Kbb262
Junior Member
 
Location: Houston

Join Date: Jul 2011
Posts: 1
Default

AndyG, I know this post is a little bit late, but were you able to find out the problem? I've been getting these double peaks for my Illumina ChIP-seq preps and they are not artifacts or bubble products (they persist through denaturation and run similar on a RNA bioA chip). These fragments do not amplify during qPCR and also do not sequence, but yet are somehow amplified enough during library prep to appear as a significant peak. We are using a pippin prep to size select (which works extremely well) and these peaks appear outside our selected size range.
Kbb262 is offline   Reply With Quote
Old 05-19-2015, 08:21 AM   #9
drops
Junior Member
 
Location: Barcelona

Join Date: Nov 2011
Posts: 4
Default double peaks

Hi guys,
how to quantify those "double peak" results in the bioanalyzer? Take both peaks together or just measure the first one?

Is it possible to do a Kappa quantification? Should be not a problem, right? Qubit measurements I should not trust, I guess. Same as nanodrop.
Any thoughts on that?

thx
D
drops is offline   Reply With Quote
Old 05-19-2015, 10:54 AM   #10
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,313
Default

Quote:
Originally Posted by drops View Post
Hi guys,
how to quantify those "double peak" results in the bioanalyzer? Take both peaks together or just measure the first one?

Is it possible to do a Kappa quantification? Should be not a problem, right? Qubit measurements I should not trust, I guess. Same as nanodrop.
Any thoughts on that?

thx
D
Yeah, if you are going to do qPCR, ignore the second peak. Use the small peak for the amplicon size. It works fine.

But normally it is better to just cut back on your PCR cycles if you start seeing these.

--
Phillip
pmiguel is offline   Reply With Quote
Old 05-19-2015, 11:21 AM   #11
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,313
Default

Quote:
Originally Posted by AndyG View Post
Method:

Wang et al. 2011, A low-cost library construction protocol for Illumina-based strand-specific multiplex rna-seq. PLoS One 6(10) e26426. doi:10.1371/journal.pone.0026426

Basics steps:

Extract Total RNA
Extract mRNA
Fragment
1st strand synthesis
2nd strand synthesis
End repair
dA-Tail
Ligate adaptors
Size select
Excise dUTP
Amplify
Bioanalyze with DNA HS Chip

Problem:

Bioanalysis with a DNA HS Chip, reveals a double hump in every library product.



Notes:

Libraries are individual i.e. not yet pooled.
Yes Library 11 has a ~126bp adaptor dimer (already fixed).
100ng mRNA input
Qubit after cDNA synthesis (2nd strand) yields on average 8-10ng total
Qubit after final library completion, anything from 8-30ng total from 14 cycles.
Amplification step is definitely working (trace not shown).
I made many more libraries with this method and the double humps just keep on coming.

Questions:
1. What is happening here, any theories?
2. How do I fix this?


EDIT - 26 Nov 13



PCR artifacts are normally twice the expected library size as shown in the image above. Since my libraries are between 200-500bp which is exactly the expected size I don't believe my results are PCR artifact related, but I'm happy to be proven wrong. Notice in my samples the double humps are within the main library peak and not two separate peaks.

[...]

Things I have tried but made no difference:
• Denaturing the libraries multiple different ways to prevent bird nesting or this so called “bubble effect” as suggested in other posts
• I have tried re-amplifying my libraries for another 2 more cycles as suggest in other posts
• Tried 12x, 13x, 14x, 15x, 16x, 17x cycles.
I don't think there is much chance that "AndyG" is still monitoring this thread, but just in case anyone reading this now would be helped:

Those double peaks are almost certainly a main peak and a "bubble product" peak. AndyG's protestations to the contrary suggest he doesn't get the concept.

Bubble products are said to form when adapters from 2 product strands anneal to each other. Because a great number of different inserts will comprise a library, it is unlikely that any 2 product strands will be complementary in the middle. So you end up with the "bubble product", a molecule annealed and double stranded at either end but the non-complementary middle section remains as a double single-strand "bubble" domain.

