Old 08-11-2010, 07:03 AM   #1
krobison
Senior Member
 
Location: Boston area

Join Date: Nov 2007
Posts: 747
Default Library quantification: opinions?

Am curious to survey how groups quantify their libraries prior to putting them on the sequencer (or sending them to a vendor, as is our case).

qPCR would generally seem to be the winner to me; very high sensitivity & accuracy and you are measuring only the active molecules

However, it seems that Bioanalyzer and NanoDrop have their adherents. BioAnalyzer does get get you some size distribution information as well, but my impression is that with low concentration libraries one is pushing the sensitivity limits of both BioAnalyzer & NanoDrop (with a resultant hit in the precision of the measurement).
krobison is offline   Reply With Quote
Old 08-11-2010, 08:06 PM   #2
HESmith
Senior Member
 
Location: Bethesda MD

Join Date: Oct 2009
Posts: 509
Default

You're correct that qPCR is the most sensitive and accurate. However, we've gotten fairly reproducible results using PicoGreen dye fluorimetry for quantification. We switched from Nanodrop b/c, in addition to poor sensitivity, some libraries contained contaminants that absorbed in the UV range. The dye is specific for dsDNA, and accurate for samples down to at least 0.4 ng/ul. We follow up with BioAnalyzer high-sensitivity DNA chips for size determination and to detect adapter dimer contamination (which will affect qPCR measurements). However, its concentration values can vary 2-3X from reality.
HESmith is offline   Reply With Quote
Old 08-12-2010, 08:05 AM   #3
GW_OK
Senior Member
 
Location: Oklahoma

Join Date: Sep 2009
Posts: 411
Default

The Nanodrop is no good for exactly the reasons HESmith gave.
The Bioanalyzer is good for sizing, but only gives ballpark quantification I've found.
Fluorometry is acceptable with some variability. I've used the Qubit system.
qPCR is the very best overall and is now what I use exclusively (with the BioA for sizing). I use the Kapa Library quantification kits on a Roche 480.
GW_OK is offline   Reply With Quote
Old 08-12-2010, 08:39 AM   #4
lee
Junior Member
 
Location: Toronto

Join Date: Jun 2008
Posts: 2
Default

We use Bioanalyzer for sizing only.

Kapa qPCR for Illumina library quantification. As of August 1, Kapa kits are now compatible with the small RNA and GEX libraries. We consistently hit our target of 300-400K clusters/tile.

SOLiD qPCR assay (from Life Tech) for SOLiD library quantification.
lee is offline   Reply With Quote
Old 08-12-2010, 02:44 PM   #5
VanessaS
Member
 
Location: Dallass

Join Date: Nov 2009
Posts: 49
Default

We also use qPCR for quantification and the Bioanalyzer for sizing. We use the nanodrop for other parts of prep.
VanessaS is offline   Reply With Quote
Old 08-13-2010, 11:56 AM   #6
Baseless
Member
 
Location: Germany

Join Date: Feb 2010
Posts: 32
Default

Bioanalyzer for sizing, pico green for quantitation. We will test the qPCR once that kit arrives. I really would not miss titration if pPCR is reproducible and hits the right numbers...
Baseless is offline   Reply With Quote
Old 08-17-2010, 07:55 AM   #7
janejane
Member
 
Location: Midwest

Join Date: Aug 2010
Posts: 15
Default

Quote:
Originally Posted by lee View Post
We use Bioanalyzer for sizing only.

Kapa qPCR for Illumina library quantification. As of August 1, Kapa kits are now compatible with the small RNA and GEX libraries. We consistently hit our target of 300-400K clusters/tile.

SOLiD qPCR assay (from Life Tech) for SOLiD library quantification.
I saw that Kapa also has a SOLiD library quantification kit, does anyone ever compared it with the ABI qPCR reagents?
janejane is offline   Reply With Quote
Old 10-07-2010, 02:59 AM   #8
suludana
Member
 
Location: Spain

Join Date: May 2008
Posts: 61
Default

Quote:
Originally Posted by lee View Post
We use Bioanalyzer for sizing only.

Kapa qPCR for Illumina library quantification. As of August 1, Kapa kits are now compatible with the small RNA and GEX libraries. We consistently hit our target of 300-400K clusters/tile.

SOLiD qPCR assay (from Life Tech) for SOLiD library quantification.

Hi Lee,
could you please tell me your experience with the Kappa standards? I am working with Kapa kits to perform the qPCR but not with their standards.
Thanks
suludana is offline   Reply With Quote
Old 10-11-2010, 03:57 AM   #9
anunn
Junior Member
 
Location: London

Join Date: Sep 2010
Posts: 7
Default

I'm trying to quantify a library produced using the DSN normalisation protocol, which looks to be very low concentration...

Is it necessary to use the kit from Kapa Biosystems or could I use the qPCR reagents we have in the lab (SYBR GreenER from Invitrogen on an Applied Biosystems 7500)?

Many thanks,

Adam
anunn is offline   Reply With Quote
Old 04-08-2011, 02:18 AM   #10
ktabbada
Junior Member
 
Location: UK

Join Date: Sep 2009
Posts: 7
Default

Our lab uses both Bioanalyzer and qPCR (the Taqman assay described in Quail et al, 2009 -- Current Protocols) for quality control and quantitation. As HESmith observed, the Bioanalyzer quantitation results are often 2-3X higher than qPCR values. I'm relieved to find that we aren't the only ones getting results of this type. However, we have occasionally had libraries where the Bioanalyzer quantitation was 10x higher than qPCR. Upon titration we found that the qPCR value gave a closer approximation of the actual concentration of cluster-forming fragments in the library.

However, I was wondering if anyone else has observed this (Bioanalyzer quantitation 10 times higher than qPCR) and if they can suggest any causes for this?

Thanks in advance,

KAT
ktabbada is offline   Reply With Quote
Old 04-08-2011, 03:06 AM   #11
TonyBrooks
Senior Member
 
Location: London

Join Date: Jun 2009
Posts: 298
Default

I think the accuracy is also dependent on the library prep method. Generally, the more PCR, the more accurate it's going to be to quantify by Bioanalyser/Picogreen. I think this is because amplification only occurs on those DNAs that have the correct adapter sequences ligated. Unfortunately, more PCR also means more PCR duplicates.

We mainly run target enriched samples that have undergone quite a few cycles of PCR (sometimes over 20 cycles in total) and the Bioanalyser plus Picogreen has been fairly consistent. When we run ChIPSeq and MeDipSeq we've seen more variablility, probably because the protocols used have limited PCR (4 cycles).
Running 454 genomic libraries show the most variation as there is no PCR involved, so you're at the mercy of your ligation efficiency - we would always quantify using qPCR for these libraries as the variation is way too high. For example plotting the estimated concentration of using qPCR vs Bioanalyser for our last 20 libraries gave an R^2 of 0.43, with Bioanalyser quantitation between 2 and 6 times higher than qPCR. Alternatively 454 amplicon libraries can be quantified by Bioanalyser as in theory, all molecules in the library should contain the adpaters - we found these to be consistent when using Bioanalyser.
TonyBrooks is offline   Reply With Quote
Old 04-14-2011, 03:46 AM   #12
Morty
Junior Member
 
Location: Norway

Join Date: Jan 2010
Posts: 1
Default

We have used the Kapa 454 kit for both Rapid libraries (gDNA & cDNA) and for amplicons for some time now, but its not working perfectly. Well, it work fine for RL, but amplicons is more of a challenge. As we are a core facility people send us amplicons produced with fusion primers, nested PCR and ligation products. Even if qPCR results and DNA chip results are in agreement, the emPCR often give very varying enrichment numbers, and we really have no good explanation for it, but we see a differense between fusion and nested generated products. Im considering skipping qPCR for amplicons and stick to chip + qubit.

Cheers!
M.
Morty is offline   Reply With Quote
Old 04-16-2011, 11:15 AM   #13
Aaron Cooper
Member
 
Location: Los Angeles

Join Date: Feb 2011
Posts: 15
Default

I think that if you have access to a qPCR machine, there really is no reason not to do qPCR.

I prefer probe-based qPCR to SYBR Green qPCR, though, since with a probe the signal is not affected by the amplicon length. This allows you to use the same standard curve for all of your libraries.
Aaron Cooper is offline   Reply With Quote
Old 07-18-2011, 07:37 AM   #14
anant
Junior Member
 
Location: malaysia

Join Date: May 2010
Posts: 2
Default

I am wondering is any one observed that the qPCR reading is higher than the Qubit? I mean this is impossible right as the qPCR only measure the molecule which ligate with adapter.
anant is offline   Reply With Quote
Old 07-18-2011, 08:50 AM   #15
Aaron Cooper
Member
 
Location: Los Angeles

Join Date: Feb 2011
Posts: 15
Default

Quote:
Originally Posted by anant View Post
I am wondering is any one observed that the qPCR reading is higher than the Qubit? I mean this is impossible right as the qPCR only measure the molecule which ligate with adapter.
I think if you're using the stock Illumina adapters and primers, qPCR should be accurate. It has worked very well so far for us. Are you sure you are considering the correct total size for your library and not just insert size? Also, what are you using for a qPCR standard?