As you might imagine, this floppy thing migrates more slowly than its more compact fully double-stranded brethren. An Agilent chip will not correctly estimate its molecular weight by a variable amount. The fact that the two peaks overlap one another is not a sign that one is not a bubble peak.

Normally bubble peaks cause no issues for sequencing as long as you use qPCR to estimate the concentration of your libraries.

But, if you see bubble peaks, it is an indication that your PCR primer concentration has become limiting in the PCR reaction. At high primer concentrations, the primer annealing to a product strand will be kinetically favored over two product strands annealing.

Strand melting your library and allowing the strands to find their correct complements is a waste of time. Don't do that.

If you can't stand the idea of bubble products in you sample, you could probable get rid of most of them by adding your library to a fresh PCR reaction, with lots of primer and then doing a single cycle -- starting with a denaturation. This would allow the high amounts of primers to anneal and be extended into a double-stranded product.

But this is also a waste of time. Again, the bubble products don't actually cause problems.

They are an indication that you used more PCR cycles than were necessary. So you could cut back the number of cycles next time you made libraries.

--
Phillip
pmiguel is offline   Reply With Quote
Old 05-20-2015, 05:45 AM   #12
NextGenSeq
Senior Member
 
Location: USA

Join Date: Apr 2009
Posts: 482
Default

It's a heteroduplex. Decades ago people used heteroduplex analysis for mutation screening.

https://www.nationaldiagnostics.com/...uplex-analysis
NextGenSeq is offline   Reply With Quote
Old 05-20-2015, 06:24 AM   #13
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,313
Default

Quote:
Originally Posted by NextGenSeq View Post
It's a heteroduplex. Decades ago people used heteroduplex analysis for mutation screening.

https://www.nationaldiagnostics.com/...uplex-analysis
Uh, I think you are pushing the boundaries of the definition of "heteroduplex" here. But, same idea as far as migration goes. The large area of non-duplex double stranded DNA in the middle of the amplicon would be expected to retard migration though the matrix.

Of course this is largely supposition. I haven't seen anyone actually test the "bubble product" hypothesis by digesting a double peak sample with something like S1 nuclease to show that the 2nd peak disappears and is replaced by a large amount of annealed adapter ends 60bp peak.

Initially I had assumed that the retarded band was from "daisy chains" of products that were largely double stranded, but associated with each other by annealing at their adapters. That still seems like a possibility to me.

--
Phillip
pmiguel is offline   Reply With Quote
Old 07-28-2015, 11:49 PM   #14
EmiVig
Member
 
Location: Berlin

Join Date: May 2014
Posts: 10
Default

Hi pmiguel
I always find your post so helpful!!
Indeed, I would like to ask just one more clarification about the second HMW peak "bubble product" appearing in my ChIP-seq libraries prep as well...
You stressed more times that it shouldn´t be an issue for sequencing using Illumina platforms, as long as the quantification is performed by qPCR.
I´ve submitted my ChIP-seq libraries but it didn´t work out so good (very low cluster density...). I´m afraid the problem might be the quantification step: I´ve quantified them by Qubit, because our facility requests this kind of quantification (our facility has never optimized the flow cell loading based on qPCR...).
Could you give me some more hints about qPCR quantification vs Qubit!?
I´m struggling since too long with this ChIP-seq...I need to make it work!!!
Thank you!
Emilia
EmiVig is offline   Reply With Quote
Old 07-29-2015, 04:28 AM   #15
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,313
Default

Hi Emilia,
It is strange that your cluster density is too low. Are you sure that it is really too low? When you really overload a lane the cluster-calling software is overwhelmed and can report a low density.

How to tell the difference between a truly low cluster density and an a badly overclustered lane? If it is truly low cluster density, usually your pass-filter% and overall quality values will be very high. If your lane was massively over-clustered then the pass-filter % and quality values will be low.

Here is why I think you may have over-clustered. If you take a library with a large component of "bubble-products" and check its concentration with a fluorimeter using a double-strand-specific dye, then the bubble products will be under-reported, right? The central portion of each bubble product amplicon will have completely un-annealed central portions and thus be effectively single-stranded. However these bubble products each contain 2 amplicons, each capable of clustering. This should cause you to underestimate the true concentration of your library using a Qubit if a large amount of your library is bubble products.