I have gotten fewer spots than expected from qPCR, but this was when using a custom read 1 primer.
Aaron Cooper is offline   Reply With Quote
Old 07-18-2011, 09:09 AM   #16
niceday
Member
 
Location: cambridge

Join Date: Apr 2010
Posts: 68
Default

We do get qPCR results higher than Qubit and bioanalyser. I spoke to some-one at Kapa and they did say others had observed this. We use the qPCR as standard and we get reproducible results.
niceday is offline   Reply With Quote
Old 07-18-2011, 09:31 AM   #17
HESmith
Senior Member
 
Location: Bethesda MD

Join Date: Oct 2009
Posts: 509
Default

Quote:
Originally Posted by anant View Post
I am wondering is any one observed that the qPCR reading is higher than the Qubit? I mean this is impossible right as the qPCR only measure the molecule which ligate with adapter.
The Qubit is specific for dsDNA. If you sample contains ssDNA with adapter ends, it will be detectable by qPCR but not Qubit. (Note: it will also form clusters, which makes qPCR more accurate for quantification.)
HESmith is offline   Reply With Quote
Old 07-20-2011, 05:12 AM   #18
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Quote:
Originally Posted by HESmith View Post
The Qubit is specific for dsDNA. If you sample contains ssDNA with adapter ends, it will be detectable by qPCR but not Qubit. (Note: it will also form clusters, which makes qPCR more accurate for quantification.)
In addition to ssDNA amplicons in your library, adapter dimers (ssDNA or dsDNA) could also be a source of higher molar concentration readings from qPCR versus fluorimetry.

Adapter dimers (zero or very short insert amplicons) will form clusters but will be detected at different sensitivities by different QC assay methods. Probe-based qPCR would likely detect them accurately. SYBR-green based qPCR might underestimate their molar concentration.

As detailed up-thread, Agilent 2100 Bioanalyzer results will detect non-amplicons (one or zero adapter ends) the same as amplicons -- other than the (probably minor) shift in length. Other confounding aspects include:

(1) ssDNA runs slower than one might expect on a High Sensitivity DNA Bioanalyzer chip. (Probably also on other DNA Bioanalyzer chips). Run a lane of nano RNA ladder on your DNA chip to get an idea. Anyway, this can be confusing under some circumstances.

(2) dsDNA runs faster than one might expect on RNA (single-stranded) Bioanalyzer chips. You can denature your sample (95 oC, 2 minutes followed by "snap chill" on ice) prior to loading, but the sometimes crazy results can be hard to interpret. Also, you may end up wondering whether your sample renatured, despite your efforts to prevent this from happening.

As far as fluorimetry goes, there are also single-stranded fluors, like ribo green, that can be used. But this can be a mine-field as well. The main, perhaps unexpected, issue is that ribo green fluoresces somewhat more brightly in the presence of double stranded DNA than single stranded. No where near the difference one sees with pico green, but still ~2x. We currently are trying to use ribo green, but using double-stranded DNA as our standard (phage lambda). We are denaturing our standard and our samples using the method mentioned above.

My assessment: all the library QC assays have significant issues. Beware.

--
Phillip
pmiguel is offline   Reply With Quote
Old 10-03-2011, 12:15 AM   #19
kulukulas
Junior Member
 
Location: Denmark

Join Date: Oct 2010
Posts: 4
Default

Quote:
Originally Posted by pmiguel View Post
(1) ssDNA runs slower than one might expect on a High Sensitivity DNA Bioanalyzer chip. (Probably also on other DNA Bioanalyzer chips). Run a lane of nano RNA ladder on your DNA chip to get an idea. Anyway, this can be confusing under some circumstances.
do you expect to detect ssDNA in DNA high sensitivity assay at all? isn't it a dsDNA binding dye used there?
kulukulas is offline   Reply With Quote
Old 10-03-2011, 06:42 AM   #20
pmiguel
Senior Member
 
Location: Purdue University, West Lafayette, Indiana

Join Date: Aug 2008
Posts: 2,317
Default

Quote:
Originally Posted by kulukulas View Post
do you expect to detect ssDNA in DNA high sensitivity assay at all? isn't it a dsDNA binding dye used there?
Regardless of what my expectations were, we have run the ssRNA ladder on a DNA high sensitivity chip. We did detect it.

See:

http://seqanswers.com/forums/showthread.php?t=12852

for details.

I should add that when we asked Agilent (prior to running this chip) what results we should expect when running single stranded molecules on a double stranded chip, we were told that the chip was not appropriate for that purpose. (Thanks for nothing, Agilent...)

--
Phillip
pmiguel is offline   Reply With Quote
Reply

Thread Tools

Posting Rules
You may not post new threads
You may not post replies
You may not post attachments
You may not edit your posts

BB code is On
Smilies are On
[IMG] code is On
HTML code is Off




All times are GMT -8. The time now is 08:52 PM.


Powered by vBulletin® Version 3.8.9
Copyright ©2000 - 2020, vBulletin Solutions, Inc.
Single Sign On provided by vBSSO