Of course this is all wanton speculation on my part. There are a dozen other factors I don't know about that could be causing your issue.

--
Phillip
pmiguel is offline   Reply With Quote
Old 08-24-2015, 07:36 AM   #16
EmiVig
Member
 
Location: Berlin

Join Date: May 2014
Posts: 10
Default

Hi pmiguel,

thank you very much for your hilightening reply and here I am for an updating!
As suggested, I measured my histone ChIP-seq libraries using the KAPA kit (I used the DNA standards provided by the kit and I followed the instructions for size-adjustement) and I ended up with almost double the concentrations than the Qubit, therefore this would sustain the hypothesis of flowcells overloading. Without size-adjustment instead, the concentration resulted instead lower. So my first question is: is it right to perform a size adjustment? I would do it, since indeed my libraries have a different average size then the DNA standards.
At the same time, I've generated a standard curve using PhiX, as suggested by Illumina, because our sequencing facility tested the cluster generation using PhiX. I obtained a lower concentration (without size-adjustment even lower...) than what I obtained initially with the Qubit!!
At the end, I got back two opposite results ...I'm a bit stuck now...I would be more prone to believe to the high concentrations measured by KAPA qPCR (using KAPA DNA standards and size-adjustemnt), since I might have a good portion of "ssDNA bubble product" which the Qubit is not able to detect ...but how could I be sure of that? I'm thinking of sequencing in parallel the samples taking in consideration one or the other concentrations...but it would be a waste of material/time/money..
Moreover, last time I multiplexed 3 samples per lane...might also the low-complexity being an issue? I would like to barcode 6 samples per lane this time.
Please, find attached a summary of the concentration values obtained with the different methods + I also attached the Sequenicng Analyzer Viewer of my failed sequencing (lane 7 and 8). I'll attached as soon as possible the Bioanalyzer traces of the libraries.
I hope I'm not bothering too much, but I think this might also be helpful to somebody else encountering similar problems...
I would really appreciate any help! Thank you so much in advance!

Emilia
Attached Files
File Type: pdf Seq_analysis_Viewer_July_2015.pdf (140.9 KB, 64 views)
File Type: pdf Summary_ChIPseq_conc_240815.pdf (27.8 KB, 40 views)
EmiVig is offline   Reply With Quote
Old 08-24-2015, 08:18 AM   #17
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,313
Default

Hi Emilia,
With the additional information you provide I'm 90% certain that your samples were overclustered. My advice is to trust the KAPA results an size adjust based on the size of the shorter-length peak.
If you want to know whether high bias is an issue, then set your SAV "data by cycle" pane to "% base" and it will be immediately apparent.

--
Phillip
pmiguel is offline   Reply With Quote
Old 08-24-2015, 08:34 AM   #18
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,313
Default

By the way, don't trust the concentration of Illumina phiX, it is always much below the 10nM it is supposed to be. The KAPA standards work fine.

--
Phillip
pmiguel is offline   Reply With Quote
Old 10-24-2015, 03:43 PM   #19
aelhage
Junior Member
 
Location: Edinburgh

Join Date: May 2014
Posts: 1
Default

Dear all
I do have double peaks in my dsDNA HIGH SENSITIVITY bioanalyser traces of rna-seq libraries (prepared using NEBNext ultradirectional RNA library prep for illumina cat. no. E7420). The smallest peak around 250-300bp and the largest one around 1500bp.
I can understand from your very interesting posts that I should send these libraries for illumina sequencing providing accurate measurements of DNA concentrations would be performed by KAPA q-PCR kits? I hope I did get this right?
Alternatively, I could add extra primers and perform a second standard PCR on my library with only 1 cycle with a 2 min denaturation step, hoping that the extra peak would disappear!?
Thanking you very much for the great advice you are providing!
Best wishes
AEH

Last edited by aelhage; 10-26-2015 at 12:44 AM.
aelhage is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 09:52 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2019, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO